Hello people, thanks for your interest on my problem ..
I have just finish the battle with my demons , and I had won ..

The solution was not simple ...
About the part that has to do with 87-5 :The named as issue was at the "High Resolution" mode .
The DMM at the 600K range measure my 100K 0.1% resistor as 100K ,
by entering at the "High Resolution" mode , the resistor was measuring 99.95 .
This deference made me think about an possible issue with the meter.
Thanks God , after the 87-5 I got and the bench meter Fluke 8050A 20.000 counts,
that it did confirm the 99.95 K reading.
Something that helped my confusion to grow, was that at the "High Resolution" mode the 87-5 , it needed more than 12 seconds to settle , it started measuring the resistor as 100K and then it started slowly to drop like 99.99 99.98 99.97 99.96 some play between 99.96 & 99.95 so to stay there.
The 8050A was spot on from the first second . 99.60 thats it .. this is the measurement .
No play , no waiting .. spot on at fast.
About the part that has to do with 9 positions switch I had bust my head for three days dealing with all this complex issue ,
and come to the point to totally dismantle it ...
Found inside excessive amount of Flux , so much that all the contacts was having an bath in it.
I used solder with Silver , it was hard to melt , and I was using lots of flux so to help it .
The switch are almost sealed , so the fumes of the flux was unable to escape , and they was remaining trapped in it.
When I was heating the external contacts so to solder or de-solder , the flux inside was becoming wet again, causing all sort of problems in my measurements . ( leakage )
I took the specific apart , wash anything inside with alcohol , in depth cleaning, and then I assembled it again.
About the part that has to do with the resistors I trusted the crop of the cream " Dale " 100K 0.1% , but even so, not all resistors was correct .
I had to check them , one by one .
I did not do that at the beginning , because of blind trust !! I payed an large price for them,
so to not need to test them ... Testing garbage its logical ... testing the top dog ? ( looked as illogical to me )
Any way since the time , that I started to isolate and remove my problems above,
DMM readings & wet switch full of Flux.....
I started to calibrate the last stage of my decade the 100K range,
got many film metal resistors from 10 Ohm to 85 Ohm ,
and started to adjust the six from the nine resistors , until to see the 100.01 K in all .
Now I am dead tired but an clear winner .
The 8050A helped me allot , with out it , with out the second confirmation , I would become nuts.
Any way, the decade are ready , and the other stages like 10K and lower , they did not effected by the extra Flux in them.
I know that they are full of it , like the last one.
Few words about the Auto-range , the Fluke 87-5 its my first Auto-range meter ,
I have to admit , that its a very fast meter .
@rf-loop had speak about some possible Auto-range issues,
well I think that the 87-5 its so damn fast , that its unable to cause damage,
due the speed that it haves.
An very slow DMM , It could be called as danger for sensitive components, but not the 87-5.
The happy end, about my seven stages decade, was the interconnections lowest resistance.
That is at 00,04 ohms . ( 20 ohms range)
I did verify that number , by the 87-5 , and the 8012A bench meter that I had fall in love with it,
because it has the "Low Range Ohm " extension that found very handy in this case,
and the most amazing part , are that the 87-5 it did agree with the ancient 8012A, at 100% .
So all my meters are finally agree , that I had succeed