Author Topic: Anritsu sitemaster S331D  (Read 2450 times)

0 Members and 1 Guest are viewing this topic.

Offline 1VOXTopic starter

  • Newbie
  • Posts: 6
  • Country: cz
Anritsu sitemaster S331D
« on: May 30, 2023, 12:41:16 pm »
I have a faulty S331D. It reports integrator failed at start and VNA integrator failed during self test. When measuring the return loss, I feel it works. You don't know where to look for the error.
Thank you
 

Offline vishaldotgupta

  • Regular Contributor
  • *
  • Posts: 96
  • Country: in
Re: Anritsu sitemaster S331D
« Reply #1 on: April 18, 2024, 07:07:59 am »
your unit has faulty RF section.  most probably MMIC is cause of failure in such cases.  is your unit a single port or two port version.  I mean the two port unit comes with internal power meter option and has two N female test ports on the front panel
 

Offline 1VOXTopic starter

  • Newbie
  • Posts: 6
  • Country: cz
Re: Anritsu sitemaster S331D
« Reply #2 on: June 01, 2025, 06:56:32 pm »
I have one port.
 

Offline ME

  • Frequent Contributor
  • **
  • Posts: 687
  • Country: gb
Re: Anritsu sitemaster S331D
« Reply #3 on: June 01, 2025, 07:28:04 pm »
How about doing a sweep and see if it has rf coming out of its ass.
 

Offline 1VOXTopic starter

  • Newbie
  • Posts: 6
  • Country: cz
Re: Anritsu sitemaster S331D
« Reply #4 on: June 02, 2025, 07:33:43 am »
We will try it. The results of the measurement are probably correct. For loads of 35, 75 and 100 Ohm, it shows the correct return loss.
 