EEVblog Electronics Community Forum

EEVblog => EEVblog Specific => Topic started by: EEVblog on January 01, 2014, 01:11:35 am

Title: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 01:11:35 am
Dave tries to find out what's wrong with a Tektronix TDS3054 500MHz 4 channel oscilloscope.
SPOILER: there is no ending...
Auction score video: https://www.youtube.com/watch?v=uqURkr9YVvM (https://www.youtube.com/watch?v=uqURkr9YVvM)

EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair (https://www.youtube.com/watch?v=leC-zSMbtfg#ws)
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: mikeselectricstuff on January 01, 2014, 01:36:33 am
"Distributed element filter" at 18:30 looks more like it's a laser-trimmed part (R or cap/inductor array)  - you can see the laser cut-line on it
 
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: mikeselectricstuff on January 01, 2014, 01:45:43 am
Would at least be worth probing around that diff output & comparing to another channel with the same input.
Also try heating & freezing, and flexing the PCB a little
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Vgkid on January 01, 2014, 01:46:13 am
Thanks for the upload, the maxtek marked chips in the hybrid are rather interesting. It appears that Maxtek(now Infinicon) only manufacturers instruments relating to thin film deposition, and various quartz based sensor assemblies.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: gxti on January 01, 2014, 01:59:06 am
Sample memory would be a good guess. When you have billions of cells made as cheaply as possible and bought on the commodity market it's no surprise that RAM is often the first thing to fail in a computer. That said, from the zoomed in look at the noise it doesn't look much like a single bit error.

I was also going to suggest probing the hybrid output but Mike beat me to it.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: free_electron on January 01, 2014, 02:01:49 am
Maxtec was a spinoff of Maxim.
Maxim for a long time made some custom asics for tektronix. They were eventually borged by Tektronix hence the MaxTec name.

Tek spat em back out after the danaher borging and part of it went back to Maxim.

The hybrid contains Ac/Dc gnd and attenuator relays. the maxtec has an amplifier and divider chain ( voltage divider)

the Maxtec chip also splits the signal in a HF component and LF component. From the signal corruption it looks like the problem is in the HF section as the lf signal is perfect.
I would reflow the analog devices ADG361 square chip of that channel. at 25:58 you see the two differntial pairs. one is hf the other lf. check the pins inbetween. one of the rails may be gone. they are fed using ferrite beads.


The trimming mark you see on the resistor is actually called an r-cut. you have to look at it form the long side. the reason they do that is because blasting it with the laser evaporates the resistance material. as it cools down it partially can re-sublimate in the cut. by going lengthwise they avoid it from getting back in the cut area.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: kxenos on January 01, 2014, 02:12:29 am
Dave, if you probe the differential output of the hybrid would it be possible to see if the problem is already there?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 02:15:58 am
the Maxtec chip also splits the signal in a HF component and LF component. From the signal corruption it looks like the problem is in the HF section as the lf signal is perfect.

Interesting, thanks.

Quote
I would reflow the analog devices ADG361 square chip of that channel. at 25:58 you see the two differntial pairs. one is hf the other lf. check the pins inbetween. one of the rails may be gone. they are fed using ferrite beads.

No ferrite beads in sight.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Jay_Diddy_B on January 01, 2014, 02:22:13 am
Hi,
I would guess that the front end is o.k.

What happens if you move the trace up and down the screen?

I would guess that there is something wrong after the ADC. I have seen similar displays on TDS5xx series if there is a bad connection in the acquisition memory.

Regards,

Jay_Diddy_B
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: free_electron on January 01, 2014, 02:30:04 am
backside board… actually side facing the front panel.

Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Hydrawerk on January 01, 2014, 02:43:12 am
Well, the TDS3054 internal construction is not solid... For such an expensive scope (even today) i would expect more steel parts... Compare it to the DSOX2000, that is full of sheet metal.
EEVblog #144 - Agilent 2000 X Series Infiniivision Oscilloscope Teardown (https://www.youtube.com/watch?v=fGxusFBMDR0#ws)
Or even Tektronix MDO4000 series.
Tektronix MDO4000 Oscilloscope Teardown - EEVblog #199 (https://www.youtube.com/watch?v=J9_JxRiuS5A#ws)
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: david77 on January 01, 2014, 03:25:37 am
I have no idea what I'm talking about, but from the video it seems to me that only the positive side of the signal is affected by that weird noise.
So I'm left thinking could it  be some issue with biasing/decoupling of something in the front end? Maybe a dead cap or a very noisy power rail?

Seeing as what people still pay for these scopes on ebay it is probably even worth your while to look a bit further.

I guess a proper manual is not available for this machine, right?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 04:03:49 am
backside board… actually side facing the front panel.

There are only 4 SMD ferrites on the whole board, all under the 5th chip between the ADC's, and they test fine.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: NiHaoMike on January 01, 2014, 04:14:05 am
I was once assigned a 12GHz Agilent scope to use for testing. There were many problems with it, as in one channel would not properly recognize active probes, another channel was completely dead (flat line), and yet another channel was noisy. They just swapped out the scope for a new one and I never heard what was wrong with the old one.

Try tapping the board with a pen or other plastic object to see if it's an intermittent connection.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 04:24:16 am
Just shot a Part 2 video on this.
No surprises for guessing the outcome...
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: member_xyz on January 01, 2014, 04:52:55 am
I find it frustrating to watch your fault finding/repair videos.
Initial mechanical inspection and some speculation is OK to a certain point, but don't go overboard.
I don't kow why you avoid using electronic test equipment to do at least some basic measurements?
You have a multi channel scope, you could easily do some signal comparisons on the front end.
At the end of the day it may not resolve the issue, but for **** sake you are running en electronics blog/forum.








Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: calexanian on January 01, 2014, 05:01:49 am
I find it frustrating to watch your fault finding/repair videos.
Initial mechanical inspection and some speculation is OK to a certain point, but don't go overboard.
I don't kow why you avoid using electronic test equipment to do at least some basic measurements?
You have a multi channel scope, you could easily do some signal comparisons on the front end.
At the end of the day it may not resolve the issue, but for **** sake you are running en electronics blog/forum.


I like the commentary on the mechanical stuff. If I wanted to look at a waveform on a scope all day I would just stay at work and get something done. This is more fun!  O0
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: TooOldForThis on January 01, 2014, 05:32:05 am
What does the channel look like in single sweep mode?  If there is any sign of a changing signal when no new data is flowing into the sample memory, then the problem must be in the sample memory or display logic.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: taemun on January 01, 2014, 05:57:50 am
The waveform disappearing seems to be linked to the acquisition speed (SPS). At 5:20, you go into the Acquire menu, and on "Normal" horizontal resolution, you're pulling 10MSPS (look at the bottom right), and it has a display, and then "Fast" is 500kSPS, which doesn't have a display. At 5:25 you increase sample rate to 5MSPS (10us/div @ "Fast") and that has a display again.

At 5:54 it works on "Normal" down to 5MSPS (200us/div @ "Normal"), then dies below that.

I'd guess that something changes mode below 5MSPS.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Kryoclasm on January 01, 2014, 06:13:10 am
Pull hybrid 3 & 4 and plop 4 in 3's spot. 

That will confirm or eliminate that part of the board.

PITA, but would be interesting to see.  :o
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Eliminateur on January 01, 2014, 06:13:28 am
Dave, if you probe the differential output of the hybrid would it be possible to see if the problem is already there?
same thing i was thinking, why not probe the diff output when fed with the test signal and compare it to the same output from one of the good channels?, it might help pinpoint the problem better.
You could also probe the rest of the pins as well and compare
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: taemun on January 01, 2014, 06:16:17 am
Pull hybrid 3 & 4 and plop 4 in 3's spot. 

That will confirm or eliminate the part of the board.

