I will tear my credentials up!
All of them!!!
Having dismantled the DS1052E in order to figure it out, I connected the PSU, the LCD and the Keypad PCBs to the mainboard to take some live readings, in this non-trivial task of repairing this sort of a
black box. Well, I realise that this kind of repair is probably next to impossible, due to the lack of any schematics or any kind of service documentation. But not everyone is an easy quitter, especially not someone who loves to play with his toys!
I dismantled the problematic DS1052E to trace down the source of the instrument's excessive noise. After having taken some readings, I decided to power it up for further investigation of its noise issue.
Well, here is something not so commonly encountered, even in the Internet Era:
Noise levels of the DS1052E dismantled!
Please, compare the noise levels of the device in this state, to the original noise levels posted at the first message of this thread:
Noise 5: Ch1 probe on the probe calibration output.
It is obvious that the noise level of the dismantled DS1052E is less than half in amplitude of the properly assembled unit noise level!
This can be partially explained, since the noise generated by the
switching PSU is further away from the ADC and the Analog sections of the oscilloscope's mainboard, and in a different orientation (polarization).
Operating DS1052E dismantled!
In the picture above, it can be seen that the oscilloscope operates outside its metal shielding (its Faraday Cage), in front of a non-trivial in EMI emission terms LCD monitor, and it has its Analog Front-End wide open(!). It can also be seen the current shunt resistors directly connected to the PSU output, as well as the non-EMI protected workspace.
I think it is time for me to have a drink, to clear my mind up!
-George