Analogic kit is well known for quality and references like that are specifically designed to have long term stability and accuracy, even if not calibrated for years. If that's the only reference you have, and providing it works and the outputs make sense across the range, check against the best dvm you have, then it's already probably the most accurate thing you have in your lab, at least until you have something to compare it to. More than adequate to calibrate analog meters and probably dvm up to 4.5 digits. A good system dvm is essential. Older Hp models like the 3456, 6.5 digit, are an affordable option and are so old now, the aging rate is very low. Have a couple in the lab here, neither formally calibrated in years, but track together within a couple of microvolts once warmed up and are within a handful agaist a Fluke dc standard. The 3457 are 6.5 digits via front panel, 7.5 via hpib commands, but from tests, not convinced they are much better than the 3456. Afaics, Fluke make the best voltage standards, while HP make the best voltmeters. There also low cost options, such as online sellers offering precalibrated fixed output references. Quality might vary, but some of them look pretty good and could be usefull for an initial reference to compare against. A handful of microvolts is more than good enough for many. Any better than that and thermal ambient, lead and connector variations start to affect results. If you think you need better than that, then there should be good reason, as it will be an expensive journey. If the work doesn't care if the measurement is miliivolts out, then it's a waste of money to pay for single digit microvolt accuracy, however much we all strive for the best...