Author Topic: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed  (Read 28238 times)

0 Members and 1 Guest are viewing this topic.

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Dave cracks out the whiteboard and side cutters to explain two lesser known facts about x1 oscilloscope probes:
- Why the bandwidth is much lower in x1 mode than x10 mode
- Why oscilloscope probe cables aren't like ordinary coax cables


 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #1 on: April 12, 2013, 01:35:07 pm »
To complement this nice video, a nice article written by Dave's mate Doug Ford ..

-> The Secret World of Oscilloscope Probes (PDF), recommended for download.  :-+

Offline ConKbot

  • Super Contributor
  • ***
  • Posts: 1384
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #2 on: April 12, 2013, 03:23:46 pm »
I was going to say Ive seen a pdf on the high frequency loss transmission line bit of oscilloscope probes before, but couldnt find it, glad to see its handled already.
 

Offline Winston

  • Regular Contributor
  • *
  • Posts: 121
  • Country: us
    • IC Die Photography
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #3 on: April 12, 2013, 03:27:40 pm »
Doubleplus good video!  I never would have guessed that lossy transmission line cause.
 

Offline free_electron

  • Super Contributor
  • ***
  • Posts: 8517
  • Country: us
    • SiliconValleyGarage
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #4 on: April 12, 2013, 04:41:21 pm »
The reason for having x1 probes is simple: measuring in the millivolts range...
You dont want to atteniate a signal and then send it through an amplifier at the backend.
Professional Electron Wrangler.
Any comments, or points of view expressed, are my own and not endorsed , induced or compensated by my employer(s).
 

Offline nitro2k01

  • Frequent Contributor
  • **
  • Posts: 843
  • Country: 00
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #5 on: April 12, 2013, 04:54:18 pm »
Am I right in saying that the equivalent circuit for this type of coax is reminiscent of a Butterworth filter topology?
Whoa! How the hell did Dave know that Bob is my uncle? Amazing!
 

Offline CarlG

  • Regular Contributor
  • *
  • Posts: 155
  • Country: se
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #6 on: April 12, 2013, 07:52:05 pm »
Very interesting video! I knew that probe cables aren't ordinary coaxes, and that one shall treat them with care (as well as ordinary coaxes!) but I haven't bothered to find out about the details. Didn't know about the wiggling of the center lead, for example. Gonna tip my coworkers about this video :)
 

Offline KedasProbe

  • Frequent Contributor
  • **
  • Posts: 646
  • Country: be
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #7 on: April 13, 2013, 06:48:17 am »
So is it easy to attach a normal coax to a cheap probe? (to make a separate X1 probe)
And how much more bandwidth can be expected.
or maybe the 9Mohm can be replaced with an 1Mohm?

I assume there does not exist a very cheap 1X only probe with like 50Mhz bandwidth.
or maybe X2 with higher bandwidth.


Edit: this is claimed to have a bandwidth for 60Mhz (no probe just clips)
http://www.amazon.com/P1011-Alligator-Oscilloscope-Probe-Coaxial/dp/B00AK9GCQW
I bought 2 x100Mhz probes for only 5 euro each on ebay, will do some 'surgery' myself.
See if I can make a X2 probe.

Edit2: found this: X2 500Mhz (very expensive >500Euro)
TPP0502  http://www.tek.com/sites/tek.com/files/media/media/resources/TPP1000_TPP0500_TPP0502_Passive_Voltage_Probes_Datasheet_51W-26151-3.pdf
« Last Edit: April 13, 2013, 08:32:50 am by KedasProbe »
Not everything that counts can be measured. Not everything that can be measured counts.
[W. Bruce Cameron]
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #8 on: April 13, 2013, 10:24:02 am »
So is it easy to attach a normal coax to a cheap probe? (to make a separate X1 probe)
And how much more bandwidth can be expected.
Dave simplified (in my opinion oversimplified) this to stay away from transmission line theory. For best high frequency performance, you want to terminate a 50 ohm coax cable in 50 ohms, otherwise you get reflections and ringing. The scope input is 1 Mohm. There is no such thing as 1 Mohm coax cable. Therefore resistive coax is necessary (not just in 10x probes!) to dampen these effects. Attaching a piece of 50 Ohm coax to a 1 Mohm scope input will have a higher bandwidth, but it will also have some nasty ringing in the pass band. The amplitude of signals at certain frequencies might go up by a factor of two! That can introduce some severe distortion. It will also present a huge capacitive load to your circuit under test.

