Replace Multimeter Fuse
« on: July 17, 2020, 10:57:51 am »
Hello all,

I need to replace both fuses on a Hioki DT4282 which requires a 11A 1000V 10.3 x 38mm ceramic fuse and a 630mA 1000V 10.3 x 38mm ceramic fuse. I can only find the same make (HOLLY) on peaktech and they only seem to do up to 10A. Is is ok to use a 10A fuse instead of 11A?

Many thanks,

Rui
 

Offline wraper

  • Supporter
  • ****
  • Posts: 18238
  • Country: lv
Re: Replace Multimeter Fuse
« Reply #1 on: July 17, 2020, 11:21:23 am »
It's a specialized multimeter fuse and it does not necessarily need to be ceramic. For example you could use Busmann DMM-B-11A which is also sold as Fluke F-11A. Or Littlefuse 0FLU011.T, Siba 50 199 06.11, just don't buy it from dodgy places, ebay is full of counterfeits. 10A fuse most likely is simply a usual fast acting fuse which is not nearly as fast as multimeter fuse.
« Last Edit: July 17, 2020, 11:23:58 am by wraper »
 

Offline ruibcaTopic starter

  • Newbie
  • Posts: 5
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #2 on: July 17, 2020, 11:38:36 am »
Thank you wrapper. PeakTech seem to be a reputable company. I've ordered the 630mA 1000v from their "Fuses for Multimeters" section (https://peaktech-rce.com/en/fuses-for-multimeters/543-peaktech-7520-ceramic-fuse-ff-630ma-1000v-103x38mm-10-pieces.html).

What you're saying is that it's best not to use the 10A fuse if the meter asks for 11A? Even if on the outside of the meter is says 10A FUSED.

« Last Edit: July 18, 2020, 09:43:49 am by ruibca »
 

Offline ledtester

  • Super Contributor
  • ***
  • Posts: 3463
  • Country: us
Re: Replace Multimeter Fuse
« Reply #3 on: July 17, 2020, 11:45:04 am »
What is the fuse protecting?

Seems like it is just protecting the 10A current measurement input. So using a lower amperage fuse is fine -- it will only reduce the maximum current you can measure.
 
The following users thanked this post: ruibca

Offline wraper

  • Supporter
  • ****
  • Posts: 18238
  • Country: lv
Re: Replace Multimeter Fuse
« Reply #4 on: July 17, 2020, 11:49:23 am »
What is the fuse protecting?

Seems like it is just protecting the 10A current measurement input. So using a lower amperage fuse is fine -- it will only reduce the maximum current you can measure.
Nope, as I said multimeter fuses are much faster than usual fuses.
What you're saying is that it's best not to use the 10A fuse if the meter asks for 11A? Even if on the outside of the meter is says 10A FUSED.
It's multimeter nominal input rating, not fuse rating. Again, multimeter fuses are different from usual fuses, even if they are HRC type.
 
The following users thanked this post: tooki

Offline ruibcaTopic starter

  • Newbie
  • Posts: 5
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #5 on: July 17, 2020, 12:39:24 pm »
Seems like it is just protecting the 10A current measurement input.

Yes, it's just for the A input. The mA and µA have a 630mA socket (image attached).

It's multimeter nominal input rating, not fuse rating. Again, multimeter fuses are different from usual fuses, even if they are HRC type.

I don't mind if it can only handle 9A instead of 10A as I use it exclusively for small electronics projects. I just want to make sure it's safe to do so (for me and the multimeter). The ones I was looking into seem to be specific for multimeters and seem to be good quality - the same that came with the multimeter - and a good price for 10 pieces (https://peaktech-rce.com/en/fuses-for-multimeters/545-peaktech-7530-ceramic-fuse-ff-10a-1000v-103x38mm-10-pieces.html). Unfortunately they don't seem to do 11A.
« Last Edit: July 17, 2020, 12:58:27 pm by ruibca »
 

Offline Jwillis

  • Super Contributor
  • ***
  • Posts: 1740
  • Country: ca
Re: Replace Multimeter Fuse
« Reply #6 on: July 17, 2020, 01:20:37 pm »
You could try https://www.tester.co.uk/catalogsearch/result/?q=fuse&p=2  for the UK  .They have the  Fluke 803293 Fuse - A:11A (F), V:1000V, IR:20kA, 10.3x38.1 mm . When I checked they had one in stock.
 
