EEVblog => EEVblog Specific => Topic started by: EEVblog on July 29, 2012, 03:28:50 am

Title: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: EEVblog on July 29, 2012, 03:28:50 am
EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4 (https://www.youtube.com/watch?v=ZVJ5uuvAlSo#ws)

Dave.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Psi on July 29, 2012, 05:11:10 am
Nice DC-DC converter. I didn't expect it to work down to such a low voltage.

Could the overshoot be caused by the electronic load taking its time to detect the voltage and apply a load.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: IanB on July 29, 2012, 05:21:39 am
I'm with Psi. I don't see how it makes sense to use an electronic load in constant power mode during turn down and switch on tests. At switch on it will appear like a short circuit, and how many actual loads behave that way? Most real loads have lower power demand at lower supply voltages.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: gxti on July 29, 2012, 06:02:52 am
This is relevant to my interests! I was just looking at using non-isolated modules as an easy way to power projects from the 10-15VDC of a lead acid battery backup (worst case discharge to maximum charging voltage). The isolated same-voltage ones like in the video are also neat for powering isolated RS-485 transceivers and whatnot.

Is there somewhere in the US to buy modules like this that might be cheaper/better than the usual distributors (Digi-Key etc.)? DK's prices are still a savings over rolling my own but I suspect there's even better out there.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Skippy on July 29, 2012, 09:26:36 am
Noob question, but what is the point of these things? You put 5v in, and get 5v out. Is it something to do with the isolation?
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: SeanB on July 29, 2012, 09:36:03 am
Isolation, and a totally separate ground for the secondary side. Helps with noise, and is needed if you want to power say a amplifier or whatever that is on a high voltage rail, like reading the current flowing in a high side switch and converting it to digital to send to a grounded MCU, or to power a high side switch that is fed a control signal via an optoisolator.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: OndraSter on July 29, 2012, 10:01:05 am
What was it near the end?

0.6V in, 0.8A in. 0.067A out and 0.00V out (says the load). It slowly crawls up.
Then you unplug the DC-DC and it goes back to full 5.15V...
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Bored@Work on July 29, 2012, 10:31:55 am
What was it near the end?

The thing blowing up.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: eV1Te on July 29, 2012, 11:22:34 am
Nice DC-DC converter. I didn't expect it to work down to such a low voltage.

Could the overshoot be caused by the electronic load taking its time to detect the voltage and apply a load.

I would be interested in knowing how the electronic load behaves in constant resistance mode instead of constant power or current. Since constant power/current requires the load to adjust it self to the observed input voltage, at the same time the DC-DC converter tries to adjust its output voltage.

Hence they are fighting each other (which can easily explain your  transients during start-up or at the very end of the current range of the electronic load)

(How does the constant resistance mode work in an electronic load, is it still a feedback circuit that adjusts the resistance or is the resistance truly constant during start-up/shut down of a power supply?)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: OndraSter on July 29, 2012, 11:23:38 am
The thing blowing up.

I have figured that much, but how could the bench power supply drop down to 0.6V?
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: johnmx on July 29, 2012, 12:11:46 pm
For my USB power supply I will use the DCP020505P which is also a 2W isolator DC/DC converter with an efficiency of 80%. But there’s no minimum load and several devices can be connected in parallel.

Data sheet:
http://www.ti.com/lit/ds/sbvs011k/sbvs011k.pdf (http://www.ti.com/lit/ds/sbvs011k/sbvs011k.pdf)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Bored@Work on July 29, 2012, 12:19:00 pm
The thing blowing up.

I have figured that much, but how could the bench power supply drop down to 0.6V?

Because of its build in current limit or because all the voltage was dropping over the ammeter.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: OndraSter on July 29, 2012, 12:36:33 pm
0.8A as current limit.. sounds about right then. Thanks.

It couldn't have been ammeter drop. It would have to be rather big resistor which on 10A range it is not. Also he had over 400mA with almost unnoticeable drop. 800mA is two unnoticeables only, not 4.2V. 3.36W in ammeter I don't think would be a good thing.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: mlopes on July 29, 2012, 12:38:54 pm
Just one consideration about the MTBF:
A MTBF of 350,000 hours means the component will have a reliability (working without any failure) of just 37% in the first 350,000 hours. If you do the math considering a reliability of 95% (probability of working without any failure), this value is reduced to just 2 years! [ R(t) = e^(-t / MTBF) ].
Thank you very much for all your effort developing the blog.
You do a really nice job.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: madires on July 29, 2012, 01:32:26 pm
Just one consideration about the MTBF:
A MTBF of 350,000 hours means the component will have a reliability (working without any failure) of just 37% in the first 350,000 hours. If you do the math considering a reliability of 95% (probability of working without any failure), this value is reduced to just 2 years! [ R(t) = e^(-t / MTBF) ].
Thank you very much for all your effort developing the blog.
You do a really nice job.

