Author Topic: FS UK: Agilent 3458A 8.5 digit DMM  (Read 5535 times)

0 Members and 1 Guest are viewing this topic.

Offline factory

  • Super Contributor
  • ***
  • Posts: 2884
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #25 on: October 27, 2023, 03:25:25 pm »
The company work outsourced calibration to still puts cal-due labels on failed items, even though it also states failed.

If I reject something it gets a cal date but where the due date is I put rejected. But I know from a 17025 (UKAS) that they feel it should be for the customer to decide so I have to use a "Limited Calibration see Certificate" label. But they also get funny about us putting due dates on things as it is for the customer to decide, the 12-month thing is a really lazy option and was mostly promoted by companies that wanted you to send the stuff back to them each year. If a customer doesn't state a frequency we just leave it blank.

In the past we had several people employed full time doing internal cal (with very little done externally) this got slowly reduced, by sending stuff to various external different external cal companies, with less done in house.
More recently those in charge want rid of the internal cal lab, they choose one company to send everything to, with a fixed yearly contract price (repairs not included), not sure if they choose based on, what was easiest for them, or the cheapest option.
They seem to be giving the same cal intervals, six or twelve months as in house did. When you say "customer" this will mean whoever is in charge will decide if they even notice.
It'll probably fall apart when a auditor finds one of the barely noticeable "fail"s on a cal label, some of us know what to look for & put out of use, but not all will.

David
 

Offline mendip_discovery

  • Frequent Contributor
  • **
  • Posts: 855
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #26 on: October 27, 2023, 05:43:21 pm »
It sold, for £3275

David, lots of companies have gone for this method, as they want to get rid of the hassle of dealing with calibration. The thing for me is very few of them ever look at a certificate and ask is this fit for purpose and if this item meets the requirements for being used? This I feel in a sensible sized company could almost be a full time job.
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
--
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.
 

Offline pope

  • Frequent Contributor
  • **
  • Posts: 297
  • Country: pl
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #27 on: October 27, 2023, 08:31:07 pm »
It sold, for £3275

David, lots of companies have gone for this method, as they want to get rid of the hassle of dealing with calibration. The thing for me is very few of them ever look at a certificate and ask is this fit for purpose and if this item meets the requirements for being used? This I feel in a sensible sized company could almost be a full time job.
To me it shows that it sold for £2,325
 

Offline chilternview

  • Regular Contributor
  • *
  • Posts: 146
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #28 on: October 27, 2023, 09:08:49 pm »
To me it shows that it sold for £2,325

 

Offline factory

  • Super Contributor
  • ***
  • Posts: 2884
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #29 on: October 27, 2023, 09:47:12 pm »
£2325 + VAT + auction fees was the one from PP auction, not ePay.

David
 

Online bastl_r

  • Regular Contributor
  • *
  • Posts: 92
  • Country: de
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #30 on: October 28, 2023, 01:00:24 am »
Apart from the fact that the device has become too expensive for me, no matter how good it still is.
How would it compare to a 7.5 digit meter. My feeling is that with the possible errors discussed, it should still be at least half a decade more accurate than a 7.5 digit multimeter.
Or?
 

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 2306
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #31 on: October 28, 2023, 08:27:16 am »
An out of spec 3458A 8.5 digit meter could be as accurate as a 7.5 digit DMM, but you would still need the ~£2K Keysight cal to show show how out of spec it was ;D
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #32 on: October 28, 2023, 10:31:37 am »
We dont hear a lot of the ADCMT  Transmill  or the Fluke 8.5 digits or any 8.5 digits meters ...

If they have some quircks like the 3458 ???

EDIT

Advantest R6581 ??
« Last Edit: October 30, 2023, 04:36:04 pm by coromonadalix »
 

Offline Oldtestgear

  • Regular Contributor
  • *
  • Posts: 144
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #33 on: October 30, 2023, 11:27:26 am »
Or the Solartron 7081 *.5 digit DMM. Although there are a few threads about these on the forum.  Nice clear display but definitely a bit quirky.


Phil
 

Offline MadTux

  • Frequent Contributor
  • **
  • Posts: 785
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #34 on: October 30, 2023, 02:40:02 pm »
We dont hear a lot of the ADCMT  Transmill  or the Fluke 8.5 digits or any 8.5 digits meters ...

If they have some quircks like the 3458 ???