The following users thanked this post: ruibca

Offline eblc1388

  • Frequent Contributor
  • **
  • Posts: 404
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #7 on: July 18, 2020, 03:16:16 am »
The question: "What is the fuse protecting?" is really an important one, as it determine what type of fuse is to be used.

For the fuse in series with the 10A current range, its main purpose is to protect the user and cut off the current when user connects the test leads across a voltage source by mistake. Its secondary function is to interrupt the current should a dumb user tried to pass excess current through the meter continously.

While the fault current in homes are usually limited, one can expect current of thousands of amps if the meter is used wrongly on system busbar.  This mistake would easily means live or dead for the user. Thus the critical data to check on the replacement fuse is fault interrupting capacity, expressed in KA. The replacement fuse should have the same or higher KA than the original. Chances are they are already being the highest KA type available.

With that done, then the voltage and current rating come into consideration.
 

Offline ruibcaTopic starter

  • Newbie
  • Posts: 5
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #8 on: July 18, 2020, 09:31:17 am »
Thank you eblc1388. The interrupting capacity is a useful bit of information. I was wondering what that was. On the original fuse there are values for AC and DC: l1=50KA AC and l1=30KA DC.
On the new one there's only one value: l1=50kA (image attached - old and new fuse). Not sure if the 50kA applies for both AC and DC in this case. But like I mentioned, I'm not planning to measure ac current any time soon. However, as I understand 4S LiPo batteries (14.8V) can put out a lot of current as well (possibly more than 10A).
So back to my original question, can I assume that, based on everything else being the same, it's safe to use a 10A fuse instead of a11A fuse? Many thanks.
« Last Edit: July 18, 2020, 09:54:32 am by ruibca »
 

Offline Per Hansson

  • Supporter
  • ****
  • Posts: 448
  • Country: se
Re: Replace Multimeter Fuse
« Reply #9 on: July 18, 2020, 10:06:46 am »
The important thing is that the fuse is fast so it interrupts before the tracks and semiconductors in the multi meter explodes.
This is designated by the "diode" icon on the fuse which is for the "aR" rating, which means semiconductor fuse.
That takes care of the protection of the multi meter and is also explained by Dave in the video linked below.
But the more important designation is breaking capacity as has already been discussed in this thread.
So that the fuse is also able to quench an plasma arc formed if you by mistake short out a high energy circuit.
So these are the two most important factors when selecting a multi meter fuse, not its ampere rating...

 
The following users thanked this post: tooki

Offline ruibcaTopic starter

  • Newbie
  • Posts: 5
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #10 on: July 18, 2020, 10:18:03 am »
Brilliant! I guess that sums up all the parameters in a fuse. Thank you for the insight Per Hansson and everyone else for their input.
This is my first post in this forum and I'm well impressed with the generosity!
 

Offline tooki

  • Super Contributor
  • ***
  • Posts: 13523
  • Country: ch
Re: Replace Multimeter Fuse
« Reply #11 on: July 18, 2020, 07:00:05 pm »
The important thing is that the fuse is fast so it interrupts before the tracks and semiconductors in the multi meter explodes.
This is designated by the "diode" icon on the fuse which is for the "aR" rating, which means semiconductor fuse.
That takes care of the protection of the multi meter and is also explained by Dave in the video linked below.
But the more important designation is breaking capacity as has already been discussed in this thread.
So that the fuse is also able to quench an plasma arc formed if you by mistake short out a high energy circuit.
So these are the two most important factors when selecting a multi meter fuse, not its ampere rating...


