EEVblog => EEVblog Specific => Topic started by: gordo51 on October 23, 2011, 06:44:02 pm

Title: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: gordo51 on October 23, 2011, 06:44:02 pm
Hey what do you know, I have the same unit on my bench except according to the date stamp on the LM741 it was built in 1974 !
Pretty much exactly the same layout of components except I have a Teledyne Philbrick 1703 chopper stabilized op amp module in place of the plug-in board.  Beside the reference zener I have a small paper label stuck on with hand written voltage and current values.  Thanks for showing the schematic.  I was able to adjust pots R1, R3 and R4 so now I have 3 zeros after the decimal point on the 10V range and the other two ranges are spot on.  Of course this is according to my uncalibrated HP3456A voltmeter.  As they say "other voltmeters may give you different mileage..."
Keep up the great work.
Gordo51

PS: I didn't realize it at first but my unit has an unlabelled knob to the left of the range switch that works a potentiometer. I didn't think to check out its effect when I was making my adjustments.  You wouldn't think this would affect the calibration settings?  Wonder what it is for?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: Nermash on October 23, 2011, 08:14:55 pm
It is a nice catch for 20ish bucks, if you are lucky enough:)

BTW, what do they use to calibrate that 3458A?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: SgtRock on October 23, 2011, 09:10:15 pm
Dear Gordo51:

--Why is it I cannot find EEVBLOG #210. Do you need a DJ decoder ring, or what?

"The biggest competitive advantage is to do the right thing at the worst time." Bill & Dave

Best Regards
Clear Ether
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: McMonster on October 23, 2011, 09:13:28 pm
You need to look on EEVblog's Youtube channel, there's usually a delay between uploading a new video and posting it on the site.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: SgtRock on October 23, 2011, 09:20:37 pm
Dear McMonster:

--Good to know. Thanks.

"The telegraph is a kind of a very long cat. You pull his tail in New York and his head is meowing in Los Angeles. Radio operates the same way, except there is no cat." Albert Einstein

Best Regards
Clear Ether
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: Kiriakos-GR on October 25, 2011, 01:43:31 pm
My own voltage - current source, looks to use similar technology.

But those English guys had cover with black hard glue the PCB with the IC and the diode, and so I did not take any pictures.
I did open it up yesterday, so to check it out.

And I managed to tweak it, by adding on the main potentiometer , my decade resistor box in parallel,
and I created one high quality resistor divider, that helps out to set voltages with so great detail speaking about bringing to zero all the digits,
at any voltage between  mv to 8V.

https://www.eevblog.com/forum/index.php?topic=1032.0 (https://www.eevblog.com/forum/index.php?topic=1032.0)

I will add pictures about it soon, in the main thread about it.  :)


 
Title: EEVBLOG #210 - Mystery Device
Post by: Kozmyk on October 28, 2011, 01:59:28 am
What about the Mystery Device?
Did you get any closer to identifying it?
Title: Re: EEVBLOG #210 - Mystery Device
Post by: EEVblog on October 28, 2011, 06:21:01 am
What about the Mystery Device?
Did you get any closer to identifying it?

Haven't really looked into it, sorry.

Dave.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: amspire on October 28, 2011, 06:35:45 am
I think the mystery device is a precision 3 ohm resistor. 4 terminal so two terminals can be used for the series current through the resistor, and two outputs to provide the precise 3 ohm output without any lead resistance error.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: Kozmyk on October 28, 2011, 01:17:57 pm
Well ferreted that man. 8^)
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on November 20, 2011, 01:20:08 pm
Nice review!  I didn't know about this vblog until today.  I was bidding on the same model as EDC, last month but I was away on vacation.  Perchance were be bidding on the same device, Dave?  Did you get yours from the USA?

Anyway, I love how its made and Dave's video was immensely helpful.  I didn't know it was so simple inside, its very KISS so its likely to last forever.  Its tolerances are huge, from the casing to the PCB: large and crude but solid as a rock.  Those old IC have such clearly made labels and nice trimming, compare to any LM741 today.   Comparing it the 3458a is the ultimate, so its just fantastic this simple device is so accurate and precise.

I managed to snag one of those today so I'll post how I fair with with.

@Gordo51, I have the same set up as yours for testing the MV106.  But I maintain 4 3456a to cross reference each other for accuracy, its a volt-nut fever I picked up last year. 

Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on November 20, 2011, 01:27:29 pm
Calibrating the 3458a is very difficult, its the ultimate DMM.  Only a Josephson Junction does it justice, so comparing it to a metcal labs 3458a is a very authoritative measurement, assuming their 3458a is properly calibrated & maintained, which on the video appears to be.

Calibrating the 3458a is one reason I don't own one, even if you could find an eBay version of it and patch it up to factory condition, one can't get its famed accuracy without shelling the $1200 to get it calibrated, annually.  Thereafter, it has to be auto-cal everyday and kept in a temp and preferably humidity controlled room to insure its 0.05 ppm dcV transfer accuracy. and 4-8ppm yearly accuracy is maintained.