Another important issue is the source impedance. Probes are usually specced when driven from a terminated 50 ohm source, with a 25 ohm source impedance (50 ohm source in parallel with 50 ohm termination). The 100 pF 1X probe will have an impedance of about 160 ohm at 10 MHz, so the voltage at the probe tip will be about 85% of the open circuit voltage due to probe loading. Now connect the same probe to a voltage divider consisting of two 5 kohm resistors across the same 10 MHz sine wave. The source impedance of this point is 2.5 kohm. The signal at the probe tip is now only 6% of the open circuit voltage. This is why 1x probes can be useless even for 1 MHz signals.

or maybe the 9Mohm can be replaced with an 1Mohm?
Only if you also change the rest of the compensation network.

I assume there does not exist a very cheap 1X only probe with like 50Mhz bandwidth.
Not at 1 Mohm. 1X-only probes will use the same resistive coax (it's not just there to optimize the 10X probe) and will have similar ~10 MHz / 100 pF specs. You can get more bandwidth with low-impedance probes, like the 50 ohm coax that Dave suggested. A piece of 50 ohm coax terminated into 50 ohms will have a very high bandwidth, but the input impedance is only 50 ohms.

Edit: this is claimed to have a bandwidth for 60Mhz (no probe just clips)
Terminated into 50 ohms maybe. Terminated into 1 Mohm no.

Some further reading: Tektronix Oscilloscope Probe Circuits. By far the most detailed work on scope probes. Also contains some information about 1x probes.
 

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #9 on: April 13, 2013, 10:26:32 am »
Dave simplified (in my opinion oversimplified) this to stay away from transmission line theory.

Yes, of course. The video was already long enough, I thought it wasn't the place for transmission line theory or simulation.
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #10 on: April 13, 2013, 10:29:28 am »
I agree, but claiming that the resistive coax is only there to optimize 10x performance is wrong. The issues that you mention for 10x probes (like ringing in the pass band) also exist for 1x probes.
 

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #11 on: April 13, 2013, 10:42:41 am »
I agree, but claiming that the resistive coax is only there to optimize 10x performance is wrong.

I did not mean to imply that (sorry, can't remember my exact words off-hand), what I was saying is that the resistive coax design and the associated compensation network are carefully designed to be optimised for the x10 mode, in order to get the greatest bandwidth in x10 mode, as that is the main performance criteria for the probe. The x1 mode simply shorts out the 9M resistor and comp cap in the probe, and then you "get what you get" in terms of x1 performance. Yes, it will help dampen the signal in x1 mode too.
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #12 on: April 13, 2013, 01:26:52 pm »
A switchable probe is definitely a compromise, which is why you don't see many 500 MHz switchable scope probes. A dedicated 1x probe will have almost all of the issues you mention, though. It has the same (or at least very similar) resistive coax and parasitic capacitance. You don't generally find 1x hi-Z probes with >> 10 MHz bandwidth either.

For a future episode it might be instructive to compare a piece of Z0 coax and a piece of resistive coax, both terminated in 1 Mohm // 15 pF. You could do either simulation, actual measurements or both. Should only take you about 5 minutes ;).
 

Offline JackOfVA

  • Supporter
  • ****
  • Posts: 350
  • Country: us
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #13 on: April 13, 2013, 01:30:57 pm »
The resistive loss cable probe was invented at Tektronix with the key patent filed in 1956. Well worth reading -- http://www.google.com/patents/US2883619

Interestingly, HP had a patent issued many years later for a broadband scope probe with standard lossless (or at least negligible loss) cable. http://www.google.com/patents/US5172051

 

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #14 on: April 13, 2013, 01:55:41 pm »
A switchable probe is definitely a compromise, which is why you don't see many 500 MHz switchable scope probes.

Yokogawa do a 400Mhz switchable, with inversely terrible rated x1
Can't say I've ever seen better than that.
http://tmi.yokogawa.com/ca/discontinued-products/oscilloscopes/voltage-probes/700988-400-mhz-passive-probe/
 

Offline LapTop006

  • Supporter
  • ****
  • Posts: 467
  • Country: au
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #15 on: April 13, 2013, 02:37:11 pm »
The reason for having x1 probes is simple: measuring in the millivolts range...
You dont want to atteniate a signal and then send it through an amplifier at the backend.

