Great review as always.
I've been contemplating a VNA purchase for a while now, and while I can't afford a really high end unit, I have the 8753ES on my list. How does that type of unit compare, in your opinion, to this Siglent (other than obviously having more bandwidth with the 6 GHz model)?
A couple of comments which might help.
The 8753ES is a full 2 port device, i.e. both ports can measure in both directions.
The SVA 1000's are T/R devices, which has one bidirectional port, and one input only port. In practice this means you need to reverse the device manually to measure all S-params, and it also limits the accuracy of S21 measurements a bit, but that can be mitigated somewhat.
The source assembly on any 8753 is prone to failure, is not repairable without a wire-bonder, and can be very expensive to replace.
A table of comparisons between the various 8753 units is here:
https://www.kirkbymicrowave.co.uk/Everything-you-wanted-to-know-about-the-HP-8753-VNA/Do you know what the part number of the surface mount directional coupler used in these is? I haven't been able to find one with similar specs.
The SVA's use an internal calibration, which means the raw performance of the coupler isn't what you see - unlike an HP VNA. Also, the spec mentions 'calibrated directivity' which is the effective directivity after a calibration, not the raw directivity of the uncorrected hardware.
This was suggested as a possible match to the part used in the SVA1015X.
https://www.mouser.de/datasheet/2/400/tdk_02022018_HHM22152A2_ver4_1(Oct.2017)-1284660.pdfWhile I don't know, it could be used in the 1032 as well. I do know that the 1015X coupler starts to affect performance noticeably about 2-3 MHz.
In the 1032X review at about 34:50 the device is showing full span. The very first data point at 100kHz is clearly bad on the Smith chart, and also on the RL chart. Perhaps that is caused by the coupler. The step size is 4MHz so the second point is back into its happy place.