Picked up a used 34401a off ebay. The unit passes self test without errors.

RS232 works but GPIB does not.

Is there a common cause of failure for 34401a GPIB ports?

I'm guessing either U703 (75ALS162) or U702 (75ALS160) is bad. Is there any way to tell which one?
« Last Edit: April 23, 2017, 03:16:56 am by usagi »
 

Online Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2396
  • Country: de
Re: Agilent 34401a repair - dead GPIB
« Reply #1 on: April 18, 2017, 09:56:49 am »
Hello usagi,

see manual p. 92: 'Only one interface can be enabled at a time'.

So I assume, that you have to enable 'HP-IB / 488' first, in the I/O menu, see p. 162, and then it should work.

Frank
« Last Edit: April 18, 2017, 10:00:03 am by Dr. Frank »
 

Offline usagiTopic starter

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Agilent 34401a repair - dead GPIB
« Reply #2 on: April 18, 2017, 05:11:57 pm »
Hello usagi,

see manual p. 92: 'Only one interface can be enabled at a time'.

So I assume, that you have to enable 'HP-IB / 488' first, in the I/O menu, see p. 162, and then it should work.

Frank

i know this already  :) i had to go into those menus to enable RS232 in the first place, since it came with GPIB interface enabled by default.

so I can confirm GPIB is 100% busted.  :scared:  GPIB works on my other devices - keithley 2015 and R&S cmu200.

so do you know which exact 34401a IC to verify for GPIB troubleshooting?
« Last Edit: April 18, 2017, 05:18:32 pm by usagi »
 

Offline Tony_G

  • Frequent Contributor
  • **
  • Posts: 922
  • Country: us
  • Checkout my old test gear channel (link in sig)
    • TGSoapbox
Re: Agilent 34401a repair - dead GPIB
« Reply #3 on: April 19, 2017, 08:38:01 pm »
I just repaired an 11713A that didn't work with GPIB by resoldering the GPIB connector - Might be worth giving it a try if the connector is soldered direct to the board in those devices.

Offline usagiTopic starter

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Agilent 34401a repair - dead GPIB
« Reply #4 on: April 20, 2017, 07:24:27 am »
100% confirmed the connector is perfect.

100% confirmed an IC is blown. waiting on mouser order to arrive now.

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5491
  • Country: de
Re: Agilent 34401a repair - dead GPIB
« Reply #5 on: April 20, 2017, 07:39:32 am »
Once you exchange the GPIB controller it should work perfect again.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline usagiTopic starter

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Agilent 34401a repair - dead GPIB
« Reply #6 on: April 20, 2017, 08:25:08 am »
Once you exchange the GPIB controller it should work perfect again.

U701 is fine.

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5491
  • Country: de
Re: Agilent 34401a repair - dead GPIB
« Reply #7 on: April 20, 2017, 08:44:40 am »

U701 is fine.
Ahh, which IC do you think is blown?
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline usagiTopic starter

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Agilent 34401a repair - dead GPIB
« Reply #8 on: April 23, 2017, 03:16:40 am »
replaced blown IC, GPIB now works. again :-DMM


Online TheSteve

  • Supporter
  • ****
  • Posts: 3757
  • Country: ca
  • Living the Dream
Re: [FIXED] Agilent 34401a repair - dead GPIB
« Reply #9 on: April 23, 2017, 03:39:30 am »
Excellent!!!

Please do post which IC you actually changed to make it work though.