Author Topic: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE  (Read 2084 times)

0 Members and 1 Guest are viewing this topic.

Offline EEVblog

  • Administrator
  • *****
  • Posts: 34400
  • Country: au
    • EEVblog
EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« on: January 23, 2022, 10:28:50 pm »
A $99 Fluke PM3370B Combiscope score from Ebay!
Will it work?

 

Offline BreakingOhmsLaw

  • Frequent Contributor
  • **
  • Posts: 261
  • Country: de
  • Certified solder fume addict
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #1 on: January 25, 2022, 10:02:34 am »
I used to work for Philips back in the 90s in a factory for CRTs.
We made absolutely *everything* when it came to parts. Open any parts drawer in the workshop, Philips, Philips, Philips, Philips ,Philips. You could source practically everything you wanted from within the company. Good times.
All the Test Equipment: Philips, Philips, Philips, Philips ,Philips. If any test equipment would fail, you'd go to an office where a guy would pull out a microfiche of the circuit diagram and service manual of any Philips device in existence. Pure madness to keep it up-to-date, but hey, this was before the internets. You could then take it and view it on these ridiculously large microfiche viewers. 
Philips was an absolute behemoth. Then they started selling everything and it now is only shadow of itself. Fond memories.

By the way, you are lucky: The analog counterpart - PM3070 - still uses the old power supply board with RIFA timebombs. I have one here in my lab that I got to take home when the factory was dismantled. It's cost a fortune back then. Still use it now and then, the trigger section in these units is absolutely brilliant. You can tell that these were designed by engineers that used them themselves.
 

Offline Teo

  • Newbie
  • Posts: 2
  • Country: it
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #2 on: January 25, 2022, 10:57:43 am »
Hi Dave! I got a fully functional PM3380A as gift some time ago and the seller was so kind to also send in all of its resources, so I have every kind of manual (compatible with your 3370B  :-+), including the service manual and even several programs  that unfortunately I cannot run on win10, but whatever...If you want I can send you these, just let me know how to send you all that stuff 'cause it's quite heavy in terms of memory (the sevice manual it's 50MB alone)
 

Offline Shock

  • Super Contributor
  • ***
  • Posts: 3961
  • Country: au
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #3 on: January 25, 2022, 11:34:08 am »
Welcome to the forums. Upload manuals and software to these websites, just check first it's not already uploaded and compress the software in a rar or zip and make it clear with an additional uploaded .txt file what the software is for and does. Same for firmware images etc (include the versions and models). Alternatively throw it in an open access google share and link to it.

Most companies don't have a problem with it being uploaded, I only recall Apple and Toshiba whinging about distribution of laptop service manual and training material. Fluke/Phillips would probably encourage you to upload it.

http://www.ko4bb.com/getsimple/index.php?id=manuals
https://xdevs.com/contact/
Soldering/Rework: Pace ADS200, Pace MBT350
Multimeters: Fluke 189, 87V, 117, 112   >>> WANTED STUFF <<<
Oszilloskopen: Lecroy 9314, Phillips PM3065, Tektronix 2215a, 314
 
The following users thanked this post: Teo

Offline Teo

  • Newbie
  • Posts: 2
  • Country: it
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #4 on: January 25, 2022, 01:31:49 pm »
Thank you! I'll post the link down below ASAP, it will take some time though. Thanks for your suggestion  :-+
 

Offline free_electron

  • Super Contributor
  • ***
  • Posts: 8299
  • Country: us
    • SiliconValleyGarage
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #5 on: January 25, 2022, 05:52:04 pm »
Philips had a penchant to make everything in their test equipment custom. all those oQ parts are unobtainium. They rarely break though.

Many of these scopes are plagued by the EHT cascade (high voltage generator for the tube) flashing over. ( which is most likely what is happening here. the cascade arcs , causing the power supply to go in overload and shutting down. )
My first 'digital' scope was a philips PM3050 with the LCD display next to the picture tube. The entire control is ran over an i2c bus. every single chip was wired into that and the entire scope was run from an 8031 microcontroller. it had 2k of acquisition memory per channel.

The nice thing about these machines is that they are true analog scopes. the 'digital ' portion is simply a recorder in front of the analog shebang. it digitizes the signal and keeps playing it back. by turning off the block you have the analog machine.
They are also dual beam scopes. in analog mode the traces have individual beams. once you switch to digital one beam is used for waveform playback , the other for text and menus ( hence the individual intensity control for text and trace. )

I have fond memories of that machine. traded it for a hp54645d , never looked back.
Professional Electron Wrangler.
Any comments, or points of view expressed, are my own and not endorsed , induced or compensated by my employer(s).
 