Indeed. I have a suspicion that the oddball 440mA and 11A values (which don’t seem to exist outside of DMM fuses) are specifically to discourage people from substituting other fuse types.
 

Offline slugrustle

  • Frequent Contributor
  • **
  • Posts: 293
  • Country: us
Re: Replace Multimeter Fuse
« Reply #12 on: December 01, 2024, 05:33:52 pm »
There are a couple of sellers on eBay with these Hollyland HC10aR fuses in stock for about $10 USD each.  I think the eBay tip came from an amazon review of the DT4282...

I found the datasheet at http://www.hollyfuse.com/UploadFile/230925ProductDLLixin/HC10aR%20250mA-32A%20HLDL-2307-02_74.pdf.

EDIT: The fuses from eBay did not appear to be genuine.  They had no UL recognized mark and different resistance from DT4282 original fuses. I was able to buy new fuses from a local partner of Hioki by filing a repair and calibration request through a link on the webpage https://www.hioki.com/us-en/support/service.
« Last Edit: March 04, 2025, 01:25:09 am by slugrustle »
 

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #13 on: February 24, 2025, 12:51:41 pm »
Hello guys,
I've just received an "original fuse" for Hioki DT4282 from Ebay. I don't really know if it's the real thing. Have some doubts. Can you take a look and give me your thougts ?
On the left, the blown original fuse.
On the right, Ebay fuse.
Thanks !
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #14 on: February 24, 2025, 05:37:16 pm »
I wonder if the fuse resistance has an effect on the meter accuracy. The shunt resistor for the 10A range is very small and a little resistance from the fuse can have big impact.
 

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #15 on: February 24, 2025, 05:55:13 pm »
The fuse in the pic is the 630 ma/1000v. The 10A one, measures 0.1 ohm.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 8868
  • Country: us
Re: Replace Multimeter Fuse
« Reply #16 on: February 24, 2025, 06:21:18 pm »
The question: "What is the fuse protecting?" is really an important one, as it determine what type of fuse is to be used.

It is an important question but you also have to know how it is achieving that goal.

The primary purpose of the fuse is to protect the user from thermal injuries from an accidental arc-flash event.  Without rehashing the entire process, an arc-flash may develop when you short out a high-energy mains power source, typically three-phase 480VAC service panels or other areas with high power connections.  The arc-flash is essentially a plasma cloud that continues to short out the source until the power is removed by an upstream protection device or the whole thing explodes or burns down.  This accident can happen if you mistakenly connect your DMM to a high-energy source with the leads in the current jacks and this is the scenario where the fuse is intended to step in and protect you from your own stupidity.

There are two places where the arc-flash is likely to start--at the tips of your probes or within the fuse itself as it blows. The arc-flash doesn't form instantly, but takes a few milliseconds to grow to the point where it continues.  If you can interrupt the fault quickly enough you will hopefully avoid the arc flash.  The fuse itself is sand-filled and designed to contain any arcs that arise within it.  However, the only way to protect the user from an arc-flash starting at the probe tips is to interrupt the current flow very quickly.  The typical 11A fuses used in Fluke and other decent DMMs has a rated "total clearing time" of 10ms at 200A.  This presumably decreases with higher currents, but the chart bottoms out at 10ms so I suppose nothing is guaranteed past that point. 

Multimeter-specific fuses are not unique regarding clearing time. There are other specialty fuses that behave similarly, such as solar protection fuses and other "fast" fuses that have a 1kV AC/DC and 10kA+ interrupt ratings.  Fuses with lower interrupt ratings are not going to perform as well regardless of their time/current curve because they will have a longer arc time.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 8868
  • Country: us
Re: Replace Multimeter Fuse
« Reply #17 on: February 24, 2025, 06:22:43 pm »
I wonder if the fuse resistance has an effect on the meter accuracy. The shunt resistor for the 10A range is very small and a little resistance from the fuse can have big impact.

