Author Topic: EEVBlog #1103 - Omicron Labs Bode 100 Review & Experiments  (Read 7236 times)
Online EEVblog
« on: July 09, 2018, 11:15:04 pm »
Review of the Omicron Labs Bode 100 50MHz Vector Network Analyser / Frequency Response Analyser and some experiments measuring bypass capacitors and characterizing a quartz crystal.
https://www.picotest.com/products_BODE100.html
Offline sibeen
Re: EEVBlog #1103 - Omicron Labs Bode 100 Review & Experiments
« Reply #1 on: July 10, 2018, 06:19:07 am »
Omicron just makes nice gear. I'm quite familiar with their protection relay testgear. That comes in at least 10x the cost of the Bode but for what it does it is worth it.
Offline PA4TIM
« Reply #2 on: July 10, 2018, 09:14:12 am »
Nice to see more network analysis stuff. This is a nice peace of kit and seeing you get usable results without ( i think) working through the manual and many, many hours practice, the software must be user friendly.
VNA's are great for impedance around 50 ohm. For very high or very low impedance you use IV measurements (impedance meters/analysers use that) and for some VNA's (like the SDR-kits VNWA) there are adapters to do that. The software needs to support that (like the VNWA does)  if you do not want to do the math. I did that once just for fun with a 1968 build HP VNA (the ESR of a 1nF cap) See the attached photo. But there are also tricks like measuring a 100 K resistor parallel to a 50 ohm or a 0,001 ohm in series with a 50 ohm. The 50 ohm can be your cal load or just measure it, a 50 ohm resistor can be measured extremely well with a good calibrated VNA. In both cases I use the custom trace function so I do not need top do the math.
I would expect the Monicron supports IV measurements in the bodeplot function and "normal" wave parameters for the VNA part ?
Dave, the results of a vna are at the best as good as the calibration.
This is often a rather time consuming thing and needs a very good calkit (a male and a female based kit and then the things you used fort that fixtures) and very advanced software implementation to do n-term error corrections. That is why VNA's have very extensive calibration functions and things like de-embeding and port extension functions. Even a hobby VNA as the SDR-kits VNWA can do that. I missed that in your review.
Are the fixtures and their cal-kit parts already characterized in the software ? I would expect a cal certificate with the numbers
The resistor value does not have to be a precision one, as long as you know the resistance and are able to tell that to the VNA. Things like open and short are more important (OK, for 0-50 MHz probably everything will do  ;) ) I use my SDR-Kits VNWA often in the span 1 KHz tot 10 MHz and calibration there is not really a big thing. Above 500 MHz it becomes the most important part of the measurement.
I like the fact it can make bode plots and do s-parameters, that combination is very nice.
The DUT fixtures look cheap if you compare them with the keysight ones. But I do not know the prices from those, probably much higher.
http://literature.cdn.keysight.com/litweb/pdf/5968-5329E.pdf this pdf tell a lot about the things to know when using fixtures.
When I started playing around with VNA's there was almost no information for non RF graduated EE's That is why I made a series of tutorials. They are not flawless but I tried to make it easy to understand and for most practical. I used the SDR-kits VNWA for it, back then more or less the only choice.
For those who are new in this field, check them out, they are free pdf's. http://www.pa4tim.nl/?p=1594 All in English and some are translated in French and German by other VNWA users.
Keysight has a free PDF, the impedance measurement handbook. https://literature.cdn.keysight.com/litweb/pdf/5950-3000.pdf  very interesting and "easy" to understand
Offline timgiles
  • Programmer, DB architect
« Reply #3 on: July 10, 2018, 12:02:17 pm »
Thanks Dave - great first video - looking forward to the others. Flexibility looks off the scale.
Offline Smokey
Offline Wolfgang
    • Electronic Projects for Fun
« Reply #5 on: July 12, 2018, 04:19:31 pm »
I would say that the creation of cal kits is a very tricky task in the GHz region, but not really at 50MHz where the Bode 100 is operating. I have made a lot of cal kits for ham radio use, and I calibrated them by comparison with a Keysight E5071C using a Keysight eCal module. Below 100MHz 30dB or better can be reached even with homebrew open, shorts and loads. Even leaded components could be used.
The accessories that come with the Bode 100 are all made using very classic construction techniques and no Gigahertz trickery. Everything more seems to be luxury.
Offline Bud
« Reply #6 on: July 12, 2018, 04:43:53 pm »
The isolation transformer seems to use a regular Hammond type aluminum cast enclosure. Perhaps $8 a piece or something. The main unit seems to use a standard enclosure as well. Nothing against it in terms of performance but I'd think for the selling price something custom could be made to stand out better.
Offline jancumps
  • New Low
« Reply #7 on: July 12, 2018, 05:17:01 pm »
Do it :)
Offline envisionelec
« Reply #8 on: July 12, 2018, 05:41:14 pm »
As I mentioned in another, similarly named thread, I broke down the entire BOM and came up with about $110 for a single unit. I'm building one and will characterize it - then probably start selling them because, why not?
Offline Wolfgang
« Reply #9 on: July 12, 2018, 05:45:14 pm »
Much luck ! IMHO you will be able to reproduce the hardware, but all the software and GUI is another story. This could well be  a few mythical manmonths minimum ... >:D
The following users thanked this post: jancumps
Offline jancumps
  • New Low
« Reply #10 on: July 12, 2018, 05:54:04 pm »
This. And the promise that you will be available when something is wrong. And the design work that you do not have to earn back. And the samples that you don't have to send to Australala...
Offline Brumby
« Reply #11 on: July 13, 2018, 03:28:52 am »
... that you don't have to send to Australala...
La la land is on the west coast of the USA.   ;)
Online EEVblog
« Reply #12 on: July 13, 2018, 06:48:35 am »
Doh!
I complete missed the AD834 multiplier in the receiver can.
The part number wasn't really visible direct on
Offline Kleinstein
« Reply #13 on: July 13, 2018, 07:03:58 am »
Using that analog multiplier is kind of a surprise: Its an expensive part and helps to reduce spurious signals due to harmonics of the LO. However its also limited accuracy and stability. So the 24 Bit ADC used later on does not make that much sense anymore. At least the analog multiplier (likely as a mixer) reduces the need for filtering.  So maybe they use some of the CMOS switches for the possibility to do extensive internal adjustment runs to compensate for the limitation of the analog multiplier. It does not even look like they could bypass the AD834 for the low frequencies, where there is no absolute need for a mixer.
Online EEVblog
Offline Kleinstein
« Reply #15 on: July 13, 2018, 07:20:09 am »
Ups - I think the last few posts should have gone to the tear-down thread and not here.
Offline JS
  • Country: ar
« Reply #16 on: July 13, 2018, 08:46:20 am »
Looks like so, I ended up here as someone mentioned in the other post...
JS
If I don't know how it works, I prefer not to turn it on.
Offline Berni
  • Country: si
« Reply #17 on: July 13, 2018, 09:12:23 am »
Sweet bit od kit. The price point might be a bit steep, but dang does that software look good. It does exatly what its suposed to without exesive messing about, setting stuff, reading manuals etc. If only otger test gear came with software that us even half as good as this.
Seing a mutiplier makes perfect sense to down convert to an IF that the ADC can sample, and we can make very very nice ADCs for 100KHz due to audio. Mixer harmonics can be compensated for with calibration and the 24bit ADC gives you exelent dynamic range(they are not too expensive ether)
I do have a Agilent Vector Signal Analyzer that does DC to 10MHz with similar capability(signal gen, dual input) but the software is simply not there to do much more than gain/phase and spectrum analysis(tho has exelent dynamic range). It could do all this can do too with some extra test fixtures, but i neved had a big enugh need to make such software using GPIB, nor would it be anywhere near as nice as this.
Offline Alextsu
  • Country: ru
« Reply #18 on: July 14, 2018, 07:15:12 am »
Thanks for this post, Dave!
If anyone would try to re-eng the input receiver channel schematics, or have some ideas how to build one, please, share.
Couldn't afford such an expensive FRA, but would like to try to make something similar for homelab.
« Last Edit: July 15, 2018, 07:44:18 am by Alextsu »
Online EEVblog
« Reply #19 on: July 14, 2018, 07:49:40 am »
+1
Offline Kleinstein
« Reply #20 on: July 14, 2018, 11:55:43 am »
Could not afford such an expensive FRAs, but would like to try to make something similar for homelab.
I am also looking for Venable Industries FRA 32xx series analyzers older versions of software (vers. 4 for WinXp).
The first half of the receiver part should be similar to an scope input stage, and a few extra switches to choose between the input and an internal cal signal / the source.  For the rather low frequency range a simple mixer, some AA filtering (not that much with an SD-ADC) and that the good resolution ADC could be all it takes. A difficult part would be to get a well defined frequency response for all gain settings - because of this I would prefer to have some of the gain switching behind the mixer. 50 MHz is not that high in frequency, but a IF in the 10 kHz range is even easier to handle. Both Channels seem to share a single LO that is likes driven from the same master clock. For a low cost version I would imagine one could get away with a switching mixer, maybe CMOS switches.
There are a few DIY VNA projects around, that also use DDS chips and high resolution ADCs - as the software is about the largest part of the project, it might be easier to start with such a circuit so that maybe one could have a start on the software. It may take changes with the input stages to go all the way to the DC coupled range.
The other alternative would be something like the red pitayar or analog discovery HW, using the faster ADCs and direct sampling. This common HW might help to find some open source SW, at least for a start.
Offline voltsandjolts
« Reply #21 on: July 14, 2018, 12:58:44 pm »
I realise the Analog Discovery has nowhere near the breadth of capability the Omicron has but it would be interesting to see a few comparisons.
Does the Analog Discovery FRA give comparable results, at least within it's comfort zone?
Offline Bud
« Reply #22 on: July 15, 2018, 01:58:16 am »
It is a delta sigma ADC, this type of ADC typically have hardware very narrow band selectable digital filter, which greatly reduces requirements for external analog filters. This comes at the expense of increased conversion time though.
Offline Bud
« Reply #23 on: July 15, 2018, 02:00:47 am »
Do it :)
Sorry do what?
Offline Bud
« Reply #24 on: July 15, 2018, 02:02:44 am »
What are your guesses why MELF resistors were used?