PITA, but would be interesting to see.  :o


Or maybe probe the outputs first. Noise as on the display would be super easy to see, surely. The EEVblog slogan isn't "don't probe it, solder it mindlessly!".
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: georges80 on January 01, 2014, 06:38:59 am
Well, if google has anything to say:

http://www.seekchip.com/icstock-m/MM9576-VJG.html (http://www.seekchip.com/icstock-m/MM9576-VJG.html)

Implies those aren't ADC chips but some kind of fancy fifo/ram controller.

My 'guess' is that the maxtek chips would have the adc (or equivalent) - I'd assume one would keep all the analog stuff on the hybrid module versus trying to feed it down to the 'digital' board.

The MM9577 again, if you believe google:

http://www.seekchip.com/icstock-%EF%BC%AD/MM9577-ACE.html (http://www.seekchip.com/icstock-%EF%BC%AD/MM9577-ACE.html)

is an ADC (not very fast).

Good luck :)

cheers,
george.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Maxlor on January 01, 2014, 06:41:12 am
I find it frustrating to watch your fault finding/repair videos.
Initial mechanical inspection and some speculation is OK to a certain point, but don't go overboard.
I don't kow why you avoid using electronic test equipment to do at least some basic measurements?
You have a multi channel scope, you could easily do some signal comparisons on the front end.
At the end of the day it may not resolve the issue, but for **** sake you are running en electronics blog/forum.
Yeah... as a software engineer, the random poking around kind of makes me cringe. The term we use for that is shotgun debugging. My approach would be to use bisection to find the problem area. In this instance, you've guessed at what some of the traces on the boards do, let's probe them with a scope at various points to see whether the fault exists before or after that point. Why waste time with random wild shot guesses.

On the other hand, since Dave is commenting on the various bits and pieces he examines as part of the random wild shot guesses (Dave: hey, at least they appear that way in the video. If you've done a more methodical problem analysis, it's not in the video. If I've stepped on your toes here, I'm sorry :) ), I find the video interesting to watch. There's still so much I don't know, and what might be an off-hand side comment to you is really an information nugget to me.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 06:41:19 am
I don't kow why you avoid using electronic test equipment to do at least some basic measurements?
You have a multi channel scope, you could easily do some signal comparisons on the front end.
At the end of the day it may not resolve the issue, but for **** sake you are running en electronics blog/forum.

Because:
a) That's what you do first
b) The video was already long enough
c) A ran out of time (it was new years eve)

Now, I could have not uploaded the video, and kept it until I had finished the repair or done some more stuff, and that's how some bloggers do repair videos. But that's not how I do it. I shoot some content and I upload it, even if it's not finished.
I have already shot a part 2 that does that stuff, and once again I have ran out time to finish it off (new years day + other things to do), but I'm ging to upload that, still an unfinished repair.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Maxlor on January 01, 2014, 06:44:34 am
One other thing just came to mind: it might be interesting to see a sinus signal in dots mode. I'm thinking that if the erroneous dots are still all in the same region at the top, it might be a problem in the analog section, i.e. something causes the signal to run into the maximum. If it mirrors the actual signal, a bit error in the digital signal seems more likely.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: calexanian on January 01, 2014, 06:57:17 am
Well, if google has anything to say:

http://www.seekchip.com/icstock-m/MM9576-VJG.html (http://www.seekchip.com/icstock-m/MM9576-VJG.html)

Implies those aren't ADC chips but some kind of fancy fifo/ram controller.

My 'guess' is that the maxtek chips would have the adc (or equivalent) - I'd assume one would keep all the analog stuff on the hybrid module versus trying to feed it down to the 'digital' board.

The MM9577 again, if you believe google:

http://www.seekchip.com/icstock-%EF%BC%AD/MM9577-ACE.html (http://www.seekchip.com/icstock-%EF%BC%AD/MM9577-ACE.html)

is an ADC (not very fast).

Good luck :)

cheers,
george.