It causes a burden voltage on the circuit being tested but shouldn't affect the meters ability to measure whatever current is flowing since the meter only looks at the voltage across the shunt.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: tooki

Offline Whales

  • Super Contributor
  • ***
  • Posts: 2236
  • Country: au
    • Halestrom
Re: Replace Multimeter Fuse
« Reply #18 on: February 24, 2025, 07:39:43 pm »
Hello guys,
I've just received an "original fuse" for Hioki DT4282 from Ebay. I don't really know if it's the real thing. Have some doubts. Can you take a look and give me your thougts ?
On the left, the blown original fuse.
On the right, Ebay fuse.
Thanks !

On one fuse the "630mA" and its box have the same line weight.  On the other the text is noticeably thicker than the box line.  This suggests that they have different art source files.

Still could be legit, no idea.  The letter kerning and serifs seems the same.

I'd normally recommend cutting them open to see if they have the same fill inside, but I expect you won't want to do that  ;)

Offline watchmaker

  • Frequent Contributor
  • **
  • Posts: 819
  • Country: us
  • Self Study in EE
Re: Replace Multimeter Fuse
« Reply #19 on: February 24, 2025, 07:55:18 pm »
I watched Dave's video on the DMM fuses and input protection.  Enlightening.

So, are counterfeit fuses (Littlefuse or Bussman) being sold on Amazon?  The price difference between Amazon and Digikey (as well as wait time) is enormous.

I have been very naive.

THANKS!

 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 8868
  • Country: us
Re: Replace Multimeter Fuse
« Reply #20 on: February 24, 2025, 09:31:36 pm »
So, are counterfeit fuses (Littlefuse or Bussman) being sold on Amazon?  The price difference between Amazon and Digikey (as well as wait time) is enormous.

Probably, but you can still find reputable places to buy them for non-stupid prices.  For the 11A 10x38 model, there are several different part numbers that Fluke used and they all look a bit different.  There's BUSS-labeled DMM-11A that appears in all my older (2000-2010s) Fluke meters, the Littelfuse FLU-11A that they are using now and an Eaton/Bussmann DMM-B-11A that I bought from Witonics but haven't seen in an actual meter yet (not that I should have...)  They all look a bit different and some of the DMM-11A units actually have the caps on a bit crooked but are not fake. 

These appear to have a list price from Fluke of $9.95, $12.62 or $21.00 depending on where you look.  You can get the latest FLU-11A from TEquipment for $8.46.  This isn't a $100 part and anyone showing you a price like that is playing some game or another. 

https://www.tequipment.net/FlukeFuses943121.html

Or Witonics on Amazon has a 5-pack of the DMM-B-11A for $20. 

https://www.amazon.com/Multimeter-DMM-11A-DMM11-Replace-Acting/dp/B07ZGW4J8C/ref=sr_1_2?dib=eyJ2IjoiMSJ9.YLFtw0Fr-MrVpUEvEyPZEneCunXodCBY4H9YuQ_RA-SAiXi9KsZ1DEkdtBnDNLJH6cPQFT7HkCeRgE9sFGTFaGkS53Zerf_t-NTC9sbh8N7DSApc5_qpYUB8mxodm7HLU-ZNH9Kuog28Aitx3mN14IdIrRzZA-VsCbFWn3qX7ANcJB9E8vQI9r03D4_i-sGKUdDApT41Sts87DF_i-dND6ZILg1WsqbU3gD82ef1Cn8.rToIo7qfKYxqPuXpFRDJ9e9FA4uajGf_LB-v-5x5eM4&dib_tag=se&keywords=fluke+fuse&m=A1UOBWSFLOXPFW&qid=1740432382&s=merchant-items&sr=1-2

I couldn't shorten that link without losing the sale price, IDK why.  It's not an affiliate link, at least not mine.
« Last Edit: February 24, 2025, 09:34:35 pm by bdunham7 »
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: watchmaker

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #21 on: February 24, 2025, 09:55:13 pm »
I wonder if the fuse resistance has an effect on the meter accuracy. The shunt resistor for the 10A range is very small and a little resistance from the fuse can have big impact.

