Hi Aiy,
Thanks for the information, I took a look over your previous post about the faulty diode you had and was making some measurements on my probe so I have some photos and info to share.
My probe is a little different but very close to yours, see images attached.
It covers -70 to +20 dBm using three three diodes, probably because each has a different square law response.
On top side :
The same TI 2202AI opamp
Maxim MAX412 instead of the MAX410 dual opamp
Microchip 24AA64 same as yours
On the bottom we have:
Large IC with label (GAL16V8Z programable gate array) not on yours, this seems to be for diode switching.
Maxim MAX393 switching device (whereas yours is MAX383)
5 x diodes (W5D) and 3x ESD protection diodes (SMDA05 & LCDA05)
On the main meter PCB, J1 (the ribbon connecting the probe to Motherboard) I have recorded some interesting voltage readings.
I measured here because it was just easier that opening the probe again (see final image attached, J1 is the left side)
I also fitted a step attenuator between the 0dBm refence and the sensor.
Pin 1 is +5.5v when the sensor input power goes > -10dBm, otherwise it is low.
Pin 11 is +5.5v when the sensor input power goes < -10dBm, otherwise it is low.
This appears to be switch drive logic and depends on the power level detected, there may be a third state which I suspect is triggered at higher powers like +10 dBm etc.
Three states would match the three front end diodes.
Pins 9 and 10 carry the 245 Hz chopped & amplified signal from the sensor diodes, putting a scope on these shows 500mV p-p square wave when the probe is connected to the 50MHz 0dBm reference.
500mv seems quite low for a 0 dBm input signal and is probably why the meter is reading -10 dBm .. don't you think?
I will investigate the switching logic a bit more, I expect a third state will arise.
Then I will see if the logic is actually switching the diodes correctly.