Author Topic: [WTB] Fluke 8010a / 8012a for parts  (Read 727 times)

0 Members and 1 Guest are viewing this topic.

Offline HendoNZTopic starter

  • Contributor
  • Posts: 44
  • Country: nz
[WTB] Fluke 8010a / 8012a for parts
« on: April 15, 2021, 07:27:47 pm »
Hi all,

I recently got a Fluke 8010a, but unfortunately the AC side reads overrange, no matter the input. I did some hunting around the unit and found there's a problem with the U8 board.

So now I'm on the lookout for a replacement board. So if anyone out there has a parts unit  that they are will to let go of - I would be very grateful.

Cheers
 

Offline strawberry

  • Super Contributor
  • ***
  • Posts: 1155
  • Country: lv
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #1 on: April 16, 2021, 07:05:01 pm »
https://www.eevblog.com/forum/repair/fluke-8050a-true-rms-u32/
could be -V supply switch not powering U8
carbon composite resistors might be bad
 

Offline HendoNZTopic starter

  • Contributor
  • Posts: 44
  • Country: nz
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #2 on: April 17, 2021, 09:01:54 am »
Thanks for the reply. U8 seems to be getting the power it needs. But it's hard to see where the problem is occurring.

Still looking for a doner board. Cheers
 

Offline strawberry

  • Super Contributor
  • ***
  • Posts: 1155
  • Country: lv
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #3 on: April 17, 2021, 09:05:46 pm »
AD636 or similar
 

Offline HendoNZTopic starter

  • Contributor
  • Posts: 44
  • Country: nz
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #4 on: April 18, 2021, 12:42:06 am »
Cheers for the response. Are you suggesting that to replace the CA3046?

I think the best solution is to replace the whole board - I don't know where the issue is occurring except for it have to do with the U8 board. Cheers
 

Offline strawberry

  • Super Contributor
  • ***
  • Posts: 1155
  • Country: lv
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #5 on: April 18, 2021, 02:16:20 pm »
easy to test if 5 bjts in CA3046
maybe output buffer OP output blown short?

by the way made in EU AS3046
https://www.alfarzpp.lv/eng/sc/transistors.php
 

Offline HendoNZTopic starter

  • Contributor
  • Posts: 44
  • Country: nz
Re: [WTB] Fluke 8010a / 8012a for parts
« Reply #6 on: April 18, 2021, 08:33:14 pm »
Thanks for the info.

I did a quick test this morning (haven't done the whole chip) but looks like pin 4 (B) is open to pin 5 (C) using a diode test and reads 1.5ohm.


Looks like I was wrong - it's designed to have a short between 5 and 4. I checked the rest of the chip and all is okay. One this I have seen - pin 22 on U8 is reading 10.6v after a zener diode of 4.3v - pin 1 on U8 gets 14v from VA then is split to pin 22.

Is that drop enough or is the zener playing up?