If ADC is build without hybrids like U180, at least it's most likely more DIY fixable.
With HP3458A, you totally depend on Keysight for support, something I'd rather avoid.
 

Online alm

  • Super Contributor
  • ***
  • Posts: 2897
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #35 on: October 30, 2023, 04:12:46 pm »
If ADC is build without hybrids like U180, at least it's most likely more DIY fixable.
With HP3458A, you totally depend on Keysight for support, something I'd rather avoid.
Are schematics available for the Fluke or Transmille meters like they are for the 3458A?

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #36 on: October 30, 2023, 04:32:32 pm »
dont think so,  never heard of leaked ones,   you have some   but no schematics for sure  loll

Fluke 8xxx               never saw
Advantest r6581     never saw         https://xdevs.com/fix/r6581t/
Transmille 8081       never saw

the 3458A            is known  loll

K2002                  cant remember ...  don't think so,   k2000 was partially
Solartron 7081    i think it is known,  some  float on Archive.org
« Last Edit: October 30, 2023, 04:43:56 pm by coromonadalix »
 

Offline Kleinstein

  • Super Contributor
  • ***
  • Posts: 14262
  • Country: de
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #37 on: October 30, 2023, 04:50:20 pm »
The Solartron 7081 eventually gets 8 digits, but with less than 1/10 the speed and a little residual drift and somewhat troubled auto zero to correct for this.
AFAIK there are no schematics avialable for the new fluke and the transmile meters.  For the slightly older (AFAIK not current) transmile meter there is a thread here in the forum that has some testing and reverse engeniering one the ADC. From that it looks a bit like the Solartron with added oddities. Without a schematics and service hints this could be hard to fix / test and a programmed µC or CPLD can already be a show stopper if bad.

There are full service docs on the Datron1281, the predecessor to the high end Fluke meters.
For the Solartron 8081 there is a plan around.
For the K2002 there is partial rev. engeniering (ADC, ohms part and crude voltage part) and a useful service manual to at least explain the self test errors.
For the R6581 there are quite good rev. eng. plans and even firmware analysis around here in the forum.
Still not all 8 digit meters are at the same level.
 

Offline alligatorblues

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #38 on: December 30, 2023, 03:32:58 pm »
I think somebody on this forum said (half-joking) the only 3458A's on ebay are the drifty ones no one wants, so buyer beware.

Semiconductor manufacturers use many 3458As. When they get drifty, not even out of spec, they turn them over to liquidators, and some find their way to eBay. The A3 boards come up on eBay a few times a year. You just need to save the search, and when one comes up, you'll get an email.

You just never know. Last time I needed one, a soldier on a military base had 25 of them. He probably found a box of them and knew what they are. If you go on military calibration forums, those guys can get anything you want. But you have to get to know who's who.

Defense contractors also have cal labs everywhere. So, their techs are there too. Lockheed Martin does calibration work for commercial labs, but it's very hush hush how to find the labs that are open to the outside world. I found one in FL USA. But it only does pick up and drop off.   
 

Offline Oldtestgear

  • Regular Contributor
  • *
  • Posts: 144
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #39 on: December 31, 2023, 12:02:41 pm »
Quote
An out of spec 3458A 8.5 digit meter could be as accurate as a 7.5 digit DMM, but you would still need the ~£2K Keysight cal to show show how out of spec it was ;D

I just paid for Keysight UK to calibrate my 3458A, at the bargain price of "only" £716.40. Note no adjustments performed.
Getting a 34470A calibrated cost £247.20 which included pre & post adjustment results.  The Instrument was within specification before adjustment but this was performed as part of the job.

All prices included VAT
 

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 2306
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #40 on: December 31, 2023, 12:24:10 pm »
Oh, bargain!

You state prices including VAT but they wouldn't deal with you as a private individual, have I got that right?

I have "given in" to Keysight's demand that I incorporate a company as the only way to get my 34470A calibrated.  It is definitely not a sensible option but it is pragmatic. I have completed the paperwork & notified Keysight. They have acknowledged my email & stated that full support is now available.  I hope to get the 34470A calibrated next week, with one of my newly acqured 3458As planned for late October assuming that both pass the dreaded long term drift test.

Not sure if I can help anyone else with parts or calibrations but am happy to try.  My new company is not intended to be profitable butcannot be run at a loss either.