It's even a little bit more complex :-) Most people understand MTBF the wrong way. MTBF = MTTF (mean time to failure) + MTTR (mean time to repair). A MTBF of 1000h could mean that a device brakes just after 1 hour usage and the repair takes 999 hours on average. If we want to specify the average time a device works until it breaks we should use MTTF! MTBF also implies that a broken device can be repaired (not just swapping the broken device).
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: HLA-27b on July 29, 2012, 01:58:54 pm
Sagan has grown up!

The day is not far when he will say "Don tuwn it on take it apawt"  :)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: KTP on July 29, 2012, 02:01:10 pm
In Australia you can give a 2 year old kid a sharp flat head screwdriver and let him play with live circuits.

In the USA, adults can't even buy magnets:

http://www.washingtonpost.com/business/economy/federal-suit-targets-maker-of-buckyballs-magnets-citing-dangers-for-children/2012/07/25/gJQA1lDg9W_story.html (http://www.washingtonpost.com/business/economy/federal-suit-targets-maker-of-buckyballs-magnets-citing-dangers-for-children/2012/07/25/gJQA1lDg9W_story.html)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: TerminalJack505 on July 29, 2012, 02:23:16 pm
In Australia you can give a 2 year old kid a sharp flat head screwdriver and let him play with live circuits.

In the USA, adults can't even buy magnets:

http://www.washingtonpost.com/business/economy/federal-suit-targets-maker-of-buckyballs-magnets-citing-dangers-for-children/2012/07/25/gJQA1lDg9W_story.html (http://www.washingtonpost.com/business/economy/federal-suit-targets-maker-of-buckyballs-magnets-citing-dangers-for-children/2012/07/25/gJQA1lDg9W_story.html)

This is the same government that will let a person buy an assault rifle, though, so... so... yeah.   :o
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: SeanB on July 29, 2012, 02:33:24 pm
In Australia you cannot even change a light fitting unless you are a sparkie, and you need a plumber to change a tap washer. Then of course you are getting $hafted by Al (live as I say not in the mansion I have, and I will sell you the credits I legislated to make you pay) as well.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: nitro2k01 on July 29, 2012, 03:22:41 pm
I'm thinking the same thing as eV1Te. I would like to see how the DC/DC converter behaves on startup when connected to an actual resistor instead of the electronic load.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: sagdahl on July 29, 2012, 03:40:44 pm
Why in Gods name doesn't Dave use his uCurrent thingy?  ???
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Bored@Work on July 29, 2012, 03:54:20 pm
Why in Gods name doesn't Dave use his uCurrent thingy?  ???

Why should he? And does it work up to 400 mA?
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: eV1Te on July 29, 2012, 05:25:05 pm
Why in Gods name doesn't Dave use his uCurrent thingy?  ???

Why should he? And does it work up to 400 mA?

Unfortunately it only goes to 300 mA, but just change one resistor I believe and it could go higher. Then the burden voltage-drop would be significantly lower than that of the multimeter directly.

Dave have you reviewed the electronic loads performance under different conditions in any video? (e.g. how constant resistance vs. current/power modes behaves compared to a real resistor?)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: SeanB on July 29, 2012, 05:46:14 pm
Load will not behave as a resistor under any circumstance. It is designed to be non linear, either keeping a constant voltage, constanty current or constant power over a range of input. A resistor is very linear, increase voltage and current increases, and power squares.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: eV1Te on July 29, 2012, 06:09:02 pm
Load will not behave as a resistor under any circumstance. It is designed to be non linear, either keeping a constant voltage, constanty current or constant power over a range of input. A resistor is very linear, increase voltage and current increases, and power squares.

I assumed that the constant-resistance-mode of the load actually simulates a linear resistor (mosfet with controlled gate voltage that gives specific on-resistance or similar).

If the constant-resistance-mode is some kind of feedback-loop that is non-linear during switch-on/off then I do not understand what the purpose if it is?!

