Author Topic: Tek 2755Ap: missing +9V  (Read 1452 times)

0 Members and 1 Guest are viewing this topic.

Offline Martin.MTopic starter

  • Frequent Contributor
  • **
  • Posts: 957
  • Country: de
  • in Tek we trust
    • vintage Tek collection
Tek 2755Ap: missing +9V
« on: March 29, 2024, 01:46:51 pm »
hello friends and happy easter,

I have the following problem:

Tektronix Spec: 2755Ap

missing +9V

I have not found application or manual.
There is a rest voltage around +1,2V, so it may be possible anywhere is a LM317 with a bad 240 ohms to ground: the only logical answer to this voltage.
But I don´t want to make shitty work here, this is a wonderful instrument. So please dear insiders, where I have to look for this problem?
It`s a ESD home, so I don`t like unscrew too much parts.

thank you for a tip and greetings from germany
Martin
 

Offline Testtech

  • Regular Contributor
  • *
  • Posts: 121
  • Country: us
Re: Tek 2755Ap: missing +9V
« Reply #1 on: March 29, 2024, 09:25:27 pm »
Check the fuses under the little cover on top of the PSU.
 
The following users thanked this post: Martin.M

Offline shakalnokturn

  • Super Contributor
  • ***
  • Posts: 2127
  • Country: fr
Re: Tek 2755Ap: missing +9V
« Reply #2 on: March 29, 2024, 11:04:00 pm »
You'll find at least the 494 schematic on KO4BB's website (think of donating once in a while, all the traffic and site maintenance cost him).
On these SA's à lot of tantalum capacitor problems with age.
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1916
  • Country: us
    • KE5FX.COM
Re: Tek 2755Ap: missing +9V
« Reply #3 on: March 30, 2024, 02:41:38 am »
All of the 275x models are rack-mount versions of one 49x model or another. The 2755AP is a repackaged 492BP if I recall correctly.  Basically a 494AP with 100 Hz minimum resolution rather than 10 Hz. 

 I'd probably use the 494AP manuals from ko4bb.com to troubleshoot it; the 494 manuals may be a little too old.
 

Offline Martin.MTopic starter

  • Frequent Contributor
  • **
  • Posts: 957
  • Country: de
  • in Tek we trust
    • vintage Tek collection
Re: Tek 2755Ap: missing +9V
« Reply #4 on: April 01, 2024, 11:05:37 am »
Check the fuses under the little cover on top of the PSU.

thank you. 2A was blown.
LM317 at front board was defect, also the driver transistor for ON/OFF the light.

 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf