Author Topic: Multimeter rotary switch contacts  (Read 12761 times)

0 Members and 1 Guest are viewing this topic.

Offline PygmalionTopic starter

  • Regular Contributor
  • *
  • Posts: 76
  • Country: 00
Multimeter rotary switch contacts
« on: November 09, 2023, 04:25:39 pm »
I have a multimeter that I am very happy with, but lately it has been having problems measuring small resistances. The shortcut measurement sometimes gives 0.2Ω and sometimes tens of Ω.

After inspection of the multimeter, I have come to the conclusion that the problem are the contacts at the rotary switch. The rotary switch contains metal parts that connect copper strips on the circuit board.

My first idea was to clean the metal parts and copper strips with isopropanol and then spray everything with contact spray Kontakt 61.

When cleaning with isopropanol, I realised that the copper strips were already covered with a greasy black substance, and by cleaning them I removed some of it. I suspect that this is some kind of contact lubricant.

Now I do not know how to proceed. Should I find/buy some kind of contact lubricant or perhaps continue with the cleaning and replace this substance with contact spray? What is the best solution to achieve good contacts in this particular case?
 
The following users thanked this post: BILLPOD

Offline BILLPOD

  • Frequent Contributor
  • **
  • Posts: 682
  • Country: us
Re: Multimeter rotary switch contacts
« Reply #1 on: November 09, 2023, 07:27:59 pm »
Good Morning Pygmalion,   Please post the make and model of your multimeter, so others here on the forum, that have them can advise you.
     Also, compare readings on your meter with a known good meter,
and come back and let us know your progress.    :popcorn:
 

Offline TimFox

  • Super Contributor
  • ***
  • Posts: 11257
  • Country: us
  • Retired, now restoring antique test equipment
Re: Multimeter rotary switch contacts
« Reply #2 on: November 09, 2023, 07:57:10 pm »
Is your short-circuit measurement done with the usual meter leads or with a very short wire at the meter panel?
0.2 ohms is not out of the question for meter leads shorted together at the probe.
 

Offline PygmalionTopic starter

  • Regular Contributor
  • *
  • Posts: 76
  • Country: 00
Re: Multimeter rotary switch contacts
« Reply #3 on: November 09, 2023, 08:20:23 pm »
Is your short-circuit measurement done with the usual meter leads or with a very short wire at the meter panel?
0.2 ohms is not out of the question for meter leads shorted together at the probe.

I did shortcut tests with the usual meter leads. Usually I got 0.2Ω - 0.5Ω, which I think is satisfactory. But lately I have been getting from 0.2Ω up to tens of Ω. And when I touch the rotary switch, the value changes immediately.

So it's definitely a contact problem between the metal parts of the rotary switch and the copper tracks on the circuit board.

I am not asking for advice on how to find the source of the fault, but how best to improve the contact between the metal parts and the copper tracks.

The model name is therefore of little importance here IMHO.
 

Offline Jeff eelcr

  • Frequent Contributor
  • **
  • Posts: 372
  • Country: us
Re: Multimeter rotary switch contacts
« Reply #4 on: November 09, 2023, 11:42:02 pm »
Clean it and then apply a dilectric grease to the contacts.
Jeff
 

Offline DavidAlfa

  • Super Contributor
  • ***
  • Posts: 6845
  • Country: es
Re: Multimeter rotary switch contacts
« Reply #5 on: November 10, 2023, 04:53:15 am »
Clean, polish with a soft scotchbrite if you see grooves (Barely touch it until shiny, don't remove the gold layer).
If you see deep grooves and/or the copper, it'll be a bit too late, will affect low resistance readings.
Hantek DSO2x1x            Drive        FAQ          DON'T BUY HANTEK! (Aka HALF-MADE)
Stm32 Soldering FW      Forum      Github      Donate
 

Offline tooki

  • Super Contributor
  • ***
  • Posts: 14895
  • Country: ch
Re: Multimeter rotary switch contacts
« Reply #6 on: November 10, 2023, 02:12:33 pm »
The model name is therefore of little importance here IMHO.
You posted this in the Beginners subforum. If you’re already such an expert why even ask a question?

You were asked because it’s a potentially important piece of information. You don’t need to know or understand why it is important yet. And I don’t understand what you gain by withholding this information.
 

Offline donlisms

  • Frequent Contributor
  • **
  • Posts: 335
  • Country: us
Re: Multimeter rotary switch contacts
« Reply #7 on: November 10, 2023, 05:57:22 pm »
It's possible to help people without being a jerk. If you read his question and understood it, you'd realize it has nothing to do with the model of the meter.  He is right, you are wrong.
 

