Electronics > Beginners
Is something wrong with my oscilloscope?
sincspecv:
--- Quote from: MarkF on August 06, 2019, 06:49:44 pm ---There is a calibration procedure. However, you probably don't have the tools necessary?
Starting on page 86:
http://web.mit.edu/8.13/8.13d/manuals/hp-54600a-54601a-oscilloscope-service-manual.pdf
--- End quote ---
You would be correct. I guess I'll just live with it until I can afford a nicer scope.
--- Quote from: rstofer on August 06, 2019, 06:57:29 pm ---Why do the images say DC Coupling?
--- End quote ---
I'm sorry, which images are you referring to? The calibration images are set to the appropriate coupling.
james_s:
I don't see anything super exotic in the calibration procedure, I'd probably just have the thing calibrated rather than replace the whole scope, assuming there's nothing serious wrong with it.
rstofer:
--- Quote from: sincspecv on August 06, 2019, 07:05:53 pm ---
--- Quote from: MarkF on August 06, 2019, 06:49:44 pm ---There is a calibration procedure. However, you probably don't have the tools necessary?
Starting on page 86:
http://web.mit.edu/8.13/8.13d/manuals/hp-54600a-54601a-oscilloscope-service-manual.pdf
--- End quote ---
You would be correct. I guess I'll just live with it until I can afford a nicer scope.
--- Quote from: rstofer on August 06, 2019, 06:57:29 pm ---Why do the images say DC Coupling?
--- End quote ---
I'm sorry, which images are you referring to? The calibration images are set to the appropriate coupling.
--- End quote ---
The images in Reply 5 which are purported to show AC Coupling but the image shows DC Coupling. I think...
sincspecv:
--- Quote from: rstofer on August 06, 2019, 09:26:36 pm ---
--- Quote from: sincspecv on August 06, 2019, 07:05:53 pm ---
--- Quote from: MarkF on August 06, 2019, 06:49:44 pm ---There is a calibration procedure. However, you probably don't have the tools necessary?
Starting on page 86:
http://web.mit.edu/8.13/8.13d/manuals/hp-54600a-54601a-oscilloscope-service-manual.pdf
--- End quote ---
You would be correct. I guess I'll just live with it until I can afford a nicer scope.
--- Quote from: rstofer on August 06, 2019, 06:57:29 pm ---Why do the images say DC Coupling?
--- End quote ---
I'm sorry, which images are you referring to? The calibration images are set to the appropriate coupling.
--- End quote ---
The images in Reply 5 which are purported to show AC Coupling but the image shows DC Coupling. I think...
--- End quote ---
The first one is AC and the second is DC.
Navigation
[0] Message Index
[*] Previous page
Go to full version