For me digits and counts means here the same.
Fluke has 4 digits display while Brymen has 5 digits. So Brymen 30d practically means the same as Fluke 3 counts.
Let's plug it into the example. Do you mean that a 500mV @ 50-60Hz input on the BM789, with +-(0.5% + 30d) accuracy, would give a reading of 494.5-505.5mV?
No!
You calculated it like specification would be saying 300d while it says only 30d.
Did you noticed in linked by you datasheet and in what I have written that Brymen has 5 digits (while Fuke has 4)?
Did you noticed what I have written that from that (5 digits display compared to 4 digits display) 30d is like 3c for Fluke (means is practically the same error value)?
So for 500mV:
- ideal reading is 500.00
- with 0.5% is 502.50
- with 30d is 502.80
This is possible but what are you basing this opinion on that "counts" and "digits" are the same in this context?
I just can't imagine different interpretation so I use the only one that fits in my head