Fi/Fo chips could be just streaming data straight to ram and the main chip figures out what to do with the data later. It lends credence to the problem actually lying in ram. I wonder. Does each channel have a separate ram chip? Could just be a faulty Dram.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 07:15:05 am
Fi/Fo chips could be just streaming data straight to ram and the main chip figures out what to do with the data later. It lends credence to the problem actually lying in ram. I wonder. Does each channel have a separate ram chip? Could just be a faulty Dram.

There are 3 58LC128K18B4 memory chips. I can't find any info on them.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: georges80 on January 01, 2014, 07:24:41 am
There are 3 58LC128K18B4 memory chips. I can't find any info on them.

Your search mojo must be defective :)

This datasheet should be close enough for a 1st stab:

http://www.datasheetarchive.com/dl/Scans-061/DSA2IH0096628.pdf (http://www.datasheetarchive.com/dl/Scans-061/DSA2IH0096628.pdf)

cheers,
george.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 07:29:13 am
Yep, that's the one.
128K x 18bit which was obvious form the part number. Third chip is likely intensity data?
It matches the 9bit ADC claim, needing two memory chips for the 4 channels.
Obviously not fast enough for the 5GS/s, so there must be a pipeline in the ASIC.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: calexanian on January 01, 2014, 07:36:53 am
And we have no forum members at Tek who can dust off somebody to give their thoughts?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: djsb on January 01, 2014, 08:02:55 am
Why not try asking on this yahoo group

http://groups.yahoo.com/neo/groups/TekScopes/info (http://groups.yahoo.com/neo/groups/TekScopes/info)

There are a lot of knowledgeable people in this group who could possibly help.

David.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: peter.mitchell on January 01, 2014, 08:52:12 am
Just shot a Part 2 video on this.
No surprises for guessing the outcome...
I'm guessing you mean BER?
I hope that isnt the case.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: TiN on January 01, 2014, 09:57:31 am
If you have access to high-bw scope with differential probe, maybe worth check that faulty channel diff-pair output and compare it to working channels.
Set same settings on both channels, feed same signal and outputs from front end should be very close :)

In that case you can save time fixing what is not broken (front end voodoo board), if signals to ADC chip do look properly. From scope display it seems like
it's digital logic issue (with memory or acquisition), not analog stuff.

Quote
Now, I could have not uploaded the video, and kept it until I had finished the repair or done some more stuff, and that's how some bloggers do repair videos.

Don't listen hates, just upload everything you have :)  If one have better idea to do repair videos - well, making own blog today is easy as sending two bits to web :D
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Rubi on January 01, 2014, 10:57:10 am
The term we use for that is shotgun debugging.

LOL!!!
As an prof. software engineer I love that term.

Sometimes you are so desperate because nothing works as expected that you fall back in this kind of amateur behaviour and guess what unfortunately it never helps finding the problem,...
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: mikeselectricstuff on January 01, 2014, 11:29:55 am
It might be instructive to look at a ramp waveform, as if the issue is something like a dodgy bit, you may see some regular patterns at certain timebase settings from the monotonic progression of input value - probably most useful in single-shot mode at a high speed to avoid the effects of any filtering, antialiasing etc.

Also, have you thoroughly checked the other channels at a range of timebase an memory settings - dodgy memory may manifest itself on other channels depending on how the memory is allocated in various modes.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: jgorsk on January 01, 2014, 11:32:00 am
Hi Dave,

It may be a good idea to look for help at this Tek forum

http://www1.tek.com/forum/viewforum.php?f=5 (http://www1.tek.com/forum/viewforum.php?f=5)

If I remember correctly, simillar problems with the TDS3000 series were sometimes
solved by running some kind of calibration.

What's in the error log of that scope?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: AndyC_772 on January 01, 2014, 11:57:15 am
+1 for the idea that it's a digital problem, not an analogue one.

When you viewed the waveform in dots mode, the extra random dots you see are clustered together and are well separated from the correct signal. There's a substantial distance in the y-axis between each (faulty) sample point and its neighbours.