Offline tooki

  • Super Contributor
  • ***
  • Posts: 14895
  • Country: ch
Re: Multimeter rotary switch contacts
« Reply #8 on: November 11, 2023, 09:24:07 am »
It's possible to help people without being a jerk. If you read his question and understood it, you'd realize it has nothing to do with the model of the meter.  He is right, you are wrong.
No, I’m not wrong.

Since you also don’t seem to understand why it could matter:
- if a service manual is available, it might specify a particular lubricant
- sometimes, a given model of product has known issues that affect it uniquely. For example, the particular lubricant might break down a certain way with age. Or the contacts could be known to fail in a particular way.
- someone else here might have encountered the same fault in the same device. In particular, if the cause of the problem turns out not to be the switch contacts, but something else, the other person might know what it is.

But it also just irks me when people who are asking others for help are secretive about the device they’re working on. Unless it’s a custom piece of proprietary gear made for intelligence agencies (in which case they wouldn’t be asking here), there’s nothing to be gained by withholding that information. Meanwhile, doing so deprives those who want to help of potentially useful information. So an answer like “The model name is therefore of little importance here IMHO.” strikes me as a bit arrogant and entitled.
 
The following users thanked this post: JustMeHere, BILLPOD

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 16342
  • Country: de
Re: Multimeter rotary switch contacts
« Reply #9 on: November 11, 2023, 09:46:32 am »
With many DMM models the rotary switch is not directly in series with the measured resistor. It is kind of natural to have seprate paths for the voltage reading and test current. So moderate switch problems would not be an issue.  There may be a few bad designs though - so the model of the meter can mater.
A few tenths of an ohm is not uncommon with a handhelp DMM, or 2 wire ohm measurements in general.

The problem is more likely with the terminals or probes.
 
The following users thanked this post: tooki

Offline PygmalionTopic starter

  • Regular Contributor
  • *
  • Posts: 76
  • Country: 00
Re: Multimeter rotary switch contacts
« Reply #10 on: November 11, 2023, 01:50:44 pm »
The model is not a secret.  It is Aneng AN870 if this helps.

And I repeat, again, I do not complain that the resistance on connecting measuring cables is 0.2Ω.  I complain that sometimes the resistance is tens of Ω.  And since it is changed by pushing rotary switch, I can only conclude that the rotary switch contact is in series with the measured resistor.
« Last Edit: November 11, 2023, 01:58:02 pm by Pygmalion »
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 8568
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #11 on: November 11, 2023, 06:12:49 pm »
For rotary switch care and cleaning - it always descends into a religious or holy war about what to do.
There is cleaning, lubricating and oxidation to deal with, 3 aspects. The different switch constructions matter greatly on how to do all that.

I found the ANENG multimeter rotary switches make poor connection, they are some slag filled copper on ENIG finish on the PCB. Ohms readings way up there.
Factory looked like a mineral oil film was there on the PCB.
I tried - polishing the copper, silicone dielectric grease and a few other chemicals, eventually the fix is applying PPE like (In North America) CAIG DeoxIT F5 on the contacts/PCB.
The rotary switch works perfect for over 2 years now.

Clean it with IPA and q-tips. PAPER is a great abrasive to scrub off oxide without destroying and plating or surface finish. Anything coarser (sandpaper, scouring pad etc.) is destructive.
Fluke lube was mineral oil and I'll check but a light petroleum grease on the ratchet/shaft like Lubriplate.
Silicones are not preferred because they are like CFC's and can cause a contamination problem, I'd read parts near it can get affected.
WD40 is mostly kerosene and terrible, the solvent which attacks many things, evaporates and then you don't have decent lubrication. Some people love the stuff. ICK.
« Last Edit: November 11, 2023, 06:41:33 pm by floobydust »
 
The following users thanked this post: tooki, Pygmalion

Offline PygmalionTopic starter

  • Regular Contributor
  • *
  • Posts: 76
  • Country: 00
Re: Multimeter rotary switch contacts
« Reply #12 on: November 11, 2023, 09:15:52 pm »
I found the ANENG multimeter rotary switches make poor connection, they are some slag filled copper on ENIG finish on the PCB. Ohms readings way up there.
Factory looked like a mineral oil film was there on the PCB.
I tried - polishing the copper, silicone dielectric grease and a few other chemicals, eventually the fix is applying PPE like (In North America) CAIG DeoxIT F5 on the contacts/PCB.

