Whenever you're worried about scope ground loops a DMM is a good thing to drag out and check for any voltage disparities between the probe reference point on the DUT and the reference (gnd) clip itself.
Can you please elaborate?
Assuming I have a simple circuit consisting of one resistor and LED (2V @ 18mA) where
Vin = 5V
R = 166 Ohms
- Connect scope probe between Vin & R
- Not connecting the GND clip...
- Use DMM, switch to Voltage meter. Probe the Scope's probe and GND clip and see the reading
- ...?
If the supply is NOT mains ground referenced you can connect the probe reference most anywhere......within reason.
With experience a glance at a schematic can tell where is safe to reference from and this can be confirmed with the DMM.
Now remember this; the probe reference clip is where the waveform is referenced from BUT in most scopes the reference clip is
also at mains ground potential so great care must be taken on where to reference from so to not create mains ground loops.
Most everyone refers to the probe reference clip as the ground clip and while the scope has a mains ground connection this is correct; probe ground clip.
Now most everyone also makes a mistake at some time and connect the probe clip to a point where they shouldn't and if we're lucky only a blown fuse or tripped MCB or RCD and a few sparks to wake us up and serve as a warning to be more observant.
The potential disparity between the the DUT and probe reference clip (at mains ground potential) is what gets us into trouble so always know what it is !
DMM is your friend here.