If it were an analogue problem, I'd expect to see clipping, or oscillation, or some kind of non-linear distortion - but the resulting signal would be continuous in time and would have constrained bandwidth. What you're seeing on the faulty channel is a signal which appears discontinuous in time and which appears different depending on the ADC sample rate and memory depth in use. That strongly suggests it's a digital problem.

I'd be looking for a memory fault. Maybe try replacing the memory ICs. You can buy some new ones here:

http://www.questcomp.com/questdetails.aspx?pn=MT58LC128K18B4LG-9&utm_source=Findchips&utm_medium=cpc&utm_term=MT58LC128K18B4LG-9&utm_content=PartDetails&utm_campaign=FindchipsWS (http://www.questcomp.com/questdetails.aspx?pn=MT58LC128K18B4LG-9&utm_source=Findchips&utm_medium=cpc&utm_term=MT58LC128K18B4LG-9&utm_content=PartDetails&utm_campaign=FindchipsWS)

Another option in the meantime might be to swap over the ADG361B chips between a working channel and the faulty one. If the fault moves with the chip, at least you know what it is even if a replacement IC isn't readily available.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Anks on January 01, 2014, 11:58:11 am
I would check the diff pair and if that looked good using another channel for reference I would swap chips on the main board.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 12:06:45 pm
Also, have you thoroughly checked the other channels at a range of timebase an memory settings - dodgy memory may manifest itself on other channels depending on how the memory is allocated in various modes.

The other channels seem to work just fine.
I do need to do more thorough testing and calibration etc though.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Rubi on January 01, 2014, 12:13:29 pm
You could remove the ferrites from the differential output on the hybrid and bridge the output from another channel.
So you could see for sure if the hybrid is responsible for the problem.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: tszaboo on January 01, 2014, 12:41:50 pm
My guess would be that you have one bit flipped on the digital side. This can be at the parallel communication somewhere, and the cause for this could be ground bouncing. I would check the decoupling capacitors if any of those are cracked, or if you have significant noise on the power supply right at the ASIC pins.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Rufus on January 01, 2014, 12:54:13 pm
It looks like a digital problem to me. Something like the MSb of the conversion being intermittently hot.

I think more detective work on what was actually being displayed would have been worthwhile. If it is digital and the hybrid is all analog there was no point looking at it.

The disappearing trace at low sweep rates might also give some clues. Is it really disappearing or is the glitch detector always capturing a 'hot' bit at lower rates and you end up with an offset signal above the trigger level? That would indicate the problem is before the trace memory.

If you run with the scope open there are lots of options for comparing signals and heating/cooling/poking things to see if you can make it more or less intermittent.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: PA0PBZ on January 01, 2014, 01:00:40 pm
The wires heat-shrinked with the AC are most likely mainly for the line trigger.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair - FIX?
Post by: jerryuk on January 01, 2014, 04:16:47 pm
I had exactly the same fault with a TDS3032B (Noise superimposed on the signal)
Initially I contacted Tektronix but they wanted over £1000 just to take a look so I decided to pull it apart and see what I could find.
Same pcb as the 54 but only 2 hybrid modules fitted of course.
Carried out the same tests but could find nothing so reassembled the scope.
I then found that by selecting bandwidth limit of 20Mhz or 150Mhz the fault did not show.
Switching back to Full BW and the fault was still there.
However I switched the bandwidth to 20 MHz and cycled the power and then reselected Full BW
Unbelievably the fault had gone :phew:
That was over a year ago and the fault has not come back (I did try some percussive maintenance to see if it was a poor joint etc but could not get the fault to come back).
May be worth a try
NOTE: I cannot remember the exact sequence I carried out but it was something like I describe above but it was definately the Bandwidth setting that I was changing.

