Thanks guys - good thoughts.
but are you sure you haven't missed something?
No and I have a nagging doubt that I
am missing something, hence the post.
The problem affects both X and Y deflection so it is likely to be something "global".
Back to basic checks? PSU?
Can't you measure the anode EHT? If it's too high maybe that could be the problem?
The PSU low voltage rails all check out - the 10V ref might be microscopically low at 9.99V (+/- 5mV spec - another 5mV would be within the DMM inaccuracy).
Yes, too high an EHT voltage was one of my thoughts. I haven't measured the final anode voltage directly - I do have a 40kV probe but it would be impossible to get it to the CRT connector with the tube in situ. There is a spec for a dummy load in the service manual but I don't have a 14kV rated 290Mohm resistor handy
As a surrogate for measuring the EHT I checked the cathode -2200V supply which is OK at -2230 or so (well within the 2% accuracy of the HV probe). Both supplies come off the same winding of the EHT transformer and the regulation is feedback from the -2200V rail anyway.
Intensity, focus and astigmatism controls adjust as expected so, although I haven't measured all the tube voltages I'm not currently suspicious that they are out of spec.
DC value checks throughout the schematic?
Will the position pots allow full trace movement?
* Below-spec voltages on the deflector plates,
DC checks throughout the schematic whould be a tall order and I'm not sure how much light it would shed. Strictly there aren't any pots - it's all DAC controlled but there are good ranges of trace movement.
I
have checked the deflection plate voltages - the final X amplifiers (two in antiphase) give a swing of between 80 and 96V p-p which is the same as a 3394A which works and in that 'scope about 88V of swing is needed.
I have checked that a signal is affected and that this is not just the vector generator in the 'scope and it is - I can't quite stretch a 10Vp-p 1kHz square wave to read correctly in either X or Y. I haven't checked the timebase frequency but I don't think that's it as the vector generator is entirely voltage driven and anyway that would not explain Y axis problems and the cursor system measures the signal perfectly.
* Wrong shape of post-deflection plates electric field.
Another idea: some tubes seem to have a kind of electrostatic lens
Mechanical damage isn't out of the question but to produce a perfectly symmetrical problem seems odd. The tube has some internal magnetic elements to help geometry.
The tube hasn't gone 'gassy' has it? i.e. partial loss of vacuum. This can happen with a failure around the glass envelope on CRT pins.
I don't believe so. EDIT: At least I don't believe the 1st CRT (the one with funny deflection) is gassy, the dim one with low cathode voltage on the other hand is a different kettle of bananas.
The plan was to replace the CRT as a diagnostic test as this has become a two 'scopes into one kind of exercise. Unfortunately the EHT connector on the 2nd CRT was not the same as the first so I needed to bring the multiplier over with it which in turn meant bringing the whole PSU.
That did "fix" the deflection problem but at present that is not terribly useful. The 2nd PSU is making "I am not a happy PSU" noises and the cathode supply is definitely too low by 10-15% (about -1900 - 1980V) so the final anode voltage will be too low as well. This is likely to make this CRT too sensitive to deflection voltages.
I'm going to swap the EHT multiplier this evening and see where that gets me. If the voltages are OK but I still have the deflection problem with the replacement CRT I will know to look elsewhere, if the voltage is still low then I will know that the 2nd CRT or EHT multiplier have problems (though that won't help much with the original problem).