Would very much want to see the scope measure the "real" current (through .1 ohm resistor or similar) and voltage drawn from the load during transient changes.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: tom66 on July 29, 2012, 09:33:46 pm
Due to the instability of the 50 Hz oscillation, I'm thinking it's a coincidence it lines up with the mains frequency.

I had a 5V power supply (wall wart thing) that I thought was an iron transformer based design. I measured the output as rippling around 60 Hz under no load. But I thought, we use 50 Hz in this country, what's up? I put it under load and it increased to several kHz... Maybe some control loop in the constant power circuit has a 50 Hz bandwidth.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: mlopes on July 29, 2012, 10:20:43 pm
Just one consideration about the MTBF:
A MTBF of 350,000 hours means the component will have a reliability (working without any failure) of just 37% in the first 350,000 hours. If you do the math considering a reliability of 95% (probability of working without any failure), this value is reduced to just 2 years! [ R(t) = e^(-t / MTBF) ].
Thank you very much for all your effort developing the blog.
You do a really nice job.

It's even a little bit more complex :-) Most people understand MTBF the wrong way. MTBF = MTTF (mean time to failure) + MTTR (mean time to repair). A MTBF of 1000h could mean that a device brakes just after 1 hour usage and the repair takes 999 hours on average. If we want to specify the average time a device works until it breaks we should use MTTF! MTBF also implies that a broken device can be repaired (not just swapping the broken device).
Just one consideration about the MTBF:
A MTBF of 350,000 hours means the component will have a reliability (working without any failure) of just 37% in the first 350,000 hours. If you do the math considering a reliability of 95% (probability of working without any failure), this value is reduced to just 2 years! [ R(t) = e^(-t / MTBF) ].
Thank you very much for all your effort developing the blog.
You do a really nice job.

It's even a little bit more complex :-) Most people understand MTBF the wrong way. MTBF = MTTF (mean time to failure) + MTTR (mean time to repair). A MTBF of 1000h could mean that a device brakes just after 1 hour usage and the repair takes 999 hours on average. If we want to specify the average time a device works until it breaks we should use MTTF! MTBF also implies that a broken device can be repaired (not just swapping the broken device).

In this case I am supposing the MTTR << MTTF, so I can consider MTBF approximately the MTTF. Only the MTBF is available in the manufacturer’s datasheet and it is not feasible for him to predict the MTTR (once he does not know the type of application and repair complexity).
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: mikeselectricstuff on July 29, 2012, 11:05:51 pm
I have a hard time believing that any MTBF figure has any real validity or usefulness - is it really possible to evaluate every conceivable failure mode and assign a realistic probability to it?
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: EEVblog on July 30, 2012, 01:14:33 am
I have a hard time believing that any MTBF figure has any real validity or usefulness - is it really possible to evaluate every conceivable failure mode and assign a realistic probability to it?

It looks good on paper  ;D

Dave.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: hlavac on July 30, 2012, 01:26:22 am
I have a bad feeling about the health of Dave's electronic load since the power supply incident in EEVBlog #315 (http://www.eevblog.com/2012/07/21/eevblog-315-korad-ka3005p-reviewfail/).
Dave you should check it out, the 50Hz mystery is not normal...
Or is it the scope sitting on top of it and picking something up from the load?
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: sagdahl on July 30, 2012, 02:26:21 am
I was not aware that the uCurrent limits at 300mA. What is the cure for the burden voltage at 400mA, Dave?

Maybe a power supply with sense inputs where you connect the sense-wires after the current-meter?

A power-supply with a good built-in current meter?

An external shunt-resistor of 0,1 Ohm, creating an maximum voltage of 40 mV?

uCurrent V2.0 which can handle 1A?

I agree with Dave that this is really strange that Fluke and other quality instruments hasn't solved this drawback. Fluke-guys, come up with a solution!


Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: IanB on July 30, 2012, 02:30:33 am
I was not aware that the uCurrent limits at 300mA. What is the cure for the burden voltage at 400mA, Dave?

Unless you are really in need of accuracy just use the 10 A range on the meter. That's what everyone else does.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: EEVblog on July 30, 2012, 03:35:40 am
I was not aware that the uCurrent limits at 300mA. What is the cure for the burden voltage at 400mA, Dave?

I showed it. Use the 10A range. Or roll your own solution if needed.

Dave.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: BobC on July 30, 2012, 04:09:06 am
I'm glad Dave is an EE instead of, say, a veterinarian.