Clean it with IPA and q-tips. PAPER is a great abrasive to scrub off oxide without destroying and plating or surface finish. Anything coarser (sandpaper, scouring pad etc.) is destructive.
Fluke lube was mineral oil and I'll check but a light petroleum grease on the ratchet/shaft like Lubriplate.

I can confirm that my Aneng also had some sort of black mineral oil, probably the same as your Aneng and Fluke. My idea was to clean with IPA and cotton swabs and then use Kontakt 61 lubricant (https://uk.farnell.com/kontakt-chemie/kontakt-61-400ml/lubricant-contact-61-400ml/dp/837982), which is probably something similar to CAIG. I think I came up with the same idea.
 
The following users thanked this post: tooki

Offline EPAIII

  • Super Contributor
  • ***
  • Posts: 1264
  • Country: us
Re: Multimeter rotary switch contacts
« Reply #13 on: November 12, 2023, 05:08:03 am »
I have seen this type of problem in several of my meters. I have had good luck using a product called DeoxIT D5. I have sprayed it on and exercised the switches. I did not wipe it off. Seems to work. I haven't needed to repeat the treatment on any of the meters where I used it.
Paul A.  -   SE Texas
And if you look REAL close at an analog signal,
You will find that it has discrete steps.
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 8568
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #14 on: November 12, 2023, 06:48:45 pm »
KONTACT CHEMIE has no mention of polyphenylene ether products. There were a few patents on PPE as well and it did cause a shakeup.
KONTACT make a lot of claims but no details. 61 appears to be a spray cleaner- solvent plus mineral oil lubricant. 60 not sure how it works.

So I would look at something else, like MG CHEMICALS 801C-P Contact Cleaner, Pen Applicator with Polyphenylene Ether or 801B.

I tried probably 5 different fixes and the cheeseball "copper" rotor quickly oxidizes and makes poor connections. Brymen rotary switch uses the same configuration, a copy of a 1989 Actron patent I mentioned here lol but better metal with better plastic tolerances.
I yearn for the Fluke phosphor-coated bronze, multi-leaf rotors - the best. But we all cheap now.
 

Offline Thunderer

  • Regular Contributor
  • *
  • Posts: 72
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #15 on: November 12, 2023, 06:56:18 pm »
As the OP was very reluctant to give the model, until later on... I guess is the time to look for a better multimeter. No need to invest more in fixing the unit than it was to purchase it  |O.

This is the contact in this cheapo instrument.


This is the other side of this ONE CHIP multimeter:
 
The following users thanked this post: tooki

Offline PygmalionTopic starter

  • Regular Contributor
  • *
  • Posts: 76
  • Country: 00
Re: Multimeter rotary switch contacts
« Reply #16 on: November 12, 2023, 07:39:10 pm »
As the OP was very reluctant to give the model, until later on... I guess is the time to look for a better multimeter. No need to invest more in fixing the unit than it was to purchase it  |O.

This is the contact in this cheapo instrument.

This is the other side of this ONE CHIP multimeter:

This judgemental comment is proof that I was rightly "reluctant" to provide a multimeter model.  :-//
« Last Edit: November 12, 2023, 07:47:46 pm by Pygmalion »
 

Offline Thunderer

  • Regular Contributor
  • *
  • Posts: 72
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #17 on: November 12, 2023, 08:52:10 pm »
Buddy, make it judgemental or not, by the time you realize spending money on cheap stuff costs you more money in the long run, you will have spent at least double of the cost of a decent multimeter.

You are not the first, not the last, to fall in the trap of expensive "cheapo instruments".

If you are on a budget, start looking at a UNI-T. I guarantee the cheapest/decent will last at least 10 years before you run into contacts problems.

NB: I said nothing about 3V multimeters, or unsafe ones. Dave (you know who) said enough.
 
The following users thanked this post: tooki

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 9326
  • Country: us
Re: Multimeter rotary switch contacts
« Reply #18 on: November 12, 2023, 09:15:17 pm »
by the time you realize spending money on cheap stuff costs you more money in the long run

If you are on a budget, start looking at a UNI-T

 :palm:

I don't see the issue with him cleaning up his contacts and putting a dab of something on it?  How is that a huge investment?  I putz around with junk all the time, not trying to justify my time at any hourly rate.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: tooki

Offline TimFox

  • Super Contributor
  • ***
  • Posts: 11257
  • Country: us
  • Retired, now restoring antique test equipment
Re: Multimeter rotary switch contacts
« Reply #19 on: November 12, 2023, 10:07:34 pm »
As the OP was very reluctant to give the model, until later on... I guess is the time to look for a better multimeter. No need to invest more in fixing the unit than it was to purchase it  |O.

