Can you send me a raw dump of data for:
Continuity/diode range
2.4 kV displayed when open circuit (meter is showing OL)
0-6.000V jumping around when shorted (meter is showing 0.00)
Note: On my meter the diode/continuity serial update rate is much faster, it seems like 10hz instead of 1hz
however ultraDMM doesn't seem to mind this and just displays data at the fast rate.
When the meter is showing 0.00V (The ultraDMM whole number jumps around by a few volts eg 0.0000 and 3.0000, decimal data (0000) doesn't change)
30-30-30-30-30-31-3A-30-30-0A
30-30-30-30-31-31-3A-30-30-0A
30-30-30-30-32-31-3A-30-30-0A
When the meter is showing OL (UltraDMM 2.4kV)
30-32-34-30-34-31-39-30-30-0A
30-32-34-30-35-31-39-30-30-0A
30-32-34-30-33-31-39-30-30-0A
30-32-34-30-32-31-39-30-30-0A
When the meter has a diode across it with ~0.51 showing (UltraDMM shows ~510V)
30-30-35-30-38-31-38-30-30-0A
30-30-35-30-37-31-38-30-30-0A
Seems to be out by 1000 on diode range as 2.4kV/1000 = 2.4V which would make sense for a diode range.
I guess 2.4V should be shown as OL