Hi Luze,
Based on your comments, I have amended the scope diagnostic check.
On the external ocsilloscope:
Channel 1 ( upper trace) set to 5V per division is plugged to 141T SCAN OUT
Channel 2 ( lower trace) set to 200mV/div is plugged to 141T VERTICAL OUT
Time /div is 100 msec /div, (exactly 2 * the 141T setting of 50ms/div) , so that both the 141T and external oscilloscope traces stay in sync
Trigger Source is set to Ch1, negative trigger on the SCAN OUT
Ch2 (being 141T Vertical out) is inverted, because the 141T IF section puts out a NEGATIVE signal 0 to -10V
These settings are better for diagnostic because we can now see more than one full scan, and the Output is aligned with the Scan.
I reduced the generator frequency to 2 MHz and set 141T BANDWIDTH 10 kHz and WIDTH 20 kHz /div to be similar to your RF section
That setting is not critical to just get a fat trace of the signal that is easy to find.
Here is a photo :
https://app.box.com/s/slydji1vuzj7d23sxw18xixbn7mn685gThe SCAN OUT starts at -5V and slopes to +5V at the end of that slope there is a sharp negative edge back to -5V which is used to trigger the external scope.
The Vertical output ( being the 2MHz signal generator) is within the SCAN OUT slope, and the position can be adjusted using the 141T FREQUENCY control on the RF Section.
Note on the SCAN OUT there is a period of flat -5V before the slope starts. I think this is for the Horizontal position adjustment.
I will leave this set-up for a few days in case you need any more.
Regards