Eh, as long as the signal's hot enough that's fine.

I wonder why FET probes aren't more common then, the parts shouldn't be much more then for a 10X probe, especially as the cable might be able to be simplified (that's not actually clear after some quick research).

Sure the current lot of cheap scopes don't have the aux power pins, but I wouldn't be surprised to see them in a future rev, the cost to add them is minimal for another "premium" feature.
 

Offline JackOfVA

  • Supporter
  • ****
  • Posts: 350
  • Country: us
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #16 on: April 13, 2013, 03:59:31 pm »
The reason for having x1 probes is simple: measuring in the millivolts range...
You dont want to atteniate a signal and then send it through an amplifier at the backend.

Eh, as long as the signal's hot enough that's fine.

I wonder why FET probes aren't more common then, the parts shouldn't be much more then for a 10X probe, especially as the cable might be able to be simplified (that's not actually clear after some quick research).

Sure the current lot of cheap scopes don't have the aux power pins, but I wouldn't be surprised to see them in a future rev, the cost to add them is minimal for another "premium" feature.

FET input gates are limited for maximum voltage to reasonably low values, some tens of volts at most and usually closer to +/- 10 or 20V. So long as the signal to be probed is less than maximum gate input voltage all is good. But, when the voltage exceeds the maximum gate voltage the probe will not faithfully reproduce the input voltage. And, at some level of input voltage, the protection mechanisms the probe designer included will break down leading the magic smoke contained inside the probe to be vented to the atmosphere, often accompanied by cursing from the probe owner.

To attenuate the voltage, therefore, you can't use the oscilloscope input attenuator / voltage range knob over much range (if at all), but rather have to apply attenuation directly at the probe tip. There are a few attenuated FET probes available and I seem to recall one FET probe with an optional 10x or 100x attenuator sleeve that fitted over the probe  body, but these are not all that satisfactory a solution to varying the combined probe/oscilloscope vertical gain. For the obvious practical reasons, it is both undesirable and difficult to build the attenuator control into the body of an  FET probe.
 

Offline G7PSK

  • Super Contributor
  • ***
  • Posts: 3861
  • Country: gb
  • It is hot until proved not.
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #17 on: April 13, 2013, 05:15:22 pm »
I measured my probe leads in the times one setting and found that they were only 100.1 ohms I have another od probe that was 346 ohms in the X1 My hundred times probe could not be measured with a DVM so I will have to get the megger to that.
 

Online sonic

  • Regular Contributor
  • *
  • Posts: 59
  • Country: de
    • Homepage
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #18 on: April 13, 2013, 07:53:49 pm »
For a future episode it might be instructive to compare a piece of Z0 coax and a piece of resistive coax, both terminated in 1 Mohm // 15 pF. You could do either simulation, actual measurements or both.

I tried to remove over-/undershot using a Z0 probe and made screenshots, but I'm having difficulties interpreting the results, since I'm not exactly knowing what I'm doing ::)
« Last Edit: April 13, 2013, 08:14:45 pm by sonic »
 

Offline sean87

  • Regular Contributor
  • *
  • Posts: 150
  • Country: nl
  • EE WannaBee
    • EMBEDONIX
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #19 on: April 13, 2013, 10:09:47 pm »
So, that being said, can someone explain, why there is 1x and 10x after all? what is the particular use of each? why not having it fixed on 10x ?
EMBEDONIX - Embedded Systems, Linux and good stuff!
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #20 on: April 13, 2013, 10:16:23 pm »
10x is the most commonly used because it has much more bandwidth and reduced circuit loading compared to 1x. The downside is that it attenuates the signal ten times, which is not a very good idea for low level signals. The 1x mode is traditionally used for power supply ripple measurement, for example. This ripple tends to be 120 Hz or so, with amplitudes in the mV.

But 10x is by far the most versatile. I use 10x probes by default and only occasionally reach for a 1x probe.
 

Offline ddavidebor

  • Super Contributor
  • ***
  • Posts: 1190
  • Country: gb
    • Smartbox AT
EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #21 on: April 13, 2013, 10:17:12 pm »
Because you want to misure small signal of a few millivolt sometimes, and you need the lowest vertical range possible.

