800mW is class IV (in the old scheme, I have not tried to keep up) so the direct, specular AND **DIFFUSE** reflections are considered dangerous to vision, while still clearly not being powerful enough to be seriously useful.
A safety shutter that is removable and not interlocked is not a safety shutter.
Does this thing have an FDA approval number for the US market (Ask not how lasers ended up under the FDA over there)? While I think some of the CDRH requirements can sometimes be a little over the top, the 5 second delay, emission indicator and mechanical shutter are all things I can totally support.
As to focal length providing safety, lets do some maths....
The assumed diameter of the pupil is 7mm, so how much area does a ~1W beam have to have spread to to reduce the power incident in a 7mm diameter circle to below 1mW (And that assumes a visible beam)?
7mm diameter is 38.5mm^2, so to get the power density down to 1mW in that area you are looking at 38,500mm^2 = a 111mm radius circle, that says short lenses and a lot of mechanically enforced clearance to me (Someone please check my arithmetic).
Regards, Dan.