It causes a burden voltage on the circuit being tested but shouldn't affect the meters ability to measure whatever current is flowing since the meter only looks at the voltage across the shunt.

Aha!Thank you! That's why when I use my DMM as current shunt to measure current with the scope it appears to be of higher value shunt than the specs said it is.
 

Offline inse

  • Super Contributor
  • ***
  • Posts: 1098
  • Country: de
Re: Replace Multimeter Fuse
« Reply #22 on: February 24, 2025, 10:03:46 pm »
It appears to me this discussion is going a little academic.
The type of fuse required depends very much on the use case of the measurement device.
Besides two CAT III/IV hand multimeters, I also own several benchtop multimeters (Keithley and HP) their input protection is a standard 5x20mm fuse.
Because those instruments are neither designed nor likely being carried into a transformer substation, thats why a standard device fuse is sufficient.
For a multimeter that is indended to be used (or cannot be excluded to be used) on high energy supply installations, a properly designed input protection is required.
So in my opinion if you solely plan to use your multimeter on the workbench, CATx rated fuses are not necessary, a standard fast blow device fuse is sufficient.
The LiPo battery you mentioned can deliver a lot of current but the voltage is low enough not to form a powerful arc.
Thats how I see it.
« Last Edit: February 25, 2025, 05:20:06 am by inse »
 

Offline BillyO

  • Super Contributor
  • ***
  • Posts: 2003
  • Country: ca
Re: Replace Multimeter Fuse
« Reply #23 on: February 25, 2025, 01:30:35 am »
So in my opinion if you solely plan to use your multimeter on the workbench, CATx rated fuses are not necessary, a standard fast blow device fuse is sufficient.
While I agree with this in concept, the problem is, if the meter is rated for CAT-4, like a high end Fluke, Brymen, etc.. that meter may someday get passed on to someone that will try to use on industrial switch gear by virtue of it's specifications and wind up in hospital with serious injuries.  The net-net here is, if you change the specification of the fuse, you change the specification/capability of the meter.
Bill  (Currently a Siglent fanboy)
--------------------------------------------------
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #24 on: February 25, 2025, 02:30:37 am »
I think it's best to replace the fuses with the correct ones and be careful not to blow them again.
 

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #25 on: February 25, 2025, 10:07:13 am »
I took some measures and now I'm quite sure it's a fake product (acording to Holly HC10aR datasheet). Moreover the crimping rings are missing on the fake fuse.
I've found the replacings, finally at
"https://peaktech-rce.com/en/fuses-for-multimeters/543-peaktech-7520-ceramic-fuse-ff-630ma-1000v-103x38mm-10-pieces.html"
I'll buy both 630ma and 10a (there's a limit of 75 €) and if anyone is interested, just let me know.
Thanks for your help.
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #26 on: February 25, 2025, 11:02:47 am »
I took some measures and now I'm quite sure it's a fake product (acording to Holly HC10aR datasheet). Moreover the crimping rings are missing on the fake fuse.
I've found the replacings, finally at
"https://peaktech-rce.com/en/fuses-for-multimeters/543-peaktech-7520-ceramic-fuse-ff-630ma-1000v-103x38mm-10-pieces.html"
I'll buy both 630ma and 10a (there's a limit of 75 €) and if anyone is interested, just let me know.
Thanks for your help.

You determined that the Fuses are faked just because the dimensions are a bit difference? It could be variation in production.
 