So if you are measuring a 8mV signal, in 1mV range with x10 probe you can't do it, you need x1 probe
David - Professional Engineer - Medical Devices and Tablet Computers at Smartbox AT
Side businesses: Altium Industry Expert writer, http://fermium.ltd.uk (Scientific Equiment), http://chinesecleavers.co.uk (Cutlery),
 

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #22 on: April 13, 2013, 10:21:40 pm »
10x is the most commonly used because it has much more bandwidth and reduced circuit loading compared to 1x. The downside is that it attenuates the signal ten times, which is not a very good idea for low level signals. The 1x mode is traditionally used for power supply ripple measurement, for example. This ripple tends to be 120 Hz or so, with amplitudes in the mV.

One point about that is that most PSU measurements are specified over a 20MHz bandwidth, so outside the range of usual run of the mill x1 probes.
Usually near enough, but if you want to do it properly, you'll need something else.
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #23 on: April 13, 2013, 10:52:14 pm »
Don't confuse ripple and noise. Ripple will usually have a fundamental frequency of 100 or 120 Hz for linear power supplies with full bridge rectification, and might be in the tens of kHz for switchers. Measuring ripple on power supplies is common during repair to check for dried out electrolytics.
 

Offline Salas

  • Frequent Contributor
  • **
  • Posts: 292
  • Country: gr
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #24 on: April 13, 2013, 10:55:33 pm »
You could tell a few words about the falling with frequency voltage spec on probes which many may be not aware of. A 300V at 10kHZ probe can be 100V at 1MHZ & 10V at 100MHZ...
 

Offline BiOzZ

  • Regular Contributor
  • *
  • Posts: 237
  • Country: us
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #25 on: April 13, 2013, 11:18:08 pm »
THANK YOU
my professor told us too do that once but never explained why so i forgot it and all the workbooks say use X10 so i just did it but at home i always used x1!
flipped my probe over ... 100mhz/6mhz!!!! i cant even think of how i might have fucked things up ... its like when i was 13 and always had my electrolytics upside down!
My one regret in life is learning to speak English on the internet ...
 

Online xrunner

  • Super Contributor
  • ***
  • Posts: 7517
  • Country: us
  • hp>Agilent>Keysight>???
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #26 on: April 13, 2013, 11:26:29 pm »
I've been in electronics my whole career, but I've never thought about asking this question: Why was x10 decided on in the beginning? Why not x12 or x15 or x20?

I imagine it was so it would be easier to do the math in your head, but now, perhaps a better multiplier would be realistic and/or give better performance?
I told my friends I could teach them to be funny, but they all just laughed at me.
 

Offline BiOzZ

  • Regular Contributor
  • *
  • Posts: 237
  • Country: us
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #27 on: April 13, 2013, 11:37:08 pm »
I've been in electronics my whole career, but I've never thought about asking this question: Why was x10 decided on in the beginning? Why not x12 or x15 or x20?

I imagine it was so it would be easier to do the math in your head, but now, perhaps a better multiplier would be realistic and/or give better performance?

id say its easier to do it with a 1M input impedance, easier to do the math your self, easier for old scopes to calculate and now all scopes work on the OOM scale .. 1X 100X 1000X exc ... sort of why we still use QWERTY even tho it was designed to slow us down

and besides any more than 10x at like 5V noise is going to be more of a problem and scaling exc
My one regret in life is learning to speak English on the internet ...
 

Offline rf-loop

  • Super Contributor
  • ***
  • Posts: 4105
  • Country: fi
  • Born in Finland with DLL21 in hand
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #28 on: April 14, 2013, 08:52:13 am »
Making passive 1x probe so that capasitance is low and resistance is high there is real problem. It is nearly impossible.
Ok, there is some old 34MHz model what have 32pF / 1Mohm (Tek P6101A01.  (I have not seen under 1m cable what still may be littlebit better) Perhaps also some others nearly this.) Also current Agilent N2870A is 1x and 1M, 39pF

But if it can be other than 1x but still useful with low levels, example 2x, situation is totally different.   Example:
Attenuation  2X Bandwidth  500MHz, Input Impedance at the Probe Tip 2 M?, 12.7 pF, Cable Length  1.3 m. (Tek. TPP0502)

This is best passive probe with low voltage divider ratio and "high Z" input what I know.
(but also it is not cheap. Good probes are not cheap. One probe and you can buy 2 entry level DSO) (also 12.7pF is some high for 500MHz, but this is 2x probe)

So, if break this 1x, 10x, 100x etc "rule", it may give some more.
But who want 1.27x probe or 1.57x probe. So, who want make these. No one.
(some scopes have fine adjust for probe factor)

I drive a LEC (low el. consumption) BEV car. Smoke exhaust pipes - go to museum. In Finland quite all electric power is made using nuclear, wind, solar and water.

