2.25k thermistor was added in Firmware Revision 3.00 Enhancements and Fixes. The current firmware is Revision 3.02, link below. Maybe I need to add a #notes Requires Firmware Revision 3.00 or higher.
Thanks for the link, now some of meters are up to date:
;; Keysight Technologies,34461A,MY53214390,A.03.02-02.40-03.02-00.52-01-01
;; Keysight Technologies,34465A,MY57502236,A.03.02-02.40-03.02-00.52-03-01
;; Keysight Technologies,34470A,MY54700281,A.03.02-02.40-03.02-00.52-02-01
But none of them will handle the 2.25kohm thermistor. I found out what the problem was with the null: The meter must have a valid reading when I activate it. Why it reports error in command, instead of something more like invalid input, it not something that can guess.
The 34461 has a couple of problems with temperature, it has much less support for different thermo sensors than the bigger model.
Here is one of the problems:
;; 17:23:28.006 K34461: Tx <TEMP:TRAN:RTD:OCOM?>
;; 17:23:38.006 K34461: Rx Timeout
;; 17:23:38.006 K34461: Rx as number <NaN>
;; 17:23:38.006 K34461: Tx <TEMP:TRAN:THER:POW:LIM:STAT?>
;; 17:23:48.006 K34461: Rx Timeout