This is the contact in this cheapo instrument.

This is the other side of this ONE CHIP multimeter:

This judgemental comment is proof that I was rightly "reluctant" to provide a multimeter model.  :-//

Internal construction of handheld DMMs varies quite a bit from model to model:  this is why people trying to give a useful answer want to know the model.
Why the reluctance?
 
The following users thanked this post: tooki

Offline Thunderer

  • Regular Contributor
  • *
  • Posts: 72
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #20 on: November 12, 2023, 10:56:49 pm »
I don't see the issue with him cleaning up his contacts and putting a dab of something on it?  How is that a huge investment?  I putz around with junk all the time, not trying to justify my time at any hourly rate.
Me neither, but when you do not have the cleaning agent, and you have to buy it, I guess it is far better to buy a decent one that does not need any maintenance. I bet the unit is 1-2 years old, and not even used on a daily basis.
 

Offline Jwillis

  • Super Contributor
  • ***
  • Posts: 1740
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #21 on: November 13, 2023, 12:11:28 am »
I have a multimeter that I am very happy with, but lately it has been having problems measuring small resistances. The shortcut measurement sometimes gives 0.2Ω and sometimes tens of Ω.

After inspection of the multimeter, I have come to the conclusion that the problem are the contacts at the rotary switch. The rotary switch contains metal parts that connect copper strips on the circuit board.

My first idea was to clean the metal parts and copper strips with isopropanol and then spray everything with contact spray Kontakt 61.

When cleaning with isopropanol, I realised that the copper strips were already covered with a greasy black substance, and by cleaning them I removed some of it. I suspect that this is some kind of contact lubricant.

Now I do not know how to proceed. Should I find/buy some kind of contact lubricant or perhaps continue with the cleaning and replace this substance with contact spray? What is the best solution to achieve good contacts in this particular case?
   

The black is just copper oxide. This is from moisture in the air oxidizing the copper surfaces and being scuffed off mechanically over a period of time. Probably from the contacts rather than the board. I can't tell but it looks like gold traces on the board. Isopropyl alcohol works great for cleaning that stuff off.
You can just use any Silicone based dielectric grease from an auto parts or hardware store. Mineral oil base seem to turn cloudy after time but work as well.  DeOxit F5 Fader works to but doesn't seem to last as long as the grease.Main purpose of the grease is not just for lubrication but also a barrier to moisture. 
Hope this helps and I'm glad to see you trying to keep stuff out of the land fills.
 
The following users thanked this post: tooki, Pygmalion

Offline tooki

  • Super Contributor
  • ***
  • Posts: 14895
  • Country: ch
Re: Multimeter rotary switch contacts
« Reply #22 on: November 13, 2023, 07:37:53 am »
As the OP was very reluctant to give the model, until later on... I guess is the time to look for a better multimeter. No need to invest more in fixing the unit than it was to purchase it  |O.

This is the contact in this cheapo instrument.

This is the other side of this ONE CHIP multimeter:

This judgemental comment is proof that I was rightly "reluctant" to provide a multimeter model.  :-//
Well, I don’t judge you for having an Aneng — I have an Aneng among my stable of meters.

The tone aside, the comment’s point of choosing how much to invest in a low-cost device is a very valid one. If you already have the suggested contact grease around, great! But it may not make economic sense to invest $25 in that grease for a meter that cost little more than that.

But the fact that others reported the same black grease in Aneng models and a detailed description of what did and didn’t work to fix the problem is precisely what I meant in the value of sharing model information.
 

Offline tooki

  • Super Contributor
  • ***
  • Posts: 14895
  • Country: ch
Re: Multimeter rotary switch contacts
« Reply #23 on: November 13, 2023, 07:40:53 am »
by the time you realize spending money on cheap stuff costs you more money in the long run

If you are on a budget, start looking at a UNI-T

 :palm:

I don't see the issue with him cleaning up his contacts and putting a dab of something on it?  How is that a huge investment?  I putz around with junk all the time, not trying to justify my time at any hourly rate.
I think the intent of the comment was that it may not make sense to spend $25 to buy that grease just for this purpose, if you don’t already have it.
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 8568
  • Country: ca
Re: Multimeter rotary switch contacts
« Reply #24 on: November 13, 2023, 06:59:03 pm »
Cleaning switches is part of owning any electronics. That $25 can I've used on a dozen pieces of equipment and it's still not empty.

When the maintenance costs are too high compared to the multimeter's price...  cheaper to bin it and buy a new one? Strange timeline this is.
 