Hi Martin. Will Siglent be able to decode an IR remote control signal?
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #576 on: March 01, 2020, 05:54:41 pm »
Hi,

No, whether recs 80 or rc-5 or hitachi code.
Hm... if this can do it in the future (firmware) it would be unique….

Offline Svrbinek

  • Contributor
  • Posts: 46
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #577 on: March 01, 2020, 06:06:26 pm »
Just a little wonder if SDS1104X E would not be enough, because it is significantly different. I know it has only 1GS / s and memory about 14Mb. Now I have Agilent 1GS / s but memory only 4Kb. Rather, I have no idea what the advantage of 50 Ohm inputs. It just eroded me if it would be sufficient for me cheaper 4 channel.
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #578 on: March 01, 2020, 06:18:04 pm »
As always, it´s depending from what you want to do with it.
I´ve owned a siglent sds1104X-E and was pleased about it - Maybe it got all you need, then go for this and save a lot of money.

Offline Svrbinek

  • Contributor
  • Posts: 46
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #579 on: March 01, 2020, 06:30:38 pm »
Can you briefly only basic points as the difference between them? I mean, besides memory and better sampling, what else would the 2000 model have or better? Like mathematical functions or other things?
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #580 on: March 01, 2020, 07:20:05 pm »
Hi,

Bigger screen (10" instead 7"), zone trigger, autosense probe inputs, 50ohm Inputs, formular editor (math), "real" built in awg, touchscreen, have a look at the two specsheets:

SDS1000X-E

SDS2000X-Plus


Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #581 on: March 01, 2020, 07:25:48 pm »
By the way and just recognized it now….

Test it again and lowered the timebase to 100ms - It decodes 15000 packages and this is the limit, at 200 or 500ms the amount is fixed to 15000.

LOL, the maximum frames are already given in the specs….

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #582 on: March 01, 2020, 07:41:14 pm »
Hi,

Bigger screen (10" instead 7"), zone trigger, autosense probe inputs, 50ohm Inputs, formular editor (math), "real" built in awg, touchscreen, have a look at the two specsheets:
Add mouse and keyboard control, Power Analysis, inbuilt MSO HW, smart fan ? (like SDS5000X), 2 Mpts FFT, RTC, Histograms, DMM mode, vertical zoom..............
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #583 on: March 01, 2020, 07:43:51 pm »
….1ppm timebase accuracy, instead 25ppm...and so on.

Offline Svrbinek

  • Contributor
  • Posts: 46
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #584 on: March 01, 2020, 08:13:55 pm »
So it's clear. Although I won't use everything, I would cry that I don't have it. So no compromises :-) SDS2104X +
 

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #585 on: March 01, 2020, 08:18:02 pm »
Martin, can you confirm 2kX Plus also has a smart fan ?
My SDS5kX fan is at full power at boot but slows to much quieter within a minute of boot.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #586 on: March 01, 2020, 08:22:17 pm »
Just checked it out:
No, it remains on the same (quiet) level.
 
The following users thanked this post: tautech, Michael YYZ

Offline 0xdeadbeef

  • Super Contributor
  • ***
  • Posts: 1576
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #587 on: March 02, 2020, 07:19:22 pm »
For the record: there is a thread about the SDS5000X (total mess though due to corrupt screen shots which make the forum SW hide all following postings on page 1) which indicates that measurement are not performed at all for large memory/capture buffer sizes. I.e. it seems as if even something simple as an "edge counter" measurement is not performed unless you reduce the memory size or zoom in. So it seems that the SDS5000X is not able to perform measurements on its whole deep memory - which would somewhat kill the idea of deep memory for certain applications and would be the final nail in the coffin for me. So I wonder if someone checked this on the SDS2000X+.
Trying is the first step towards failure - Homer J. Simpson
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #588 on: March 02, 2020, 08:52:19 pm »
Maybe I get it wrong, but I took a squarewave ( I borrowed my stb-3 today) of 1Mhz, zoom in it measures the edges, zoom out, it measure it too.
But it couldn´t measure more than 200000 edges.




Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #589 on: March 02, 2020, 09:02:56 pm »
Maybe I get it wrong, but I took a squarewave ( I borrowed my stb-3 today) of 1Mhz, zoom in it measures the edges, zoom out, it measure it too.
Trolls looking for bugs that ain't there !  ::)

Quote
But it couldn´t measure more than 200000 edges.
Yes and it indicates that.  ;)
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #590 on: March 02, 2020, 09:09:21 pm »
Quote
Yes and it indicates that.

Therefore the third pic.  ;)

Offline 0xdeadbeef

  • Super Contributor
  • ***
  • Posts: 1576
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #591 on: March 02, 2020, 10:55:35 pm »
At least it's measuring something but this "saturation" is somewhat alarming. I wonder if other measurements like DC and the average etc. calculated from them are also only performed for a part of the measured data.
Besides: does this counter count both edges or did you configure it that way? With only one edge counted, I would have expected 100k edges with 10ms/div and 1MHz square wave.
Trying is the first step towards failure - Homer J. Simpson
 

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #592 on: March 02, 2020, 10:59:39 pm »
At least it's measuring something but this "saturation" is somewhat alarming. I wonder if other measurements like DC and the average etc. calculated from them are also only performed for a part of the measured data.
Besides: does this counter count both edges or did you configure it that way? With only one edge counted, I would have expected 100k edges with 10ms/div and 1MHz square wave.
And what is the practical use of such a measurement ?  :-//
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Online Martin72

  • Super Contributor
  • ***
  • Posts: 5839
  • Country: de
  • Testfield Technician
Re: Siglent SDS2000X Plus
« Reply #593 on: March 02, 2020, 11:06:58 pm »
Quote
At least it's measuring something but this "saturation" is somewhat alarming

For me it´s not.
In fact, I only want to measure what´s actually on the screen.
Practice example, triangle waveform, it´s offset.
To measure it right, the whole screen was needing.
And only one period.
This scope can handle it, so no worries whatsoever.



Offline nctnico

  • Super Contributor
  • ***
  • Posts: 26906
  • Country: nl
    • NCT Developments
Re: Siglent SDS2000X Plus
« Reply #594 on: March 02, 2020, 11:13:56 pm »
At least it's measuring something but this "saturation" is somewhat alarming. I wonder if other measurements like DC and the average etc. calculated from them are also only performed for a part of the measured data.
Besides: does this counter count both edges or did you configure it that way? With only one edge counted, I would have expected 100k edges with 10ms/div and 1MHz square wave.
And what is the practical use of such a measurement ?  :-//
Think about tracking math where you can see a demodulated PWM waveform. Or as I wrote before: numerical analysis of a long train of pulses to know the min/max and average. A useful purpose is to measure the time a microcontroller is spending inside an interrupt routine measured over a longer period of time.
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #595 on: March 02, 2020, 11:18:19 pm »
At least it's measuring something but this "saturation" is somewhat alarming. I wonder if other measurements like DC and the average etc. calculated from them are also only performed for a part of the measured data.
Besides: does this counter count both edges or did you configure it that way? With only one edge counted, I would have expected 100k edges with 10ms/div and 1MHz square wave.
And what is the practical use of such a measurement ?  :-//
Think about tracking math where you can see a demodulated PWM waveform. Or as I wrote before: numerical analysis of a long train of pulses to know the min/max and average. A useful purpose is to measure the time a microcontroller is spending inside an interrupt routine measured over a longer period of time.
Statistics and Histograms are the more appropriate tools for this.
Performa01 posted screenshots using the these functions earlier in this thread I believe.
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline 0xdeadbeef

  • Super Contributor
  • ***
  • Posts: 1576
  • Country: de
Re: Siglent SDS2000X Plus
« Reply #596 on: March 02, 2020, 11:19:09 pm »
In fact, I only want to measure what´s actually on the screen.
"On the screen" is a bit vague. I'm not talking about offscreen data (as there should be any considerable amount of it if at all). But the measurements have to be done from the data buffer and not from some screen presentation of the data (like former Agilents did). The saturation hints towards a limitation of measurements, i.e. not all the data captured (and displayed on the screen) is used for measurements. This is obviously a different approach to speed up the measurements, but one that can make the deep memory measurements invalid and thus unusable for certain applications.
BTW: did you check this in single trigger or normal trigger mode (not stopped/stopped)?

Statistics and Histograms are the more appropriate tools for this.
But those are also devaluated by limiting the measurements to a certain amount. Is this threshold (200k) at least documented somewhere?
« Last Edit: March 02, 2020, 11:21:12 pm by 0xdeadbeef »
Trying is the first step towards failure - Homer J. Simpson
 

Offline nctnico

  • Super Contributor
  • ***
  • Posts: 26906
  • Country: nl
    • NCT Developments
Re: Siglent SDS2000X Plus
« Reply #597 on: March 02, 2020, 11:28:55 pm »
At least it's measuring something but this "saturation" is somewhat alarming. I wonder if other measurements like DC and the average etc. calculated from them are also only performed for a part of the measured data.
Besides: does this counter count both edges or did you configure it that way? With only one edge counted, I would have expected 100k edges with 10ms/div and 1MHz square wave.
And what is the practical use of such a measurement ?  :-//
Think about tracking math where you can see a demodulated PWM waveform. Or as I wrote before: numerical analysis of a long train of pulses to know the min/max and average. A useful purpose is to measure the time a microcontroller is spending inside an interrupt routine measured over a longer period of time.
Statistics and Histograms are the more appropriate tools for this.
But where do the statistics come from if the actual amount of data that can be analysed is limited? You keep thinking too much in terms of periodic signals and not in an entire waveform belonging to a test case which you want to analyse. In my world periodic signals are very uninteresting. The whole point of deep memory is to be able to analyse entire waveforms.
« Last Edit: March 02, 2020, 11:33:53 pm by nctnico »
There are small lies, big lies and then there is what is on the screen of your oscilloscope.
 
The following users thanked this post: egonotto

Offline tautechTopic starter

  • Super Contributor
  • ***
  • Posts: 28380
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDS2000X Plus
« Reply #598 on: March 02, 2020, 11:51:27 pm »
Statistics and Histograms are the more appropriate tools for this.
But those are also devaluated by limiting the measurements to a certain amount. Is this threshold (200k) at least documented somewhere?
Have no idea until I get one and I'm not about to investigate it with a 5kX as it may well be different.
Only Martin and other 2kX Plus owners can investigate the perceived limits to simple measurements whereas Statistics and Histograms are live measurements only constrained by the amount of time they are left to accumulate data and show it in the format they do.

If 'counts' are your thing maybe a 10+ digit counter would be a better instrument instead of robbing valuable DSO display real estate with 6+ digit numbers.  :-//
Avid Rabid Hobbyist
Siglent Youtube channel: https://www.youtube.com/@SiglentVideo/videos
 

Offline thinkfat

  • Supporter
  • ****
  • Posts: 2152
  • Country: de
  • This is just a hobby I spend too much time on.
    • Matthias' Hackerstübchen
Re: Siglent SDS2000X Plus
« Reply #599 on: March 03, 2020, 08:24:00 am »
The pot is full with a delicious soup, everyone is satisfied.

Only one is looking for the hair in the soup that doesn't exist. The one has to buy the soup he thinks there is no hair in the soup, even if it costs 10 to 20 times more than the tasty, cheap soup the others taste.

It's important to know the limits of the instrument you're using, wouldn't you agree? Especially for statistics, it's important to understand the sample size that is the base for the computation. Only then can you know if the instrument fits the intended use.
Everybody likes gadgets. Until they try to make them.
 