Why does ECG machine cost more than a digital storage oscilloscope?
Is has a good differential amplifier, but than the sample rates are low, and the LCD and processing board are the same or cheaper...
This I have actual, personal experience with. Not saying there isn't price gouging, but to assume that everyone is just trying to rip off sick people is a bit hyperbolic. Also, an ECG is not the same class of instrument as a hearing aid.
1. Because end users aren't meant to buy them (and shouldn't). One shouldn't use them outside of a specially designed environment as the noise floor will be too high for diagnostic purposes.
2. They absolutely cannot break and absolutely cannot get things wrong. In an ICU you want to worry about people, not whether the soldering on op amps was good enough, or a person vomiting on it will short it out. Liability is HUGE.
3. Thought certification was expensive? Try getting certification for a piece of emergency equipment.
4. Cal is expensive (cal people have their own certifications too)
5. Modern processing algorithms are complicated, especially in diagnostic systems (lots of DSP for noise, wavelet, HMM, NN, etc).
6. (sucks, but true) A reliable ECG readout can potentially clear a clinician from malpractice (saves $$$$$$$). See point 2.
I can build an ECG for less than $20, doesn't mean it should be used by a doctor or in a hospital. But really, you can get a solid ECG for about $10k-20k. This is definitely in the price range of a mid-range digital storage scope? The rigol/siglent/etc is hardly a piece of medical diagnostic equipment, I don't think you can really compare them.
Think medical is expensive? If I told the optics guys I know that I wanted to buy a $20k DSO they could probably just give me the money from spare change out of their pockets