Imagine the fuss if he had done all this to a kitten...
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: Rufus on July 30, 2012, 04:13:52 am
What is the cure for the burden voltage at 400mA

In this case turn up the supply voltage to compensate or connect the PSU sense inputs to do it automatically (or just read current from the power supply in the first place).
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: SteveyG on July 30, 2012, 07:06:12 am
... connect the PSU sense inputs to do it automatically (or just read current from the power supply in the first place).

While I was watching this I was surprised that he had no PSU with separate sense terminals.  :o
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: madires on July 30, 2012, 01:30:45 pm
I have a hard time believing that any MTBF figure has any real validity or usefulness - is it really possible to evaluate every conceivable failure mode and assign a realistic probability to it?

It's more like guessing a probability based on your experience, knowledge and statistics you may have. Then you check the dependencies of all atomic elements of your component or device and calculate the total reliability. But that may end up being a quite complex task. I've done that for communication networks. If you promise in your SLA such and such a reliability for your network, you should be able to deliver that. A vendor sold ATM switches with a reliability of 99.9999% (could be even some 9s more :-). Common sense tells me that's BS but I had to use that number for the calculation because there were no other numbers. If something goes wrong you can blame the vendor for giving you wrong numbers. Job done :-)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: sagdahl on July 30, 2012, 01:42:48 pm
You are of course right. There is no need for better resolution then the 10A range will give us.
Looking forward to next episode.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: mikeselectricstuff on July 30, 2012, 03:50:10 pm
I have a hard time believing that any MTBF figure has any real validity or usefulness - is it really possible to evaluate every conceivable failure mode and assign a realistic probability to it?

It's more like guessing a probability based on your experience, knowledge and statistics you may have. Then you check the dependencies of all atomic elements of your component or device and calculate the total reliability. But that may end up being a quite complex task. I've done that for communication networks. If you promise in your SLA such and such a reliability for your network, you should be able to deliver that. A vendor sold ATM switches with a reliability of 99.9999% (could be even some 9s more :-). Common sense tells me that's BS but I had to use that number for the calculation because there were no other numbers. If something goes wrong you can blame the vendor for giving you wrong numbers. Job done :-)
And any failure to meet the MTBF can be blamed on insufficient sample size...!
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: LaurenceW on July 30, 2012, 10:28:37 pm
Dave, I too think your BK electronic load may itself be contributing to these 50Hz ripples at low current, and the step overshoots at switch on. I don't think it's mains pickup, but instability in the system. Simple test - can you stick a 12V battery (simple linear chemical device! no active electronics!) and maybe a series resistor of 10R across your electronic load, and measure the voltage on your scope across the BK when set to a current of 1 mA? Also, apply the 12V at a power limit of 1W, and see what voltage it loads the battery down to, when you abruptly switch the load on?.

An interesting video, all the same. Let's see you testing that 1KV isolation voltage! :)
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: EEVblog on July 30, 2012, 11:12:26 pm
(or just read current from the power supply in the first place).

Not generally a good idea, as the current can change depending upon the actual supply voltage to the chip.
In this case when you are calculating power, you need to know the current and voltage, or you'll end up with significant error.

Dave.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: rbk17c on July 31, 2012, 07:06:32 pm
About the *evil* shorting of output: Others (ReCom) specify a max 10s shot, else-wise get the ~10% more expensive '/P' model, witch has 'Short Circuit Protection'.

Oh, and you can now get 3W in a (smaller than Dave's) SIP4 package like MuRata's MEE3S0515SC.
While this device also has a 1kV isolation, note that this is tested for only 1s, and that a continual 1kV will kill the device.
Do you think this is a common limitation? Me too.
/holger
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: RJSC on July 31, 2012, 11:21:17 pm
Oh well...
I was hoping you'd be using an Adjustable DC-DC converter as a tracking pre-regulator so that the supply would go over 5V...
Will you consider it?

Someone can argue we could use a fixed 20V converter, but than we'd be wasting a large amount of power at the lower voltages and limited to 100 mA current mo matter the voltage.
Title: Re: EEVblog #324 - DC-DC Converter Testing - USB Supply Part 4
Post by: EEVblog on August 01, 2012, 12:00:23 am
Oh well...
I was hoping you'd be using an Adjustable DC-DC converter as a tracking pre-regulator so that the supply would go over 5V...

That's exactly what I'm doing. This is just for the isolation.

Dave.