New HW Hantek DSO5102B
(test started with selfcal what have done totally in different temp becouse machine burn in tests are going on. This do not mean anything becouse this test is only for look change)
Before start scope was more around 3 hours powr off and roon temp around 22 C whole test time. Scope is on the table alone so that outside temp do not change markable.
Vertical settings: DC, 1x, 2mV/div,
20MHz limit off, Average 64, realtime.
50 ohm terminator in input.
Look only change, not absolute value.
Time min; display (mean) mV
0;2.24
1;2.32
2;2.24
7;1.68
10;1.36
15;1.12
20;1.04
25;0.96
30;0.88
35;0.80
55;0.72
120;0.64 - 0.72
But it must also think that it is nearly only vertical shift. Not gain error what is much more severe problem if it drift.
partially this is becouse big part of cooling is working as free convection trough outside case air holes. So inside case it takes long time it goes slowly higher and higher temp after it reach balance between heating and cooling.
Front end (whole front end box inside main board box) final temp is quite high.
If there is fan it do not continue warming so high and balance is reached more fast.
same time I look how my one analog tektronix drift.
2mV/div it drift total in same time 0,3mV
If there is fan in Hantek I believe that drift is less and it reach maximum drift more fast.
This is price about fanless... read noiselless scope.
yes it can also design much more better for handling thermal things without fan, but, construction is then not so cheap. In production there is lot of other costs than only components and material itself. (it need be so that it need minimum amount of human work hours)
I really do not understand why they design to run it as high temp as it is. maybe thermal design is not so wanted work... all want write code... mechanical and thermal design so that it is also clever in manufacturing needs lot of experience and knowledge...it need real professionals. Example in Tektronix and HP they have done lot of goods and bads and they have learn slowly in tens of years and still it seems be difficult sometimes....
(I think all know Tek 2400 series hybrids extremely critical thermal design... just as most bad what engineer can do. But who care... with name can sell... and who cares... equipments useable time is going more and more short. Years ago minimum is 20years. Now many company try design that machies fails just after warranty time. In some Tektronix if fan stop... scope die after half hour. There is not any watch dog for fan or thermal harakiri. After 5 euro fan stop... you just loose 10000. And this is Tektronix and some before years state of art oscilloscopes.
There are many components around 50 - 70celsius in normal temperature. What if use in 40 celsius enviroment and also if scope is place where air can not freely run so as free convection need. (yes we can do high exhaustion pipe from top of scope and do free air convection flow more powerful... yeah... put 2m high 100mm pipe...