It is a nice catch for 20ish bucks, if you are lucky enough:)

BTW, what do they use to calibrate that 3458A?
Title: Voltage Reference Standard was Re: Krohn-Hite Voltage Standard
Post by: saturation on December 02, 2011, 07:26:02 pm
I've just had a short exchange with Joe Inglis at Krohn-Hite about the MV106, and thank them for their assistance!  Anyone looking to service any EDC or KH device would do well to contact Joe.

I've been performance testing this device and its stability is very impressive. 

After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 0.0005 x 10-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.

This is far better than the spec sheet 1-8 h stability of 5ppm-10ppm + 2uV.   

But based on the above data it could be possible to use the calibrator to its lowest resolution of 10nV.

If you can find this model or its variants on eBay it well worth it to own if you need a voltage reference; components are easily obtained if repair were needed, the device is very easy to service and inspect, its a tribute to the EE who designed it that such stability and accuracy can be made this simply. 

Its a lot more accurate and stable than Geller or voltagestandard.com offerings [ which can still be used as transfer references via snail mail] and a lot easier to maintain than a Fluke 732A series if < 30ppm is adequate.  The manual states its operating environment is 20oC-30oC so it was designed to be a field instrument, chances are in a more temperature controlled setting it will be far stabler over time.


Title: Re: Voltage Reference Standard was Re: Krohn-Hite Voltage Standard
Post by: EEVblog on December 02, 2011, 10:43:37 pm
I've just had a short exchange with Joe Inglis at Krohn-Hite about the MV106, and thank them for their assistance!  Anyone looking to service any EDC or KH device would do well to contact Joe.

Yes indeed, Joe was very helpful in getting me the schematic.

Quote
I've been performance testing this device and its stability is very impressive. 

After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.

Wow, nice to know, thanks.
I presume that others would have similar order stability as well?

Quote
If you can find this model or its variants on eBay it well worth it to own if you need a voltage reference; components are easily obtained if repair were needed, the device is very easy to service and inspect, its a tribute to the EE who designed it that such stability and accuracy can be made this simply. 

I concur, an amazing device it seems, belying it's design simplicity.

Dave.
Title: Re: Voltage Reference Standard was Re: Krohn-Hite Voltage Standard
Post by: alm on December 02, 2011, 10:55:14 pm
After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.
How much does the 3456A contribute to this? In the past, I've measured short term variance (n=1000) to be about 5 * 10-15 V2 on the lowest DCV range with 100 NPLC and shorted inputs (three orders of magnitude less than your measurement), but I'm not sure how much sample variation there is between multimeters. I would expect long term variance to be larger, since I'd expect there to be some slight drift in the gain, eg. due to tempco.
Title: Re: Voltage Reference Standard was Re: Krohn-Hite Voltage Standard
Post by: amspire on December 02, 2011, 11:29:03 pm

After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.

Is that meant to be 0.005 x 10-3?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 02, 2011, 11:33:37 pm
A variance of 0.005 * 10-9 corresponds to a standard deviation of about 2ppm, not unlikely given a stable source.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: amspire on December 02, 2011, 11:44:39 pm
Ah! Variance is a squared function.

I knew that. Once upon a time.  :-[
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 03, 2011, 12:07:54 am
I'm guessing that saturation quoted variance because the HP 3456A can only calculate variance, not standard deviation. A seemingly strange omission. Maybe they didn't have the spare CPU cycles to calculate the square root so many times per second? It's not like today when you might stick an embedded ARM system in it which is idle 90% of the time.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 03, 2011, 03:55:47 pm
Thanks alm.  I made an order of magnitude error: I missed a zero, the variance on the measurement is 0.0005 x 10-9 so its more like 1ppm at this point.  Apologies to all for the missing zero; have made corrections.

For comparison, the $1000 eBay Fluke 732A voltage reference has a 10V spec'd stability of 0.3ppm at 30 days and 1ppm at 1 year

http://www.ebay.com/sch/i.html?_trkparms=65%253A12%257C66%253A4%257C39%253A1%257C72%253A5841&rt=nc&_nkw=Fluke+732A&_sticky=1&_trksid=p3286.c0.m14&_sop=15&_sc=1 (http://www.ebay.com/sch/i.html?_trkparms=65%253A12%257C66%253A4%257C39%253A1%257C72%253A5841&rt=nc&_nkw=Fluke+732A&_sticky=1&_trksid=p3286.c0.m14&_sop=15&_sc=1)

http://us.flukecal.com/products/electrical-calibration/electrical-standards/734a-reference-transfer-standards?quicktabs_product_details=2 (http://us.flukecal.com/products/electrical-calibration/electrical-standards/734a-reference-transfer-standards?quicktabs_product_details=2)

The MV 106 manual specs its best accuracy at 5ppm at 1 hour but rated at wide operating temperatures, so the accuracy could be far better if kept in a stabler test temperature.  Note, the Fluke standard is likewise rated for a wide operating temperature.

http://www.krohn-hite.com/htm/ServiceSupport/PDF/Manuals/MV106J-116J%20Manual.pdf (http://www.krohn-hite.com/htm/ServiceSupport/PDF/Manuals/MV106J-116J%20Manual.pdf)

As for the contribution of the 3456a, its a probabilistic guess, I haven't done the output short on the 3456a for a while to see how my 3456a drifts, but with your data as comparison, I'll let you know in 24h.  So far the measured specs are in line with your measurement [ the actual 3456a contribution is negligible] as the measured 1ppm variation at 1 week is below the 1 day spec best case stability for either the 3456a at 8ppm or the MV106 at 5ppm [not counting the LSD drifts and uV offset on the MV106].