Jerry
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: vikpc on January 01, 2014, 04:56:52 pm
Maybe it just relays?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: ResR on January 01, 2014, 06:55:58 pm
Have you checked the passive components? Maybe some resistor is burned out or capacitor cracked and doesn't filter the noise any longer on that channel?? Sometimes a transistor that is starting to fail gives out noise with the signal also.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Alexei.Polkhanov on January 01, 2014, 07:14:46 pm
My guess would be that you have one bit flipped on the digital side. This can be at the parallel communication somewhere, and the cause for this could be ground bouncing. I would check the decoupling capacitors if any of those are cracked, or if you have significant noise on the power supply right at the ASIC pins.
I would +1 to that. Signal on screen does show signs of what you have described.

Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: atw60444 on January 01, 2014, 07:17:22 pm
Did Dave say he'd uploaded #565, part two? No sign of it here yet :(
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Jonny on January 01, 2014, 07:38:12 pm
Did Dave say he'd uploaded #565, part two? No sign of it here yet :(

There, There.... he only said he shot it. I'm sure we'll see it soon enough  ^-^
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Rufus on January 01, 2014, 07:58:45 pm
Well, if google has anything to say:

http://www.seekchip.com/icstock-m/MM9576-VJG.html (http://www.seekchip.com/icstock-m/MM9576-VJG.html)

The MM numbers are date or batch codes the part numbers are ADG.

The ADG365 is mentioned in a Tektronix patent about some video triggering scheme. It is described as a trigger circuit controlling a sampler.

Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: vikpc on January 01, 2014, 08:00:24 pm
im think need swap hybrid modules betwen channels to check where is problem.. 8) Dave have cool desoldering station.. its about 10-20min to swap
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: georges80 on January 01, 2014, 08:42:44 pm
Well, if google has anything to say:

http://www.seekchip.com/icstock-m/MM9576-VJG.html (http://www.seekchip.com/icstock-m/MM9576-VJG.html)

The MM numbers are date or batch codes the part numbers are ADG.

The ADG365 is mentioned in a Tektronix patent about some video triggering scheme. It is described as a trigger circuit controlling a sampler.

Given the parts are NS and MM is an NS part number 'starter', I'd suggest those are NS part numbers. The ADGxxx may be a branding for Tek. MMxxxx doesn't look anything like a typical batch/date code - the stuff below the MMxxxx is more likely that info.

cheers,
george.

Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: sync on January 01, 2014, 08:56:02 pm
im think need swap hybrid modules betwen channels to check where is problem.. 8) Dave have cool desoldering station.. its about 10-20min to swap
... and risk to destroy another channel.
And probably you should not touch the hybrids which makes swapping a bit difficult.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: PA0PBZ on January 01, 2014, 09:50:09 pm
Did Dave say he'd uploaded #565, part two? No sign of it here yet :(

There, There.... he only said he shot it. I'm sure we'll see it soon enough  ^-^

There: https://www.eevblog.com/forum/blog/eevblog-565-tektronix-tds3054-oscilloscope-repair-part-2 (https://www.eevblog.com/forum/blog/eevblog-565-tektronix-tds3054-oscilloscope-repair-part-2)
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: jancumps on January 01, 2014, 10:42:26 pm
I find it frustrating to watch your fault finding/repair videos.
Initial mechanical inspection and some speculation is OK to a certain point, but don't go overboard.
I don't kow why you avoid using electronic test equipment to do at least some basic measurements?
You have a multi channel scope, you could easily do some signal comparisons on the front end.
At the end of the day it may not resolve the issue, but for **** sake you are running en electronics blog/forum.
Yeah... as a software engineer, the random poking around kind of makes me cringe. The term we use for that is shotgun debugging. My approach would be to use bisection to find the problem area. In this instance, you've guessed at what some of the traces on the boards do, let's probe them with a scope at various points to see whether the fault exists before oYr after that point. Why waste time with random wild shot guesses.

On the other hand, since Dave is commenting on the various bits and pieces he examines as part of the random wild shot guesses (Dave: hey, at least they appear that way in the video. If you've done a more methodical problem analysis, it's not in the video. If I've stepped on your toes here, I'm sorry :) ), I find the video interesting to watch. There's still so much I don't know, and what might be an off-hand side comment to you is really an information nugget to me.