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #27 on: February 25, 2025, 12:06:30 pm »
Yes. It could be. And yet...The datasheet shows a lenght value = 38.0+/-0.6 mm. And then the measured resistance value.
Actually for those who own Hioki DT4282, maybe they can measure their own original fuses and post their findings here, for reference. It would be appreciated.
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #28 on: February 25, 2025, 01:53:02 pm »
Wow so someone faked a Chinese made product. So perhaps they made them in Vietnam? Also the resistance values are very close. I don't think they can be tighter tolerance than that. For the 630mA fuse the error was only -1%. The 11A fuse the error is less than -1% and Hioki stated that it can be 0.1 or less.
« Last Edit: February 25, 2025, 02:22:23 pm by BeBuLamar »
 

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #29 on: February 25, 2025, 03:24:16 pm »
The fuse of interest is the 630ma one. I don't know what the normal resistance value is since I don't have any other to compare with. The 11A one is the original.
So what you're telling is >1 ohm for a fuse (630ma in my case) is a normal value (knowing that the original 11A is 0.1 ohm). As for the dimensions, as far the factory is issuing  a datashett with accepted tolerances, yes I think the fuse in discussion is fake.
I don't like controversial disscusions. If you have smth of value to say and you're willing to help, please do so.
I'm simply cautious in using a questionable (IMO) product in my preciuous DMM.
 

Online Monkeh

  • Super Contributor
  • ***
  • Posts: 8221
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #30 on: February 25, 2025, 03:45:33 pm »
Yes. It could be. And yet...The datasheet shows a lenght value = 38.0+/-0.6 mm. And then the measured resistance value.

And which measurement falls outside the range of 37.4 to 38.6mm?

~1ohm is quite normal for a fuse in that range. For example, here's a datasheet for a slightly smaller fuse used in other meters, complete with nominal resistance: https://astml.com/wp-content/uploads/2023/12/HV620-Series_1000VDC_20231223.pdf
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #31 on: February 25, 2025, 03:57:41 pm »
The fuse of interest is the 630ma one. I don't know what the normal resistance value is since I don't have any other to compare with. The 11A one is the original.
So what you're telling is >1 ohm for a fuse (630ma in my case) is a normal value (knowing that the original 11A is 0.1 ohm). As for the dimensions, as far the factory is issuing  a datashett with accepted tolerances, yes I think the fuse in discussion is fake.
I don't like controversial disscusions. If you have smth of value to say and you're willing to help, please do so.
I'm simply cautious in using a questionable (IMO) product in my preciuous DMM.

The specs in the Hioki manual said the 630mA fuse should have 1.2Ohms resistance and the one you measured was 1.178 that's very close. And all the fuses dimension you measure is less than 38.6mm which is within specs.
 
The following users thanked this post: mwb1100

Offline trinitty

  • Newbie
  • Posts: 8
  • Country: fr
Re: Replace Multimeter Fuse
« Reply #32 on: February 25, 2025, 04:03:14 pm »
Yes. It could be. And yet...The datasheet shows a lenght value = 38.0+/-0.6 mm. And then the measured resistance value.

And which measurement falls outside the range of 37.4 to 38.6mm?

Yes. My mistake. I mistook the tolerance for a smaller value (0.06mm)
Thank you Monkeh. And thanks for the datasheet. Thank you all.
 

Offline Whales

  • Super Contributor
  • ***
  • Posts: 2236
  • Country: au
    • Halestrom
Re: Replace Multimeter Fuse
« Reply #33 on: February 25, 2025, 10:53:56 pm »
If you're certain it's a fake then crack it open and see if it has a sand-like media inside.  AFAIK all HRC fuses are filled, except fake ones, but I could be wrong.

Offline mwb1100

  • Frequent Contributor
  • **
  • Posts: 715
  • Country: us
Re: Replace Multimeter Fuse
« Reply #34 on: February 25, 2025, 11:30:11 pm »
I have seen many posts that claim that faked name brand multimeter fuses are rife on Amazon and eBay.  But I've never seen anything approaching evidence of that.  (note: I'm not talking about cheap no-name fuses which definitely have marginal to poor performance, as you'd expect).

At worst I'd expect the "fakes" to be off-the-books factory runs.  Maybe seconds/QA non-passes - but I'd be surprised.