After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.
How much does the 3456A contribute to this? In the past, I've measured short term variance (n=1000) to be about 5 * 10-15 V2 on the lowest DCV range with 100 NPLC and shorted inputs (three orders of magnitude less than your measurement), but I'm not sure how much sample variation there is between multimeters. I would expect long term variance to be larger, since I'd expect there to be some slight drift in the gain, eg. due to tempco.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 03, 2011, 04:52:34 pm
For a company its size and age, Krohn-Hite has survived much upheaval in the industry to suggest their portfolio of gear is well respected. 

http://www.manta.com/c/mm71qq0/krohn-hite-corp (http://www.manta.com/c/mm71qq0/krohn-hite-corp)

The manuals I received from KH suggests the guts of the MV106 is a module for a series of instruments: variants exist with a voltmeter, strip chart recorder outputs, mA calibrator, 1V in lieu of 100mV, nanvolt variant with a 1mV scale instead of 10V ... again a testament to their ingenuity of taking a basic design and likely re-purposed to other arenas.  So, its very likely the accuracy is similar if not better in other variants.

100mV variant
(http://i.ebayimg.com/t/EDC-MODEL-MV106-DC-VOLTAGE-STANDARD-TESTED-GOOD-/05!B+ezlHgB2k~$%28KGrHqV,!k8Ey+jC2SdvBM+wjCVsgg~~_3.JPG)

10mV variant
(http://i.ebayimg.com/t/EDC-MV106-Low-Impedance-DC-Nano-Volt-Source-MV-106-G-/00/$%28KGrHqJ,!iwE3QOY,jB7BN1Ujh8ZSw~~0_3.JPG)

1V variant
(http://i.ebayimg.com/t/EDC-MODEL-MV116-DC-VOLTAGE-STANDARD-TESTED-GOOD-/00/s/NDUxWDEwMjQ=/$%28KGrHqF,!ncE63T09SI3BO1vEyPzY!~~60_3.JPG)

Amp calibrator variant:
(http://i.ebayimg.com/t/EDC-Electronic-Development-Corporation-CR103-voltage-and-current-calibrator-/02/!B%28vNIuQBGk~$%28KGrHgoH-C8EjlLlh!lGBKd0zGFDFw~~_12.JPG)


etc.,

There are hardly any reviews of EDC products or discussions of them [ at least via google].  So, very likely this thread will pop up in the near future when EDC or Krohn Hite are searched for.  This device and its derivatives are worthy of more interest from anyone looking a well made, portable affordable voltage standards that can be priced right for small labs.



I've just had a short exchange with Joe Inglis at Krohn-Hite about the MV106, and thank them for their assistance!  Anyone looking to service any EDC or KH device would do well to contact Joe.

Yes indeed, Joe was very helpful in getting me the schematic.

Quote
I've been performance testing this device and its stability is very impressive. 

After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.

Wow, nice to know, thanks.
I presume that others would have similar order stability as well?

Quote
If you can find this model or its variants on eBay it well worth it to own if you need a voltage reference; components are easily obtained if repair were needed, the device is very easy to service and inspect, its a tribute to the EE who designed it that such stability and accuracy can be made this simply. 

I concur, an amazing device it seems, belying it's design simplicity.

Dave.
Title: Power Consumption Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 04, 2011, 12:58:42 pm
At this post, with n ~> 12000, the variance of a shorted 3456a input is 2.9 x 10-14 using PLC 100. [ I have 3, the other two are doing long term measurements of the MV106A and Geller reference.]

It contributes about 0.17 ppm to the variation of the measurement.
 
As aside, I measured the power consumption of the MV106 at 5W using a killawatt.


After 24/7 days of monitoring fluctuations of 1.000 000 VDC at 21oC,  my 3456a is registering variance of .005 x 10^-9 using a 3456a.  This device performs similar to what Dave showed in the video but I've stretched it to 168+ hours and had the 3456a analyze variations continuously using its statistical functions.
How much does the 3456A contribute to this? In the past, I've measured short term variance (n=1000) to be about 5 * 10-15 V2 on the lowest DCV range with 100 NPLC and shorted inputs (three orders of magnitude less than your measurement), but I'm not sure how much sample variation there is between multimeters. I would expect long term variance to be larger, since I'd expect there to be some slight drift in the gain, eg. due to tempco.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 04, 2011, 02:37:24 pm
What about the 24h drift of the gain? For short term (transfer) measurements you can usually ignore drift and expect noise to limit the accuracy, but I wouldn't expect this to be the case over a 24h period. Any estimates from measuring the references?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 04, 2011, 10:37:37 pm
I checked the shorted output of the 3456a at n~ 1000 and it read like your ~ 5 x10-15.  Hours later change in variance suggests the 3456a does drift [ as expected] but its tiny,  and still underneath its rated 10ppm at 24h.   

