« Reply #175 on: March 29, 2016, 06:07:23 am »
Sorry I meant how did I manage to measure a 800MHz signal and get such good rise time on a 70/300MHz scope.
(the screenshots have some clues   :P )
Offline hendorog
« Reply #176 on: March 29, 2016, 06:23:48 am »
Doh - you even put red ellipses there.
I'm going back on holiday...
Offline hendorog
« Reply #177 on: March 29, 2016, 06:32:47 am »
So did ya hack the VGA into 900MHz mode?
Perhaps by resetting it after the scope had already started so it went into full bandwidth mode?
Offline tautech
« Reply #178 on: March 29, 2016, 07:08:20 am »
You're f***ing with our heads Bud, please put us out of our misery.  :-//
Online T3sl4co1l
    • Seven Transistor Labs
« Reply #179 on: March 29, 2016, 08:26:54 am »
I'm going to say direct injection into ADC input.
Tim
Seven Transistor Labs, LLC
Offline krivx
  • Country: ie
« Reply #180 on: March 29, 2016, 08:58:51 am »
Tim
This seems likely, should it be surprising that the firmware will happily go along with this and give measurements well outside the specs? I'm not sure what the best method for handling this would be.
Offline exe
  • self-educated hobbyist
« Reply #181 on: March 29, 2016, 09:29:48 am »
I would leave it as is because I don't see clear benefits in "sanity check" in this case because that would decrease "hackability". This also complicates the code, needless to say firmwares are not perfect, more complications potentially would make it less stable/feature rich.
Anyway, I think most measurements functions work with curves, not with measurement points. Performing sanity checks on this is just to tricky and quite pointless, imho.
Offline Circlotron
« Reply #182 on: March 29, 2016, 09:37:32 am »
Sooo..... who's going to be the first to offer a drive-in-drive-out scope update service for this baby?
Offline tautech
« Reply #183 on: March 29, 2016, 09:40:17 am »
The parts won't cost much, but the labour........
Online tggzzz
« Reply #184 on: March 29, 2016, 09:42:20 am »
Please explain the causal connection - if there is one.
I can make it infinitely "feature rich", provided it isn't required that the features actually work.
I really don't understand what you are trying to say. A digitising scope can only measure points. Waveforms are only curves, until you are dealing at the level of individual photons and electrons and quantum, physics.
If the manufacturers carelessly get that wrong, what else have they carelessly got wrong?
Old saying: "Fool me once, shame on you. Fool me twice, shame on me".
Offline exe
« Reply #185 on: March 29, 2016, 09:56:27 am »
You move cursors on a waveform, not between points. Analytic functions (or whatever they called) work on waveforms. Why should they care how and what waveform was acquired? In some cases it can be an indication that something is wrong or outside the speck. But this is not a proper self-test for this.
Why do you think this is wrong? Sorry, I feel like you just falling into "this is wrong and everyone who does not share my point of view is incompetent" without any reasoning.
Online tggzzz
« Reply #186 on: March 29, 2016, 10:26:53 am »
Ah. You mean functions that work on continuous curve inferred by interpolating between the sampled points. So that is going to depend on the interpolation chosen, the circumstances undet which the interpolation is and isn't valid, and on whether the implementation is correct.
Is there a hint there?
Please don't snip the relevant bits in order to (try to) make a point. You wrote "Performing sanity checks on this is just to tricky and quite pointless".
If an instrument doesn't pass a sanity check, should you continue to put faith in the instrument?
Offline AndyC_772
« Reply #187 on: March 29, 2016, 10:31:42 am »
Display mode in the top screen is X-Y. You're feeding in an external, slower time base on the X channel.
Offline exe
« Reply #188 on: March 29, 2016, 10:50:22 am »
Okay, I think I was too strong on statements. I don't say "no check is ever needed". I say four things 1) I don't trust rigol implementing them 2) shouldn't be part of analytical functions. 3) may affect hackability 4) sanity check is very limited when analyzing unknown waveform. And here is why I think so.
