Author Topic: Tektronix 2225 Channels showing up twice  (Read 1062 times)

0 Members and 1 Guest are viewing this topic.

Offline eloreTopic starter

  • Newbie
  • Posts: 2
Tektronix 2225 Channels showing up twice
« on: March 10, 2019, 06:05:13 am »
Hello all, I am new to the forum and to hobbyist electronics. Recently I picked up an older analog  oscilloscope to learn the basics on from ebay but I noticed something odd. The scope is a Tektronix 2225 and everything seems to be in order. I checked the knobs and buttons and they all seem to be working (as far as I can tell) but the channel signals show up twice. When I adjust the position knobs they move in pairs as if it is only one signal that is being doubled. When I adjust the seconds/division knob to the slower settings the signals appear to move slowly as expected but the duplicates appear to move a bit faster in comparison. This is the only difference I could tell between the two. The scope has not been calibrated for over a decade. I do not know if that has anything to do with it but I thought I’d share everything I know.

I attached a photo of what I mean. Both channels are on but the upper two lines are from channel 1 and the second two are from channel 2. Any input is appreciated thank you.
 

Offline macboy

  • Super Contributor
  • ***
  • Posts: 2287
  • Country: ca
Re: Tektronix 2225 Channels showing up twice
« Reply #1 on: March 10, 2019, 01:03:03 pm »
This is a dual (or delayed) timebase scope and you have that feature enabled. Turning this off may be as simple as pushing the horizontal timebase knob in (pull to enable). I do strongly encourage you to read about this feature in the manual, as it is a very powerful and desirable feature.
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23096
  • Country: gb
Re: Tektronix 2225 Channels showing up twice
« Reply #2 on: March 10, 2019, 01:12:20 pm »
2225 doesn’t have delayed timebase. It has mag instead. Above the timebase control switch it to X1 instead of alt and it will go away.

Nice scopes.
 
The following users thanked this post: elore

Offline eloreTopic starter

  • Newbie
  • Posts: 2
Re: Tektronix 2225 Channels showing up twice
« Reply #3 on: March 10, 2019, 04:22:59 pm »
I switched it to X1 instead pf ALT just like you said bd139 and it went away! I thought I got a dud there for a second! Shows I still have a lot to learn about oscilloscopes. Now I can start learning
 

Offline bd139

  • Super Contributor
  • ***
  • Posts: 23096
  • Country: gb
Re: Tektronix 2225 Channels showing up twice
« Reply #4 on: March 10, 2019, 04:55:12 pm »
Cool. Glad it worked  :-+
 

Online David Hess

  • Super Contributor
  • ***
  • Posts: 17117
  • Country: us
  • DavidH
Re: Tektronix 2225 Channels showing up twice
« Reply #5 on: March 10, 2019, 06:01:10 pm »
In my notes I refer to this feature as "alternate magnification".  When magnification is active, the magnified traces are displayed alternately with the main traces.  It is basically a poor man's alternate delayed sweep which is distinct from delayed sweep and mixed delayed sweep.  Alternate magnification was very rare or unique to the 2225.

2213/2213A   Single Delayed (!), Magnification
2215/2215A/2235   Alternate Dual Delayed, Magnification
2236/2236A   Alternate Delta Dual Delayed, Magnification
2220/2221/2221A   Single, Magnification

2225   Single, Alternate Magnification
2201/2205/2210/2211   Single, Magnification

465/464/466   Mixed Dual Delayed, Magnification
475/475A   Mixed Dual Delayed, Magnification
485   Alternate Dual Delayed, Magnification
465B/468   Alternate Dual Delayed, Magnification

Many of the various 465 models also optionally support Alternate Delta Dual Delayed.

Single delayed is like having a 1.5 delayed sweep.
 