I think I can separate the noise from the drift component.  Noise causes continuous measurement fluctuations but the net variance mean is ~ 0 [unless its not white noise!], while the drifting shifts the baseline and affects the variance; as analogy noise is to a sine wave riding a DC offset, which represents drift.

I find the biggest cause of drift despite a temperature controlled room remains diurnal shifts caused by day and night in the 24 h period, that's when my measures deteriorated to 2.9 x 10-14 at 24h, and assuming a stable climate [ i.e., no storms to drastically affect humidity or pressure .]  If you measure n=1000 at night for a few hours or in the day for a few hours assuming a stable period where ambient conditions wont' change drastically, the variances are identical, but if you string the two diurnal changes together,  which includes all the ambient condition shifts, the drift component shows itself. 

(http://www.wunderground.com/cgi-bin/histGraphAll?day=27&year=2011&month=11&dayend=3&yearend=2011&monthend=12&ID=KPHL&type=6&width=614)

What about the 24h drift of the gain? For short term (transfer) measurements you can usually ignore drift and expect noise to limit the accuracy, but I wouldn't expect this to be the case over a 24h period. Any estimates from measuring the references?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 05, 2011, 02:13:59 am
I think I can separate the noise from the drift component.  Noise causes continuous measurement fluctuations but the net variance is ~ 0 [unless its not white noise!], while the drifting shifts the baseline and affects the variance; as analogy noise is to a sine wave riding a DC offset, which represents drift.
I believe you're using variance in a non-technical sense, this is confusing in this context. Since variance is a function of the square of the deviation from the mean, it will be non-zero for anything but a constant signal (pure DC). The mean of a white noise process is zero, however, and the variance is finite.

I find the biggest cause of drift despite a temperature controlled room remains diurnal shifts caused by day and night in the 24 h period, that's when my measures deteriorated to 2.9 x 10-14 at 24h, and assuming a stable climate [ i.e., no storms to drastically affect humidity or pressure .]  If you measure n=1000 at night for a few hours or in the day for a few hours assuming a stable period where ambient conditions wont' change drastically, the variances are identical, but if you string the two diurnal changes together,  which includes all the ambient condition shifts, the drift component shows itself.
Of course both the DMM and the voltage reference will be affected by this. So taking samples at 24h intervals would give a lower variance? The drift appears to be dominated by temperature coefficient? If you could measure just the effect of the DMM, you could correlate the ambient temperature with the drift of the DMM, and attempt to correct for this. This is cheaper than to get a proper climate-controlled room ;).

Guess this is why 24h specs are for 23°C ± 1°C.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 05, 2011, 09:15:41 pm
Yes, my bad, error corrected.  white noise, statistical mean = 0, variance finite.

Yes, during a 24 hour period, its unlikely a marked shift will occur to the average of environmental variables, for example from the weather graph dates Nov 27-29 and Dec 3 have less variable temps, humidity, pressure, but different averages for each group, not including the diurnal shifts.  For a weeks worth of data, measures includes the Nov 29-Dec 2 rising barometric pressure as well as falling average temps and humidity.   I've found it only affect uV levels and smaller.

Yes, temperature is undoubted a biggest factor, but its one we can do something about, so climate control in a room will minimize larger variance, but trying to stabilize the last few LSD is a challenge.  Assuming we've put controls on other things like triboelectric, electrochemical, guarding , etc., 

Yes, I've considered it. But as you showed, if the shorted output does represent 3456a internal drift + noise, it only contributes variation in single digit nV,  at the limits of the 3456a resolution.   I didn't have a reference stable below 1uV to give it a try, but with the MV106, I'll try.   

Its possible to estimate a correction similar to these folks [its free to download, hopefully for all]:

http://www.springerlink.com/content/725668231627q136/ (http://www.springerlink.com/content/725668231627q136/)



I believe you're using variance in a non-technical sense, this is confusing in this context.

Of course both the DMM and the voltage reference will be affected by this. So taking samples at 24h intervals would give a lower variance? The drift appears to be dominated by temperature coefficient? If you could measure just the effect of the DMM, you could correlate the ambient temperature with the drift of the DMM, and attempt to correct for this. This is cheaper than to get a proper climate-controlled room ;).

Guess this is why 24h specs are for 23°C ± 1°C.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 05, 2011, 09:48:59 pm
Yes, temperature is undoubted a biggest factor, but its one we can do something about, so climate control in a room will minimize larger variance, but trying to stabilize the last few LSD is a challenge.  Assuming we've put controls on other things like triboelectric, electrochemical, guarding , etc., 
As long as the noise is uncorrelated, averaging will help, so you shouldn't have to worry about some random fluctuations. This won't help against cyclic behavior like the daily temp/humidity/pressure changes, unless you set the sample period to be a multiple of 24h, just like how they set the integration time to be a multiple of a PLC to reject the mains noise.