I'm also from software side and I disagree ;)

There is a substantial difference. Hardware has a much bigger physical component in it. Software debugging is almost always about logic that's  failing.
There is nothing wrong with (seemingly) random poking at and looking at parts of the scope - to see if something looks bad - before digging deeper into it.

chances are quite high that a failure is visible or smellable.

When that faiils it is the right time to get into a more structured approach like described above.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Saneoc on January 01, 2014, 10:55:03 pm
Hi all!
First of all, I'm sorry for my english, but I want to say something.
  The strange thing I notticed is the fact that the noise is always above the trace at the positive half cycle and mostly below the trace at the negative one (or near zero).
Probably this can be caused by some adc logic error, but I don't know how to test it.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: jancumps on January 01, 2014, 10:57:02 pm
In the follow up video, noise is on both sides with a sinus 1KHz test signal.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Saneoc on January 01, 2014, 11:07:19 pm
I saw it, but there the clamping is also present.
  You may notice that at some point the polarity of added signal is changing.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: vaualbus on January 01, 2014, 11:24:40 pm
I will not consider noise to be a problem, it could be due to a lot of factors. Remember that the scope not working in their perfect condition so it could be normal.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: EEVblog on January 01, 2014, 11:41:14 pm
Did Dave say he'd uploaded #565, part two? No sign of it here yet :(

There, There.... he only said he shot it. I'm sure we'll see it soon enough  ^-^

It usually takes 12-24 hours or overnight for the video to upload, process, and then monetise before I can release it. And that's after the editing.
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: KarolNi on January 02, 2014, 10:52:42 pm
I've watched 2nd part, but I spotted something in 2:39 in this part. Noise seemed to be periodic - I think it might be some coupling and heterodyning with other periodic signal - clock or dc converter ripple, maybe on digital side. It will not be handy, but maybe some off-circiut check of decoupling capacitors?
For 2nd video I'm curious about negative voltage on broken diode vs. positive on operating channels - but this wouldn't be issue IMHO. Dave, have you checked voltage on pads after desoldering diode?
Last thing - maybe some phone procedure (factory reset, reflash with newest firmware, if available and recalibration) will work?

Good luck!
Karol

Edit: I mean heterodyning sample clock with ripple or other clock.
Karol
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: burra7 on January 04, 2014, 10:11:48 pm
DC ripple sounds interesting. I tried to have a look and the ripple seems like 50kHz. Might be an idea to measuere the DC rails +5V1, -5V1, +15V and -15V to see if we can find the ripple there also.

Have made a block shema also:
https://bitbucket.org/burra/tds-3054/src
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: Wim_L on January 05, 2014, 06:16:30 pm
A close look at the video in dots mode. 6:21

It seems the noise is quantised, a group of dots gathering, but in discrete steps, not as a smooth noise band. Perhaps something to investigate? A certain subset of the 9 bits failing or going random?
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: M0BSW on January 05, 2014, 08:15:37 pm
And we have no forum members at Tek who can dust off somebody to give their thoughts?
Actually there is Alan W2AEW  a Tek employee, Very helpful fella , I'm sure you'd just have to ask him for his thoughts and He'd help
Title: Re: EEVblog #564 - Tektronix TDS3054 Oscilloscope Teardown Repair
Post by: R_G_B_ on March 16, 2018, 12:59:27 pm
I am to having the noise problem on high sampling rates

E.g noise when the time base is below 400us.

I narrowed this down a bit as follows.

10k points the noise appears below 400us @ 5Meg samples.

At 500 points sampling this happens below 10us at 5Meg samples

Because of the decimation in sampling points there does not seem to be as much noise in the display but it is there and it seems to be low  frequency ripple And Wonder of the signal. Could this be a fault with the triggering circuit. Open circuit resistor seems that the error message complaint is about the termination resistor being to hot. 4KTBR watts Boltzmann's constant having anything to do with the noise at higher sampling rates with intermittent or resistance ?

Just some thoughts