It cant be that T184-2 core, the AL is way too low to get the specified inductance. My bet goes out to a Vacuumschmelze W424, the colour matches, the size matches, as do the specs if you run the numbers. I knew I had one of those laying around somewhere, so I dug it up and whipped up another test. This time 50 turns (I had some wire leftover after 40
![Wink ;)](https://www.eevblog.com/forum/Smileys/default/xwink.gif.pagespeed.ic.cldandycH0.png)
) on the W424 core, the picture speaks for itself.
Once again I measured with the analog discovery. The "Cal" image shows the wavegen output connected to both CH1 and CH2.
CH2 is displayed in blue and is in relative mode, so gain and phase relative to channel one. This removes any gain variations due to wavegen output loading etc (its a 50 ohm output).
"Abs" shows the measurement as Omicron does it, I think, Here the amplitude is absolute compared to the wavegen (and thus -6db because of the 50Ohm termination), and the phase is relative to CH1.
"Rel" is the same measurement, but now in relative mode, this shows the transfer function of just the transformer.
"100mV" is the previous absolute measurement but now at 100mV instead of 1V input amplitude, this lessens the loading on the wavegen output stage.
Pri -sec capacitance is 130pF
Magnetizing inductance is 274mH
And leakage inductance is 3uH