Yes, I've considered it. But as you showed, if the shorted output does represent 3456a internal drift + noise, it only contributes variation in single digit nV,  at the limits of the 3456a resolution.   I didn't have a reference stable below 1uV to give it a try, but with the MV106, I'll try.   
Shorted output only represents noise and offset drift, not gain drift. Not sure how to characterize gain drift without taking the source into account, it's a lot easier to make good short than a good voltage source. Connecting multiple meters to the same source might help, although I'm not sure if their drift will be uncorrelated if they're the same model with the same LMx99 reference.

Its possible to estimate a correction similar to these folks [its free to download, hopefully for all]:
http://www.springerlink.com/content/725668231627q136/ (http://www.springerlink.com/content/725668231627q136/)
Except for the constant pressure/temperature chamber, I'm guessing ;). You may be able to figure it out just from correlation if you collect enough data. At a first glance their model seems straightforward and quite usable in your case.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: EEVblog on December 06, 2011, 07:01:56 am
There are hardly any reviews of EDC products or discussions of them [ at least via google].  So, very likely this thread will pop up in the near future when EDC or Krohn Hite are searched for.  This device and its derivatives are worthy of more interest from anyone looking a well made, portable affordable voltage standards that can be priced right for small labs.

Already link #2 for "Krohn-Hite voltage standard", with the video at #1

Dave.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 06, 2011, 03:34:35 pm
Good work Dave, your video and forum links pop up as the #1 item when you google "krohn hite voltage standard"  and among the first 10 using "krohn hite MV106 review."

I hope it helps them as it helps you.

The EDC equipment portfolio is not well documented on the web, if not for the KH website.  There are models of theirs that pop up now and then on eBay that have no documentation I can find.    There seems to be evolutionary models of the MV106J, such as the MV105 & MV100 that are identical to the 106 but less switchable voltage ranges.

They have several generations of AC calibrators that rival the then standard Fluke and HP models but are not huge boat anchors, even for its time, such as the CDC-32.   These models are not in the KH site.

(http://i5.ebayimg.com/05/i/06/ba/65/40_1_b.JPG)


FWIW, I'm now testing the MV106 stability at 1uV output and at just over 24h the statistical variance on the 3456a is amazing at 6 x 10-15, or an SD of 77 nV.


There are hardly any reviews of EDC products or discussions of them [ at least via google].  So, very likely this thread will pop up in the near future when EDC or Krohn Hite are searched for.  This device and its derivatives are worthy of more interest from anyone looking a well made, portable affordable voltage standard that can be priced right for small labs.

Already link #2 for "Krohn-Hite voltage standard", with the video at #1

Dave.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 07, 2011, 02:04:23 pm
saturation, 
I find the metrology aspect of this thread is very interesting. (my metrology knowledge is mostly mechanical I have a long way to go in electronics) Your stability testing made me wonder how stable my 8846A is on DC volts. So I followed your lead and did a shorted input test for 12 hours overnight in statistics mode at 100NPLC after a 2 hour warm-up. No special temp control on the room, just thermostat with ceiling mounted forced air heater.  In stat mode the min/max resolution is 0.1nV . After 12 hours overnight with 26000 readings the SD was 180.979nV. I think that would be 0.181 ppm. Is that a good number for a meter of this level?
I have noticed that the 8846A works internally (on the lowest range only) at 1000X higher resolution than the displayed LSD (I do know the difference between resolution and accuracy). You can see the full readings in stat mode or trend plot. Max DCV resolution is 0.1uV but internally is 0.1nV, Max Ohms resolution is 10uOhm but internally is 10nOhm etc. That 1000X resolution is there  from 0.02NPLC to 100NPLC and with digital filter on or off.  It still only has 6.5 digits but on the lowest range but it seems to internally change amplification a decade at a time based on the reading to maximize resolution with a limit of 0.1nV.  Do all 6.5 digit meters do this?

Edit: corrected to 0.1nV  was 01.nV
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 07, 2011, 04:44:35 pm
Yes concur.  To reduce the effects of cyclic variations of daily weather and then climate with the seasons, I compare the Geller SVR reference output only when the weather conditions are identical to the calibration date.  Its essentially identical at 1 year, which is not what estimated predictions suggest it would be, once variations accounted for by weather are controlled.  Thus, I don't even need a make a coefficient to correct my reference voltage.

FWIW, the MV106 is far more immune to weather, as we've had a large change in conditions this week and the statistical variance it caused are minimal.  But can't say yet as have no longer term data as the season progresses.


Can you clarify what 'gain drift' refers too?  The internal amps of the DMM?


