Question: Isolation transformer, Ground floating or Ground shorted to Neutral
<< < (5/6) > >>
IanB:

--- Quote from: Gtx21 on September 03, 2018, 10:44:08 pm ---If the earth ground for the dut is shorted to the DUT neutral, or used in some way internally, then I have a problem.
 
The particular instance I ran into is where a heat sink was earth-grounded via a support screw to the case bottom, which was connected to the Emitter of a BJT, this bring the Neutral to Earth Ground.   In this situation, I got lucky and realized it before I started probing.  Had I not, poof goes something.
--- End quote ---

If the situation you describe occurs, then the device you are looking at has an electrical fault. No part of the live wiring of a device is permitted to be connected to earth ground. There are insulation testers that exist to check for this issue. They apply a high voltage between the combined L/N conductors of the device and the earth/ground conductor/chassis/casing and test for leakage current. A high resistance of 1 MΩ or more must exist for the device to pass the test, otherwise the device is unsafe and must not be used.

Rather than investing in a differential probe, it might be better to invest in an insulation tester so you can check for device faults before working on them.
perieanuo:
Ground shorted to neutral, NEVER! Bad bad practice.Saw this in Romania.



Envoyé de mon iPad en utilisant Tapatalk
Monkeh:

--- Quote from: perieanuo on September 04, 2018, 03:26:42 pm ---Ground shorted to neutral, NEVER! Bad bad practice.Saw this in Romania.

--- End quote ---

Ponder how neutral becomes neutral. It is, in fact, normal practice - when you want a neutral. There is no neutral without it.
perieanuo:

--- Quote from: Monkeh on September 04, 2018, 03:40:40 pm ---
--- Quote from: perieanuo on September 04, 2018, 03:26:42 pm ---Ground shorted to neutral, NEVER! Bad bad practice.Saw this in Romania.

--- End quote ---

Ponder how neutral becomes neutral. It is, in fact, normal practice - when you want a neutral. There is no neutral without it.

--- End quote ---
Neutral is never neutral.Usual in triphasic systems every phase is distributed to 1/3 of the customers hoping it gets the network balanced and the neutral to 0, but every triphase network is unbalanced, so Neutral is not 0,never ever.It's almost zero [emoji4]
Ground is earth plus 2-4 ohms, end of story,protects you from dying.Ground is mandatory imho everywhere, not only in wet rooms and dangerous electrical equipment.
So keep the ground connection continuous.
Regards,pierre


Envoyé de mon iPad en utilisant Tapatalk
Monkeh:

--- Quote from: perieanuo on September 04, 2018, 03:49:39 pm ---
--- Quote from: Monkeh on September 04, 2018, 03:40:40 pm ---
--- Quote from: perieanuo on September 04, 2018, 03:26:42 pm ---Ground shorted to neutral, NEVER! Bad bad practice.Saw this in Romania.

--- End quote ---

Ponder how neutral becomes neutral. It is, in fact, normal practice - when you want a neutral. There is no neutral without it.

--- End quote ---
Neutral is never neutral.Usual in triphasic systems every phase is distributed to 1/3 of the customers hoping it gets the network balanced and the neutral to 0, but every triphase network is unbalanced, so Neutral is not 0,never ever.It's almost zero [emoji4]
Ground is earth plus 2-4 ohms, end of story,protects you from dying.Ground is mandatory imho everywhere, not only in wet rooms and dangerous electrical equipment.
So keep the ground connection continuous.
Regards,pierre

--- End quote ---