Author Topic: new UT202A+ clamp meter - incorrect readings ?!  (Read 941 times)

0 Members and 1 Guest are viewing this topic.

Offline mike_mikeTopic starter

  • Frequent Contributor
  • **
  • Posts: 898
  • Country: ro
new UT202A+ clamp meter - incorrect readings ?!
« on: November 19, 2021, 04:47:56 am »
Hello! I have a new Unit UT202A+ clampmeter. I tested it and I found that the clamp does not work OK, because when I placed the masured wire right in the center of the clamp, then it reads, for example, 6A ac. But when I move the wire to the one of the extremities of the clamp then it shows me a difference of about 0.5-0.6A ac.
I found on various forums, than some of people said that this is normal and some people that this is not normal.
Please tell me what you think about this situation ?
« Last Edit: November 19, 2021, 07:54:32 pm by mike_mike »
 

Offline mike_mikeTopic starter

  • Frequent Contributor
  • **
  • Posts: 898
  • Country: ro
Re: new UT202A+ clampmeter - incorrect readings ?!
« Reply #1 on: November 19, 2021, 05:16:38 pm »
I made a few more tests and I found that:
If I measure the current through the wire while powering a 220V AC/1000W heater which also has a fan:
1. With the wire on the upper side of the clamp I=4.710A
2. With the wire on the center side of the clamp I=4.660A
3. With the wire on the lower part of the clamp (the most close to the meters body) I=4.419A

If I just close the clamp to the wire, while powering the same load as above:
1. With the upper side of the clamp I read I=0.022A
2. With the center side of the clamp I read I=0.099A
3. With the center side of the clamp I read I=0.134A

Please have a look and tell me what you think ?
 

Offline Manul

  • Super Contributor
  • ***
  • Posts: 1109
  • Country: lt
Re: new UT202A+ clamp meter - incorrect readings ?!
« Reply #2 on: November 19, 2021, 09:38:45 pm »
The wire must be passing through the inside of a closed clamp, like a rope through a ring. Some small errors are expected, clamp meter is not a precision instrument.
 
The following users thanked this post: mike_mike

Offline Benta

  • Super Contributor
  • ***
  • Posts: 5877
  • Country: de
Re: new UT202A+ clamp meter - incorrect readings ?!
« Reply #3 on: November 19, 2021, 09:43:31 pm »
You do realise, that only ONE wire must go through the clamp, right? You need to split the power cable to do the measurement.

 
The following users thanked this post: mike_mike

Offline mike_mikeTopic starter

  • Frequent Contributor
  • **
  • Posts: 898
  • Country: ro
Re: new UT202A+ clamp meter - incorrect readings ?!
« Reply #4 on: November 20, 2021, 09:04:55 pm »
Yes, I measured the current by splitting the cable.

I am still a little bit confused, and I don't know why it shows different readings if I move the wire inside the clamp ?
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1202
  • Country: us
Re: new UT202A+ clamp meter - incorrect readings ?!
« Reply #5 on: November 20, 2021, 11:53:53 pm »
The manual did say that you should have the wire at the center of the clamp. If you don't you get another 1% of inaccuracy although your readings show more than 1%.
 