Excluding my shipping costs to/from your office, how much would you charge to get my 34470A cal'd at Keysight, with adjustments?
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #41 on: December 31, 2023, 02:16:36 pm »
there's a difference between certification AND calibration
 

Offline voltsandjoltsTopic starter

  • Supporter
  • ****
  • Posts: 2306
  • Country: gb
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #42 on: December 31, 2023, 02:55:27 pm »
I have a plethora of certificates of calibration.
I don't have any certificates of certification.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #43 on: December 31, 2023, 03:04:58 pm »
in my country, 

Certify the instrument work in the acceptable range or outside of the acceptable range,  no calbration

Calibration,  it was certified before, got the results, and calibrate it,  adjustment done internally or software made  etc ... 
a calibration can be done,  AND  it can get a failed status on some ranges, failed on 1 range  or failed on all   ....   because it still can not get under the acceptable ranges

normally the certification will tell all,  and the user have to choose if he / she goes thru the Calibration

mostly the Cal labs :   you can not calibrate until a certification was made prior, everything is logged ... that's how they work in my town
 

Online alm

  • Super Contributor
  • ***
  • Posts: 2897
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #44 on: December 31, 2023, 04:20:19 pm »
in my country,
What country is that, that is not a member of the BIPM, and does not adhere to the International Vocabulary of Metrology (see section 5.18 / page 39)?

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #45 on: December 31, 2023, 04:58:53 pm »
https://www.bipm.org/en/countries/ca

and you refer on section 3


i'm not the one who use theses terms,  but there is differences, i simply wrote what happen in some of our cal labs  other i dont know

I ship  over 10K $$ of stuff at a 1 or 2 year intervals  and they clearly state, is it a certification or a calibration ?

A certification can be issued without a calibration


I repeat

If after the certification you want to do a calibration, you can  it's your choice if it goes well or not, because there may or was possible problems,  they can or could be resolved by a calibration,  if  say:  you have a failed range,  the calibration get an incomplete status and it is not considered fully calibrated therefore  it still get some failed status in the paperwork, it as i said  it will be recorded

After some repairs if it's possible, you can retry a calibration ... or not,   bit it was already logged as having one problem


I'm not talking about Keysight labs, there are others in Canada, Quebec city, Montreal, Ontario   etc ...

Some of my Gossens meter had to be shipped very far to be kept under the DAKKS system, some of my 34401 at the time (one was fully calibrated)  the others where certified, and theses old 20 years pup where still in the ranges ...

some scopes where calibrated,  others not ... old tds220,  tds 2k series

Even some Cal labs have problems to deal with  old thanks like TDS7254B and a TDS7404, have to ship them far away too'

now shipping lots of brand new Siglent stuff  we bough, and they will be "calibrated" because they work in critical stuff

even i  i'm lost at word,   it's some basic examples ... for sure it is contestable if some wants to

even some crappy (but working well) 3x bought juntek jds2900 at 100$ usd  had been certified for 160CAD x3 ??    witch is a pure joke, i've done some battle with the certification guy, he simply dont care,  i said its not worth it

 

Online alm

  • Super Contributor
  • ***
  • Posts: 2897
  • Country: 00
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #46 on: December 31, 2023, 05:07:08 pm »
i'm not the one who use theses terms,  but there is differences, i simply wrote what happen in some of our cal labs  other i dont know

I ship  over 10K $$ of stuff at a 1 or 2 year intervals  and they clearly state, is it a certification or a calibration ?
Please do not repeat such misleading usage. It's internationally agreed (for example in the document I referred to) that calibration is a documented measurement or comparison, and adjustment of an instrument is called adjustment. You can ask for calibration with or without adjustment. For some instruments asking for calibration may also get you adjustment. See for example this video:

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5950
  • Country: ca
Re: FS UK: Agilent 3458A 8.5 digit DMM
« Reply #47 on: December 31, 2023, 05:41:18 pm »
tell it to  the cal labs we use   ... they are the ones who scrap the good terminology / definitions etc ...  and yes  i can do bad telling it here   

if they are wrong from the start   

sadly   some and  "i include myself", dont talk like this

here calibration is not a certification for sure,  and calibration as the same usage as adjustment (the moment you change a reference value into an instrument)
even got a big red / white sticker once on a psu  "No Calibration needed" and i could still use it

and some other Canadians can or could disagree too

but i'm not here to start a war of words,   if i stepped on a land mine  thats all   boom   loll


well return to the main topics as a mod would kindly ask



« Last Edit: December 31, 2023, 05:51:37 pm by coromonadalix »
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf