yes, thanks, if money was no issue the 6622A xr would be a fine piece of gear. my guess is this is arround 50k though. that does not make sense.
the 242d was actually what I was trying to emulate by using the 720a and the 721 lead compensator, and given the fact that the 720a is self-cal (the 242d resistor decade is more complicated to calibrate, also it has quite a high temp drift of several ppm/K, and I see fluctuations of 1-2 K during a measurement cycle), I would think it should not be worse in resistance comparison. at least on the paper the 720a solution looks actually better.
the nullmeter can be adjusted to better than 1uV I would think, and at 10v or 100v total (for the 10megs measurement) with 0,1ppm referenced to input should give some reasonable results. i have not done the math yet, will do soon, and am waiting for the 721 to arrive, so doing tests to see how the noise is doing will be important too. I thought that you might have had some results here overall.