Wises must compel the mad barbarians to stop their crimes against humanity. Where have the wises gone?
 

Online EEVblogTopic starter

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #29 on: April 14, 2013, 09:25:01 am »
I've been in electronics my whole career, but I've never thought about asking this question: Why was x10 decided on in the beginning? Why not x12 or x15 or x20?
I imagine it was so it would be easier to do the math in your head, but now, perhaps a better multiplier would be realistic and/or give better performance?

x10 was chosen in the days of analog scope when you had to multiply the scale factor yourself. Anything other than x10 or x100 would have been a pain.
x10 is also the classic "order of magnitude" difference that engineers love, and in this case it would have been the only logical compromise between attenuation and extended bandwidth while making the math easy.
 

Offline poida_pie

  • Regular Contributor
  • *
  • Posts: 119
  • Country: au
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #30 on: April 14, 2013, 10:42:17 am »
I found these two links very informative on CRO probes and go down the rabbit hole of transmission line considerations.

http://www.dfad.com.au/links/THE%20SECRET%20WORLD%20OF%20PROBES%20OCt09.pdf
'n
http://www.sigcon.com/Pubs/straight/probes.htm
 

Offline ddavidebor

  • Super Contributor
  • ***
  • Posts: 1190
  • Country: gb
    • Smartbox AT
EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #31 on: April 14, 2013, 11:11:11 am »
So... If i destroy an old probe and connect the tip direct to a coax, can i get a decent bandwich?
David - Professional Engineer - Medical Devices and Tablet Computers at Smartbox AT
Side businesses: Altium Industry Expert writer, http://fermium.ltd.uk (Scientific Equiment), http://chinesecleavers.co.uk (Cutlery),
 

Offline Pentium100

  • Frequent Contributor
  • **
  • Posts: 258
  • Country: lt
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #32 on: April 14, 2013, 06:19:25 pm »
Very interesting video. I always used the probes with x1 setting (unless the voltage was 50V or more), now I'll use the x10 mode unless the voltage is too low (I rarely measure very low voltages and when I do, it's usually audio frequency).
 

Offline landswipe

  • Newbie
  • Posts: 2
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #33 on: April 15, 2013, 10:57:18 am »
I have a ds2000 and the 350mhz rigol probes that came with it.

I noticed that above the 1x 10x switch there is a rubber? plug, if you remove it, there is another adjustment, how does that fit into this?

Thanks

David.
 

Offline Fezder

  • Regular Contributor
  • *
  • Posts: 104
  • Country: fi
  • Lean on, or was it learn on?
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #34 on: April 15, 2013, 12:19:52 pm »
nice video, i thought 10x was only for higher voltages....newer knew about bandwith limits, i thought they are the same at 1x/10x ranges....thanks dave! :)
Both analog/digital hobbyist, reparing stuff from time to time
 

Offline cengland0

  • Contributor
  • Posts: 49
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #35 on: April 15, 2013, 01:02:44 pm »
i absolutely love the technical nature of these videos.  I learn something every time.  You rock Dave.
 

Offline Salas

  • Frequent Contributor
  • **
  • Posts: 292
  • Country: gr
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #36 on: April 15, 2013, 04:06:59 pm »
nice video, i thought 10x was only for higher voltages....newer knew about bandwith limits, i thought they are the same at 1x/10x ranges....thanks dave! :)

As the frequency increases the probe's impedance goes down and its voltage spec goes down too. When quoted in the hundreds of Volt at 10X mode, that is only true for few decades of kHz. It ends up to being a low voltage probe near its bandwidth limit. Another not much talked about catch to watch for.
 