As long as the noise is uncorrelated, averaging will help, so you shouldn't have to worry about some random fluctuations. This won't help against cyclic behavior like the daily temp/humidity/pressure changes, unless you set the sample period to be a multiple of 24h, just like how they set the integration time to be a multiple of a PLC to reject the mains noise.

Shorted output only represents noise and offset drift, not gain drift. Not sure how to characterize gain drift without taking the source into account, it's a lot easier to make good short than a good voltage source. Connecting multiple meters to the same source might help, although I'm not sure if their drift will be uncorrelated if they're the same model with the same LMx99 reference.

Except for the constant pressure/temperature chamber, I'm guessing ;). (http://You may be able to figure it out just from correlation if you collect enough data.) At a first glance their model seems straightforward and quite usable in your case.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 07, 2011, 05:47:09 pm
Hello robrenz,

On the bold items, yes, its superb.  I don't know if all 6.5 digits DMMs do this.  Its very possible your DMM has more precision internally than the readouts show at face value. 

A question is how much system noise or drift is created by the DMM's design to make that extra resolution useful?  Assuming your measurement connections were optimal and you got 0.18ppm, its fairly close to my 3456a.  However the 3456a resolution is 100nV [ it could be 10nV, see below]; so the Fluke's extra 0.1nV resolution could be limited by other issues and to use it requires much care and planning; its just a speculation.

IIRC, the 3456a may use internally to 7 digits at the 0.1V scale, so it may really have 10nV resolution instead of the publicized 100nV, but it can only be used via the GPIB port, analogous to what you have with the Fluke.  The Fluke 8846 competes against the defacto standard bench DMM, the Agilent 34401a, so its built to meet and beat its specs. 

The 34401a has many similarities to the 3456a, but its more modern and more portable compared to the rack mounted 3456a.  01a incorporates many measurement design concepts made in the 34xx series DVM: 3455, 3456, 3457, 3458.  The 3458a is a widely used 8.5 digit DMM often found in metrology labs.  The design concepts of these DMM were published in the HP journal when it was around so, its possible competitors have studied those techniques and use them in their designs.


saturation, 
I find the metrology aspect of this thread is very interesting. (my metrology knowledge is mostly mechanical I have a long way to go in electronics) Your stability testing made me wonder how stable my 8846A is on DC volts. So I followed your lead and did a shorted input test for 12 hours overnight in statistics mode at 100NPLC after a 2 hour warm-up. No special temp control on the room, just thermostat with ceiling mounted forced air heater.  In stat mode the min/max resolution is 0.1nV . After 12 hours overnight with 26000 readings the SD was 180.979nV. I think that would be 0.181 ppm. Is that a good number for a meter of this level?
I have noticed that the 8846A works internally (on the lowest range only) at 1000X higher resolution than the displayed LSD (I do know the difference between resolution and accuracy). You can see the full readings in stat mode or trend plot. Max DCV resolution is 0.1uV but internally is 0.1nV, Max Ohms resolution is 10uOhm but internally is 10nOhm etc. That 1000X resolution is there  from 0.02NPLC to 100NPLC and with digital filter on or off.  It still only has 6.5 digits but on the lowest range but it seems to internally change amplification a decade at a time based on the reading to maximize resolution with a limit of 01.nV.  Do all 6.5 digit meters do this?
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 07, 2011, 05:59:21 pm
Thanks saturation,
I am restarting the test again to do at least 24 hours and I am recording the min/max temperature at the case during  the test.

robrenz
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 07, 2011, 06:35:06 pm
Can you clarify what 'gain drift' refers too?  The internal amps of the DMM?
Amplifiers, dividers, internal voltage reference, ADC and anything in between. All will drift (very slightly). With shorted inputs, you're just measuring offset (which is usually compensated by auto-zero) and noise. Change in gain is irrelevant for this measurement, if your voltage reference would change from 10V to 9.9V, you wouldn't notice any difference with shorted inputs. Not saying I expect much for a well-designed DMM like the HP 3456A, but I can't imagine it'd be zero, just look at the 24h and tempco specs. Of course these are worst case. The gain error is specified by the % of value uncertainty; the offset error/noise is represented by the % of full scale error.

After 12 hours overnight with 26000 readings the SD was 180.979nV.
This is very close to the 170 nV saturation reported for n ~ 12000. The number of samples is sufficiently large that both should represent be a very good estimate of the population standard deviation, assuming it's purely determined by noise. Temperature shouldn't really matter for this measurement, although you might expect slightly more noise at higher temperatures. The difference between say 293K and 300K is unlikely to make much difference, I would expect. I believe the HP 3456A has better accuracy specs than the Fluke 8846A (it was a reference class meter in its day, similar to the 3458A now), not sure about noise.