Anyway, has anyone ever seen evidence?  Missing sand fill or a typo on the labeling?  Surely if Littelfuse multimeter fuses are being faked there would be some marked "LittleFuse"

« Last Edit: February 25, 2025, 11:35:11 pm by mwb1100 »
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 8868
  • Country: us
Re: Replace Multimeter Fuse
« Reply #35 on: February 25, 2025, 11:50:20 pm »
Anyway, has anyone ever seen evidence?  Missing sand fill or a typo on the labeling?  Surely if Littelfuse multimeter fuses are being faked there would be some marked "LittleFuse"

https://www.eevblog.com/forum/testgear/fake-bussmann-multimeter-fuses-on-ebay-dmm-b-44100-r-and-dmm-b-11a/

Sort of inconclusive, but even relabeled other-brand fuses (which to me does mean "fake") may not be that bad in quality and may function close enough to normally that few will notice.  You could just order BOJACK brand fuses though--no need for fake labels.

A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 
The following users thanked this post: mwb1100

Offline slugrustle

  • Frequent Contributor
  • **
  • Posts: 293
  • Country: us
Re: Replace Multimeter Fuse
« Reply #36 on: March 01, 2025, 05:45:09 pm »
The fuse of interest is the 630ma one. I don't know what the normal resistance value is since I don't have any other to compare with. The 11A one is the original.
So what you're telling is >1 ohm for a fuse (630ma in my case) is a normal value (knowing that the original 11A is 0.1 ohm). As for the dimensions, as far the factory is issuing  a datashett with accepted tolerances, yes I think the fuse in discussion is fake.
I don't like controversial disscusions. If you have smth of value to say and you're willing to help, please do so.
I'm simply cautious in using a questionable (IMO) product in my preciuous DMM.

I measured the original 630mA fuse in my DT4282 with a Rigol DM3068 bench DMM in 4-wire mode with two-pin Kelvin probes and got 1.089Ω.  The original 11A fuse came in at around 10mΩ.  The extra 0.1Ω in your readings is probably the resistance of your DMM test leads.  Try using rel mode to subtract the test lead resistance.

Unfortunately, the replacement 630mA fuses that I purchased on eBay have a resistance of about 334mΩ, which is not right.  They have red writing instead of blue and do not bear the UL recognized mark like the original fuses.  The 11A eBay fuses come in at 13mΩ, close to the originals, but no UL recognized mark again.  Maybe they don't bother with that mark when distributing to places other than the US...

I'm registering the meter with myHioki to see if maybe I can get fuses through Hioki directly.

I took some measures and now I'm quite sure it's a fake product (acording to Holly HC10aR datasheet). Moreover the crimping rings are missing on the fake fuse.
I've found the replacings, finally at
"https://peaktech-rce.com/en/fuses-for-multimeters/543-peaktech-7520-ceramic-fuse-ff-630ma-1000v-103x38mm-10-pieces.html"
I'll buy both 630ma and 10a (there's a limit of 75 €) and if anyone is interested, just let me know.
Thanks for your help.

The high current fuse has an 11A rating, not a 10A rating.
« Last Edit: March 01, 2025, 06:03:32 pm by slugrustle »
 

Offline cte

  • Regular Contributor
  • *
  • Posts: 77
  • Country: de
Re: Replace Multimeter Fuse
« Reply #37 on: March 03, 2025, 11:13:08 am »
I got this pack of 630 mA/1kV fuses from Distrelec, sold under the PeakTech brand, but apparently made by Holly.

[Update]: Measured resistance of 1.07 Ω and weight of one fuse is approx. 10.7 grams.
« Last Edit: March 03, 2025, 02:19:00 pm by cte »
⚡ To avoid electric shock, ensure that your instrument is correctly grounded.
 

Offline slugrustle

  • Frequent Contributor
  • **
  • Posts: 293
  • Country: us
Re: Replace Multimeter Fuse
« Reply #38 on: March 04, 2025, 01:20:14 am »
I filed a repair & calibration inquiry with Hioki using the "Contact Us" link on the webpage https://www.hioki.com/us-en/support/service.  I put N/A for company name.  Someone from ETI Energy Tools responded the next business day offering replacement fuses of both sizes for $18.70 each.

I should have tried the service request form earlier.
 

Offline watchmaker

  • Frequent Contributor
  • **
  • Posts: 819
  • Country: us
  • Self Study in EE
Re: Replace Multimeter Fuse
« Reply #39 on: March 04, 2025, 12:53:56 pm »
I looked at how to replace the fuse in the Fluke 45.  Looks like at least an hour to get the job done!

Do NOT blow that fuse.  OTOH, if I do blow it, maybe replace it with a 1 A slow blow?  :-DD
 

Online BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1591
  • Country: us
Re: Replace Multimeter Fuse
« Reply #40 on: March 04, 2025, 01:57:21 pm »
I looked at how to replace the fuse in the Fluke 45.  Looks like at least an hour to get the job done!

Do NOT blow that fuse.  OTOH, if I do blow it, maybe replace it with a 1 A slow blow?  :-DD

Shouldn't take an hour to replace it. Both the power supply fuse and the 100mA can be replaced without opening the case. The 10A fuse you have to open the case but it wouldn't take that long.
 

Online Monkeh

  • Super Contributor
  • ***
  • Posts: 8221
  • Country: gb
Re: Replace Multimeter Fuse
« Reply #41 on: March 04, 2025, 01:59:53 pm »
I filed a repair & calibration inquiry with Hioki using the "Contact Us" link on the webpage https://www.hioki.com/us-en/support/service.  I put N/A for company name.  Someone from ETI Energy Tools responded the next business day offering replacement fuses of both sizes for $18.70 each.

I should have tried the service request form earlier.

How much?

I know these are fairly specific fuses but they're not that expensive.
 

Offline watchmaker

  • Frequent Contributor
  • **
  • Posts: 819
  • Country: us
  • Self Study in EE
Re: Replace Multimeter Fuse
« Reply #42 on: March 04, 2025, 06:14:34 pm »
I looked at how to replace the fuse in the Fluke 45.  Looks like at least an hour to get the job done!

Do NOT blow that fuse.  OTOH, if I do blow it, maybe replace it with a 1 A slow blow?  :-DD

Shouldn't take an hour to replace it. Both the power supply fuse and the 100mA can be replaced without opening the case. The 10A fuse you have to open the case but it wouldn't take that long.

Talking about the DMM 4410  .4mA fuse.  That is buried deep.
 

Online bdunham7

  • Super Contributor
  • ***
  • Posts: 8868
  • Country: us
Re: Replace Multimeter Fuse
« Reply #43 on: March 04, 2025, 06:57:24 pm »
Talking about the DMM 4410  .4mA fuse.  That is buried deep.

If you do have the IEC 1010 version (later, not as common) then it's just three screws, slide the chassis out of the case and you're looking right at it.  If you have the earlier non-IEC 1010 version then there is no secondary fuse for the 100mA range.  The only fuse that takes a bit of work to replace would be the one on the battery board if you have that option.
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline watchmaker

  • Frequent Contributor
  • **
  • Posts: 819
  • Country: us
  • Self Study in EE
Re: Replace Multimeter Fuse
« Reply #44 on: March 04, 2025, 09:31:36 pm »
I see what happened.  The manual goes straight thru to complete disassembly after describing F2 and F5.  Since i was just looking to see what is involved, I read it quickly and did not notice the procedure you describe. :palm:
 

Offline slugrustle

  • Frequent Contributor
  • **
  • Posts: 293
  • Country: us
Re: Replace Multimeter Fuse
« Reply #45 on: March 05, 2025, 12:46:50 am »
The part numbers for the Hioki DT4282 fuses are:

630mA/1000V: 600391390
11A/1000V: 600391380

An internet search showed them for sale at the following distributor:
https://testeeurope.eu/fuse-630-ma-1000-v-c48976
https://testeeurope.eu/fuse-11-a-1000-v-c48975
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 39836
  • Country: au
    • EEVblog
Re: Replace Multimeter Fuse
« Reply #46 on: March 05, 2025, 05:39:06 am »
FYI, there is another fuse related video:

 