Author Topic: HP 3325A error cal1.  (Read 554 times)
Offline hnjmklTopic starter
« on: April 06, 2022, 03:24:28 pm »
I have an HP3325A with errors cal1, cal2 and 3 pass, I have the operator-service manual, in which section does it help to investigate the defect?
Offline free_electron
    • SiliconValleyGarage
Re: HP 3325A error cal1.
« Reply #1 on: April 06, 2022, 03:44:58 pm »
page 8-31
« Reply #2 on: April 07, 2022, 06:24:53 am »
thank you, found the relevant section
« Reply #3 on: April 11, 2022, 05:26:17 pm »
After replacing the aborted F3, the 3325A passes the 3 tests without errors. I approached the calibration, at least in amplitude, excluding harmonics and spurious, not having an SA. I performed the D / A Converter Offset, up to punyto "e", after which I was unable to continue, point "f" I cannot execute it, at this point the explanation is not very clear.