I have noticed that the 8846A works internally (on the lowest range only) at 1000X higher resolution than the displayed LSD (I do know the difference between resolution and accuracy). You can see the full readings in stat mode or trend plot. Max DCV resolution is 0.1uV but internally is 0.1nV, Max Ohms resolution is 10uOhm but internally is 10nOhm etc. That 1000X resolution is there  from 0.02NPLC to 100NPLC and with digital filter on or off.  It still only has 6.5 digits but on the lowest range but it seems to internally change amplification a decade at a time based on the reading to maximize resolution with a limit of 01.nV.  Do all 6.5 digit meters do this?
Some do, especially Keithley. A random result via GPIB from a 5.5 digit meter with shorted inputs on its 300mV range (expect resolution 1 uV)
+000.0042E-3 (resolution .1 uV)

From a 6.5 digit meter on 100mV range:
4.75412547E-08 (.1 fV resolution!)

Especially the last one is completely ludicrous, I consider this just artifacts of the calculation and completely insignificant compared to the noise floor. The only advantage of the extra resolution is that it reduces quantization error, although it should be random for this (haven't studied it in that much detail). This is a real issue when averaging down to the last digit. Datron actually introduced noise in the measurement and then averaged it to get a better resolution. That's how they achieved 7.5 digits, I wouldn't be surprised if it was still used in the Fluke 8508, since they bought Datron to get their high accuarcy DMM technology). Most other brands return the same resolution via GPIB as on the display. Except for the 3457A which is specified as a 6.5 digit meter, but 7.5 digits of resolution via GPIB. This does not translate in improved accuracy, however, since the 3456A is superior in that regard.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 07, 2011, 07:41:20 pm
I have noticed that the 8846A works internally (on the lowest range only) at 1000X higher resolution than the displayed LSD (I do know the difference between resolution and accuracy). You can see the full readings in stat mode or trend plot. Max DCV resolution is 0.1uV but internally is 0.1nV, Max Ohms resolution is 10uOhm but internally is 10nOhm etc. That 1000X resolution is there  from 0.02NPLC to 100NPLC and with digital filter on or off.  It still only has 6.5 digits but on the lowest range but it seems to internally change amplification a decade at a time based on the reading to maximize resolution with a limit of 01.nV.  Do all 6.5 digit meters do this?
Some do, especially Keithley. A random result via GPIB from a 5.5 digit meter with shorted inputs on its 300mV range (expect resolution 1 uV)
+000.0042E-3 (resolution .1 uV)

From a 6.5 digit meter on 100mV range:
4.75412547E-08 (.1 fV resolution!)

Especially the last one is completely ludicrous, I consider this just artifacts of the calculation and completely insignificant compared to the noise floor. The only advantage of the extra resolution is that it reduces quantization error, although it should be random for this (haven't studied it in that much detail). This is a real issue when averaging down to the last digit. Datron actually introduced noise in the measurement and then averaged it to get a better resolution. That's how they achieved 7.5 digits, I wouldn't be surprised if it was still used in the Fluke 8508, since they bought Datron to get their high accuarcy DMM technology). Most other brands return the same resolution via GPIB as on the display. Except for the 3457A which is specified as a 6.5 digit meter, but 7.5 digits of resolution via GPIB. This does not translate in improved accuracy, however, since the 3456A is superior in that regard.

It seems from what I have tested so far that the extra 3 digits are real not calculated. It is the same at .02NPLC and no digital filtering so I would think in that case it is actually a single reading. This shows the 0.1nV resolution showing on the stat display(http://C:\Documents and Settings\ROBIN\Desktop\PC071625.JPG) The max and min are not calculated values and the last two trailing zero's are always there.  The avg. and Sd are calculated and show more digits.  Keep in mind as soon as the max or min value needs another most significant digit to display the value, the resolution will decrease one digit.  So this meter only does this as values approach zero. As soon as 6.5 digits are required to display the value, the stat display will have the same resolution as the main display.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 07, 2011, 08:02:57 pm
I talked to Fluke tech support and the 8846A does in fact do autoscaling to maximize the available internal resolution past the stated max main display resolution.  It does this anywhere in the range not just around zero. So if you hooked it up to a precision 10V source and the source only deviated a few uV it would autoscale to the max resolution on the stat or trend plot displays. It will always use the max internal resolution possible. I have not tried this yet at a voltage other than zero but as soon as I am done the 24hr stability test is done I will.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 07, 2011, 08:06:22 pm
As an aside here's interesting cal lab cumulative report on the stability of multiple 34401a compared to  Keithley 2000s.

There is a public discussion on a metcal technician forum about the Fluke 8845a, most of their concerns were about input jack quality.

http://www.pmelforum.com/index.php?topic=1429.15 (http://www.pmelforum.com/index.php?topic=1429.15)
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: alm on December 08, 2011, 12:02:47 am
I talked to Fluke tech support and the 8846A does in fact do autoscaling to maximize the available internal resolution past the stated max main display resolution.
Any spec for this internal resolution?

  It does this anywhere in the range not just around zero. So if you hooked it up to a precision 10V source and the source only deviated a few uV it would autoscale to the max resolution on the stat or trend plot displays. It will always use the max internal resolution possible. I have not tried this yet at a voltage other than zero but as soon as I am done the 24hr stability test is done I will.
As I stated, the value is somewhat limited if the noise floor is ~0.2 ppm, apart from reducing quantization error.