1)
analytical functions != Input validation is a different thing. Checking the input range or "space between points" is fine. But...
2)
If it is too strict on checking inputs this may greatly harm hackability. How so? Imagine you removed restriction on bandwidth. But the scope "doesn't know" about this. So it can indicate a failure in this case. Or if I "overclock" it it may refuse work faster because "it checks the specs".
3)
It happened to me I'm a software developer, so I know a little bit about software. I didn't get your point about features, but I repeat my statement: more code -- more hassle to maintain it. If I knew they have good software specialists I would welcome more checks. But since it's rigol I have a fear that they don't have expertise. They can easily degrade scope performance when adding more validation. And it's not uncommon to see "software gone mad".
4) I think best check is the analysis of a signal with a known waveform. I'm not sure it is possible to catch most failures just by analyzing user input.
Offline Gixy
« Reply #189 on: March 29, 2016, 11:29:41 am »
@AndyC:
This is the Cursor menu, X-Y is relative to the display of cursors, not the acquisition mode.
Offline Gixy
« Reply #190 on: March 29, 2016, 11:46:32 am »
Bandwidth means -3dB at max frequency. Let's suppose your input signal amplitude is 5V, you have here 4mV, so an attenuation of more than -61dB, so clearly outside bandwidth. The sampling rate being 2Gs/s, you're still in the Nyquist ratio of 2 samples/period, so the digitalized signal is still valid.
Online tggzzz
« Reply #191 on: March 29, 2016, 12:40:48 pm »
How did I make these screenshots ?
Is there any way that you can force the scope to show just the ADC samples, without interpolation / reconstruction / post-capture filtering?
For the rise/falltime, which points are being measured? What happens when the front-end is overloaded? Is it simply measuring noise and finding a signal?
« Reply #192 on: March 29, 2016, 02:16:31 pm »
Thanks guys for participating in the discussion. I will answer in the evening when get home, i do not like typing on my mobile. For now just can say there was no fiddling with the clock frequency, these are actual shots of high frequency and low rise/fall time signals. And still there is a common clue in both screenshots that seems has not been spotted yet.  :)
Offline exe
« Reply #193 on: March 29, 2016, 02:28:14 pm »
My guess is you hacked the input attenuator -- removed/modified low-pass filter.
Offline exe
« Reply #194 on: March 29, 2016, 02:33:32 pm »
I agree wit this. Removing interpolation and leaving only datapoints should show something triangle-like. Plus some input hacking imho.
Offline AndyC_772
« Reply #195 on: March 29, 2016, 02:39:36 pm »
@AndyC:
Ah, OK, fair enough.
Perhaps it's something to do with the signal level being extremely low (I see measurements of just a few mV... not sure how exactly they're being calculated)...? Very small amplitude means the system isn't limited by slew rate, because the V/sec rate is still within limits even if the input frequency is high.
Online tggzzz
« Reply #196 on: March 29, 2016, 04:07:11 pm »
Why are you choosing that specific reconstruction? Why not  a zero-order hold or, better, just the impulses?
Offline exe
« Reply #197 on: March 29, 2016, 09:41:21 pm »
I think I meant "just impulses" (=no reconstruction?). I'm no expert in this stuff, it's just a picture that popped into my head -- dots connected with straight lines. May be because I saw this on youtube when the guy chose "no interpolation" on the scope. Probably, it was in "the signal path blog" and with a review of a 100GHz scope.
Offline hendorog
« Reply #198 on: March 29, 2016, 10:21:48 pm »
The first screenshot has the bandwidth limit on - but is still measuring 800MHz...
Both traces look very sharp compared to the equivalent signal on my DS4k - is averaging turned on or is that typical when the scope is stopped?
« Reply #199 on: March 30, 2016, 02:23:09 am »
Nope, the waveform quickly becomes fuzzy after you cross the VGA set bandwidth limit.