Offline Fezder

  • Regular Contributor
  • *
  • Posts: 104
  • Country: fi
  • Lean on, or was it learn on?
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #37 on: April 15, 2013, 06:12:15 pm »
nice video, i thought 10x was only for higher voltages....newer knew about bandwith limits, i thought they are the same at 1x/10x ranges....thanks dave! :)

As the frequency increases the probe's impedance goes down and its voltage spec goes down too. When quoted in the hundreds of Volt at 10X mode, that is only true for few decades of kHz. It ends up to being a low voltage probe near its bandwidth limit. Another not much talked about catch to watch for.

hmm, interesting, and bit confusing, but makes sense thought :D.
Both analog/digital hobbyist, reparing stuff from time to time
 

alm

  • Guest
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #38 on: April 15, 2013, 07:16:34 pm »
There is usually a curve of maximum input voltage vs. frequency in the datasheet. It's often flat for the 'DC' part of the spectrum, then goes down gradually until it reaches the lower limit which might be in the order of 20 V.
 

Offline KarlMonster

  • Contributor
  • Posts: 13
  • Country: us
  • Don't harsh on my reckless idealism, man!
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #39 on: June 15, 2013, 08:10:34 pm »
So... If i destroy an old probe and connect the tip direct to a coax, can i get a decent bandwich?
Bandwidth yes, but ugly signal. As I think was mentioned in the blog, you need a 50 ohm termnator at the end of the coax before you attach a probe.

I believe (since Dave used bare coax - I didn't see any terminators) that the reason why unterminated coax works between the sig gen and the 'scope is because of the impedance of the sig gen output. The 'scope input is about 1Meg and the coax is rated at 50 ohms, but the sig gen output is something on the order of hundreds of Mega ohms. That's to prevent external signals from wandering upstream into the sig gen. As was also mentioned in the video, ohm specs that are orders of magnitude less can often be ignored; the sig gen dominates the circuit.

So bare coax needs to be terminated for circuits with less than a hojillion ohm impedance.


It's given me an idea for building an active 1x probe, but I'm new to terminating coax. This'll be fun!

Massive Edit: I did not understand what "termination" meant. I thought it was just a resistor in line with the conductor. Some coax has built-in termination, as Dave's cable probably does - though he didn't say so. Also; if you have a terminated coax cable, why attach a probe? A slightly longer conductor sticking out of the end should work fine, unless you need ground clip or other probe attachment.
« Last Edit: June 16, 2013, 03:06:58 pm by KarlMonster »
Experience gained is proportional to value of equipment ruined.
I'm ready for a new learning experience.
 

Offline Computeruser

  • Regular Contributor
  • *
  • Posts: 85
  • Country: ca
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #40 on: June 15, 2013, 10:05:45 pm »
LapTop006I wrote:  "wonder why FET probes aren't more common then"

JackOfVA  wrote:  "FET input gates are limited for maximum voltage to reasonably low values"

I have a Tektronix P6045 FET probe that I purchased used some years back. It still works. It is a 1X probe with a high impedance / low capacitance FET front end and an active amplifier/compensator at the 'scope end. It was very expensive new.

I foolishly attached it to a circuit that had high voltage on it and blew up the front end. Fortunately, when I did that, I could still get (and did get) a refurbish kit from Tektronix and fixed it. It still runs and I limit it to low voltages.

I use passive 10X probes most of the time, and if I need a 1X probe, I accept the limitations.

... C
 

Offline pfm

  • Regular Contributor
  • *
  • Posts: 92
  • Country: us
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #41 on: February 23, 2015, 12:24:59 am »
Just came across this blog and video when searching for the answer to the question of why are 1x probes so low in bandwidth. I also read the pdf and noted the comment -
 "The effect of varying the cable resistance over the range 50 ~ 200ohm per meter can also be seen in Fig.10. Low resistances (50ohm/m) allow transmission-line reflections to build up, giving a peaking response. Larger resistances (200ohm/m) give an overdamped, sagging response. The optimum cable resistance was found to be around 115ohm/m. This gave a response which is substantially flat to nearly 600MHz!"


So based on this would it be practical to just use the shortest coax possible with ~100ohm resistor in series ? Perhaps a trimmer capacitor in parallel for compensation purposes. Wouldn't that be like the perfect 1x probe ?

Just another thought - why not just use a twisted pair cable ? would that be any better (or worse) than a coax probe ?
« Last Edit: February 23, 2015, 12:30:41 am by pfm »
 

Offline zeaflal

  • Newbie
  • Posts: 1
Re: EEVblog #453 - Mysteries of x1 Oscilloscope Probes Revealed
« Reply #42 on: March 24, 2015, 05:22:52 pm »
Just wondering if anyone knows of a source to buy scope probe cable.  We had purchased some years ago and are now looking (without success) for more.

Thanks.
 