As an aside here's interesting cal lab cumulative report on the stability of multiple 34401a compared to  Keithley 2000s.
Shows why the 34401A has been the industry standard forever. Funny how the far majority of the out of tolerance points for the Keithley are in the resistance ranges, increasing with the higher ranges, but dropping again in the 10/100 Mohm ranges, probably because of the looser tolerances.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 08, 2011, 12:33:53 am
I talked to Fluke tech support and the 8846A does in fact do autoscaling to maximize the available internal resolution past the stated max main display resolution.
Any spec for this internal resolution?

Max resolution is 0.1nV but it never exceeds 6.5 digits of information. You can see an example in the picture of the screen in a post below.
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 08, 2011, 02:46:11 pm
8846A results from a 24 hour test of shorted input DcV   Much better than the first 12 hour test
Max. reading -2.9535uV  Min. reading -1.8006uV  (zero does not function in stat mode these include the meter offset)
total 24 hour span 1.1529uV
@ 1hour 2064 readings SD=60.96nV
@ 6 hours 12387 readings SD=83.47nV
@ 12 hours 24774 readings SD=86.1nV
@ 24 hours 49000 readings SD=133.32nV

Air temperature at case during test: Max. 30.89 degC Min. 22.67degC  8.2 degC temp variation.  (I forgot about my thermostat setting back the temp overnight)
Will I see a big difference if I run again with the temp more stable? (no temp setback overnight)
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 08, 2011, 05:04:55 pm
Thanks robrenz for making this available.  As expect, its definitely doing as well, if not better, than the 3456a. 

In the end, this exercise is just to get insight into how the the whole system drifts by monitoring the offset voltages, and using those results to guestimate how the DMM drift will contribute errors when measuring a high quality voltage reference.

As for the effect of setting your thermostat, its unknown until you try, it should be easy to do if you have time and and it could settle any questions you have of its stability. 

At 6.5 digits, we are edging to the limits of DMM design.


8846A results from a 24 hour test of shorted input DcV   Much better than the first 12 hour test
Max. reading -2.9535uV  Min. reading -1.8006uV  (zero does not function in stat mode these include the meter offset)
total 24 hour span 1.1529uV
@ 1hour 2064 readings SD=60.96nV
@ 6 hours 12387 readings SD=83.47nV
@ 12 hours 24774 readings SD=86.1nV
@ 24 hours 49000 readings SD=133.32nV

Air temperature at case during test: Max. 30.89 degC Min. 22.67degC  8.2 degC temp variation.  (I forgot about my thermostat setting back the temp overnight)
Will I see a big difference if I run again with the temp more stable? (no temp setback overnight)
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 08, 2011, 08:56:21 pm
Amp-hour posted this notice, Keithley's latest text on the how-and-why of low level measurements is available free:

http://www.keithley.com/data?asset=9538 (http://www.keithley.com/data?asset=9538)

If the link fails try the main search:

http://www.keithley.com/at/027.html (http://www.keithley.com/at/027.html)

Search for: Low Level Measurements Handbook
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: robrenz on December 08, 2011, 09:27:45 pm
They will send you a paperback version for free also. I got one 6 months ago.
http://www.keithley.com/knowledgecenter (http://www.keithley.com/knowledgecenter)
Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: saturation on December 16, 2011, 05:33:37 pm
A manual for the MV106 and other venerable older equipment are available here:

http://www.ko4bb.com/manuals/index.php (http://www.ko4bb.com/manuals/index.php)

Title: Re: EEVBLOG #210 Krohn-Hite Voltage Standard
Post by: Galaxyrise on November 10, 2013, 10:55:07 pm
I'm playing with an EDC 2902 now, built in 12/79, which seems to be based on a very similar design.  Documentation has been sparse, and the video was great at providing some insights for me.  Thanks, Dave :)

I think the mystery device is a precision 3 ohm resistor. 4 terminal so two terminals can be used for the series current through the resistor, and two outputs to provide the precise 3 ohm output without any lead resistance error.
My mystery device is marked A2405-1, unlike Dave's which is A3550.  A3550 does appear on Dave's schematic by the 3R resistor, and between OUT and COM, I do measure 2.984.  But between IN and OUT, I measure 296.8.  between COM and IN, I measure 299.8.  So I think the A3550 is both the 3R and the 297R.

Thanks to the video, I know what the label on the right is referring to, yay :)  (It's been cut off, but it reads 1893, 6.372, @7.5) .  However, in my unit, the zener looks a LOT different than Dave's; I mistook it for a resistor at first!  (The little black glassy component under the yellow wire, banded with grey white yellow orange.) Stability of mine is pretty poor compared to modern voltage references, and I may try replacing it with something else.

I can also offer some insight as to why Dave's chopper amp was on a daughter board.  Check out the amp on my board!  It's got the same pin arrangement on the bottom as Dave's has.  Another possible project for me :)

All the diodes in this unit (and only the diodes) have a U-bend added to the lead on the cathode side. Some kind of strain relief, I presume?