But I'd like to have something that does say 1mm field of view for general silicon dies and other objects for teardowns etc.
Speaking of which, FOV calculations for the Amscope one are:
The WF16X eyepieces are 11mm FOV
For the objectives supplied:
x5: FOV = 11mm / 5 = 2.2mm
x10 FOV = 11mm / 10 = 1.1mm
x50: FOV = 11mm / 50 = 220um
I'm going to assume the x50 is not going to be that good an optical quality (and likely extreme working distance limitations, i.e. practically touching, so flat objects only). So the x10 objective should be quite usable for starters.
And I can always buy a top quality x20 or even x50 objective.
Although Amscope claim the objectives are plan objectives, so should be flat across most the visible area and gives the best working distance. How much remains to be seen.
But that's optical view, not sure what that translates to into the trinocular camera port?
As for what the camera sees, these aren't infinity-corrected objectives*, so it's quite straightforward (once you've spent a while studying all this :-)): the eyepiece is not involved at all, the objective projects its image directly (well, via some mirrors/prisms) on the camera sensor, but 5, 10 or 50 times bigger than real life. The field of view of the supplied camera is 5.70mm(H) x 4.28mm(V), but there's a 0.5X reduction lens on it which makes it effectively 11.4mm x 8.56mm; divide that by 5x, 10x, or 50x to see the field of view that you'd get. Since your objective has a field of view of ~11mm, and the camera field of view is ~11.4mm wide, the answer is, as you would hope, almost exactly the same as the calculations you gave above. The pixel size is 3.2um, which is actually 6.4um if you factor in the reduction lens, again, divide by 5, 10 or 50 to see the "real" pixel size (remembering that each pixel is either R, G, or B, so your true resolution is considerably worse than this.)
* With infinity-corrected objectives, the image is formed by a lens in the tube; the magnification factors given by an infinity-corrected objective is stated under the
assumption that the tube lens has a 200mm focal length. If it's not, you have to scale the objective magnification accordingly, a x10 infinity objective with a 100mm tube lens is really only x5.