The following users thanked this post: EEVblog

Offline EEVblog

  • Administrator
  • *****
  • Posts: 34400
  • Country: au
    • EEVblog
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #6 on: January 27, 2022, 04:44:34 am »
Hi Dave! I got a fully functional PM3380A as gift some time ago and the seller was so kind to also send in all of its resources, so I have every kind of manual (compatible with your 3370B  :-+), including the service manual and even several programs  that unfortunately I cannot run on win10, but whatever...If you want I can send you these, just let me know how to send you all that stuff 'cause it's quite heavy in terms of memory (the sevice manual it's 50MB alone)

Thanks. I've got the service manaul PDF already.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 34400
  • Country: au
    • EEVblog
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #7 on: January 27, 2022, 04:46:04 am »
Philips had a penchant to make everything in their test equipment custom. all those oQ parts are unobtainium. They rarely break though.
Many of these scopes are plagued by the EHT cascade (high voltage generator for the tube) flashing over. ( which is most likely what is happening here. the cascade arcs , causing the power supply to go in overload and shutting down. )

Someone else mentioned this as well, supposedly RIFA caps inside?. If it is the trippler then I don't like my chances of getting a direct replacement  :(
 

Offline Shock

  • Super Contributor
  • ***
  • Posts: 3961
  • Country: au
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #8 on: January 27, 2022, 06:12:51 am »
Someone else mentioned this as well, supposedly RIFA caps inside?. If it is the trippler then I don't like my chances of getting a direct replacement  :(


There is a bunch of parts units out there, you just need to put the word out. I have two PM3065 and a parts PM3350 playing a crucial life support role, so unfortunately no donors. But here is a sexy PM3055 pic instead.

Soldering/Rework: Pace ADS200, Pace MBT350
Multimeters: Fluke 189, 87V, 117, 112   >>> WANTED STUFF <<<
Oszilloskopen: Lecroy 9314, Phillips PM3065, Tektronix 2215a, 314
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1209
  • Country: au
  • Watts in an ohm?
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #9 on: January 27, 2022, 09:47:56 am »
Philips had a penchant to make everything in their test equipment custom. all those oQ parts are unobtainium. They rarely break though.
Many of these scopes are plagued by the EHT cascade (high voltage generator for the tube) flashing over. ( which is most likely what is happening here. the cascade arcs , causing the power supply to go in overload and shutting down. )

Someone else mentioned this as well, supposedly RIFA caps inside?. If it is the trippler then I don't like my chances of getting a direct replacement  :(
I my early days of learning  to fix electronic  gadgets  TV'S from curbside  cleanups were a popular  target, and in particular  Philips was a favourite.  So the mention  of tripplers tripped a memory  cell that those were a common  failure  point in those sets and spares were available  from Wagner  electronics in Ashfield
 A quick look at the  combiscope manual  gives a part BG2000 for the  trippler which also turns up in an old Wagner  pdf search.  No idea what if any stock of old TV parts they still have  but I suspect  this  part may be easier to  get than it first  seems.
 One other possible  thought  is the  power supply  has an overtemp thermistor to monitor  the PS temp and will shut down  the  supply  if detected,  perhaps  that  is the  suspect  cct and not the  trippler!.
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 
The following users thanked this post: EEVblog

Offline EEVblog

  • Administrator
  • *****
  • Posts: 34400
  • Country: au
    • EEVblog
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #10 on: January 28, 2022, 10:04:29 am »
Someone else mentioned this as well, supposedly RIFA caps inside?. If it is the trippler then I don't like my chances of getting a direct replacement  :(


There is a bunch of parts units out there, you just need to put the word out. I have two PM3065 and a parts PM3350 playing a crucial life support role, so unfortunately no donors. But here is a sexy PM3055 pic instead.



I lusted after that scope when I was a kid!
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5011
  • Country: de
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #11 on: January 28, 2022, 11:07:46 am »
Even today these scopes are wonderful and have a place in the lab.
I still have about 20 of these scopes around.

If you need the original Software to control the scope from a PC over RS232 let me know.
You can even take screen shots over the software, which was very special in those days.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline Shock

  • Super Contributor
  • ***
  • Posts: 3961
  • Country: au
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #12 on: January 28, 2022, 12:44:22 pm »
I lusted after that scope when I was a kid!

I know, you told me 9 years ago hehe. I used to lust after Dick Smiths Multiband Receiver due to aesthetics (until I looked inside and saw chinesium). That Philips model is classy though, you could sell a clean design like that (but cut down, with an led screen) even today.
Soldering/Rework: Pace ADS200, Pace MBT350
Multimeters: Fluke 189, 87V, 117, 112   >>> WANTED STUFF <<<
Oszilloskopen: Lecroy 9314, Phillips PM3065, Tektronix 2215a, 314
 

Offline Old Fart Analog Engineer

  • Contributor
  • Posts: 12
  • Country: us
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #13 on: February 02, 2022, 07:33:46 pm »
REPAIR PM3370B:
Its the Over Voltage / Over Current protection SCR circuit in the "mains" power supply. There is a trimpot to adjust it which works as a voltage divider to set the trip point. It is likely corroded and increases or opens which triggers the SCR to to latch and shut it all down. Clean that with some contact cleaning spray, maybe also "rock" it back and forth a little bit. It is likely to have silver plated contacts, which are tarnished. If it is a "Piher" brand (used throughout European electronics) it is junk. Philips also made some which were marginally better. Also the link to the Forum is wrong.
« Last Edit: February 02, 2022, 07:36:38 pm by Old Fart Analog Engineer »
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 34400
  • Country: au
    • EEVblog
Re: EEVblog 1450- $99 Fluke Combiscope Ebay SCORE
« Reply #14 on: February 03, 2022, 04:31:18 am »
REPAIR PM3370B:
Its the Over Voltage / Over Current protection SCR circuit in the "mains" power supply. There is a trimpot to adjust it which works as a voltage divider to set the trip point.

There is no trimpot on the board.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf