Author Topic: Correct Impedance settings/circuit for Scope, Function Generator, and ADC setup  (Read 3487 times)

0 Members and 1 Guest are viewing this topic.

Offline nummyTopic starter

  • Newbie
  • Posts: 2
I have an Agilent (33220A) Function Generator set up to output a simple 60Hz sine wave.  I'm feeding this to an ADC pin on a microprocessor (TMS320F28069).  The output impedance is set to "High Z".  I also have a scope set to measure the signal being applied on the ADC pin.  The scope has an imput impedance setting of 1Mohm (you can choose 50ohm or 1Mohm).

According to the microprocessor documentation, this is how the ADC is setup internally.

Up until now, I have just been connecting the output of the function generator directly to the ADC pin.  Am I doing this wrong?

Should I be connecting the function generator to the ADC pin with a 50ohm resistor in series?  Should I have a capacitor somewhere?

Should I then change my setting on the function generator to 50ohm output impedance?  And then changing my scope to 50ohm input impedance?
« Last Edit: April 07, 2014, 01:37:33 pm by nummy »
 

Online Marco

  • Super Contributor
  • ***
  • Posts: 6939
  • Country: nl
The output impedance is set to "High Z".
No, it isn't ... the expected load impedance is set to High Z. The generator uses this to estimate what voltage the load will see (ie. it assumes it will see the full output voltage because the fixed 50 Ohm output resistance of the generator will take nearly no voltage). You should only use this at low frequencies or when using short cables. In this case (at 60 Hz with 1K input impedance for your ADC) it's fine.
Quote
The scope has an imput impedance setting of 1Mohm (you can choose 50ohm or 1Mohm).
This is fine.
Quote
Up until now, I have just been connecting the output of the function generator directly to the ADC pin.  Am I doing this wrong?
No, in this case it's fine ... the input impedance of your ADC is 1K Ohm and you're using low frequencies so it's relatively accurate. Just be aware that your ADC only sees 1000:1050 of the voltage you're setting on your generator.
Quote
Should I be connecting the function generator to the ADC pin with a 50ohm resistor in series?
No ... you could use a 50 Ohm parallel termination at higher frequencies though, you could put your scope on 50 Ohm input if you put a splitter straight on it's BNC input with a very short cable to the ADC, that way you use the scopes internal 50 Ohm termination  (not valid at very high frequencies, but with the generator only doing 20 MHz it's okay).

All that said, you should really read up on termination ... because all I said is probably voodoo to you without the necessary context.
« Last Edit: April 07, 2014, 01:52:02 pm by Marco »
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7549
  • Country: 00
  • +++ ATH1
Maybe this will help ?


Offline w2aew

  • Super Contributor
  • ***
  • Posts: 1780
  • Country: us
  • I usTa cuDnt speL enjinere, noW I aR wuN
    • My YouTube Channel

Should I be connecting the function generator to the ADC pin with a 50ohm resistor in series?  Should I have a capacitor somewhere?

Should I then change my setting on the function generator to 50ohm output impedance?  And then changing my scope to 50ohm input impedance?

Important note - the generator *always* has a 50 ohm resistor in series with the output!  Changing the *load* impedance setting on the generator only changes the way the generator calculates the voltage to be generated, it does NOT change the output impedance of the generator.
YouTube channel: https://www.youtube.com/w2aew
FAE for Tektronix
Technical Coordinator for the ARRL Northern NJ Section
 

Offline nummyTopic starter

  • Newbie
  • Posts: 2
Great tips, thanks guys! :-+
 