[Repaired (mostly)] Broken Rigol DS1054Z - Blown PSU
« on: September 21, 2016, 01:28:29 pm »
Hi there everyone.

I had a freaky accident with my Rigol DS1054Z. I accidentally pulled the probe cables and dragged the scope from the desk to the floor  |O. It dropped with the screen facing down. My home differential protection circuit kicked in so I knew some sort of a short happened in there. The fuse was totally blown. I shacked the scope to listen for any damage and a piece of plastic fell of from inside. So i opend the scope and found that the internal PSU shorted a pin of the big electrolytic cap shorted against the metal cover.
I found the broken component, but I need help identifying it (don't know if it is a varistor or a capacitor of some sort) so if someone could help it would be great. I know I should also check the big cap but I have no ESR meter.
Any other advice? Should I make this scope a proper funeral after this event?? :P

I've attached a picture of the scope inside and signaled the broken component.

Thanks in advanced.
« Last Edit: September 28, 2016, 08:54:39 am by Paciente8159 »
 

Offline silverlego3

  • Contributor
  • Posts: 13
  • Country: us
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #1 on: September 21, 2016, 02:30:48 pm »
Hard to tell from the picture but it looks like it used to be a ceramic capacitor.
"Work Smarter, not harder"

"Its hard to soar like a eagle when your flying with turkeys"
 

Offline Orange

  • Frequent Contributor
  • **
  • Posts: 346
  • Country: nl
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #2 on: September 21, 2016, 02:44:14 pm »
Maybe it is an NTC to limit the rush-in current. if so they usually have values from 5..10 ohms.
Does the tear down video of Dave give you a clue ?
 

Offline OE2WHP

  • Regular Contributor
  • *
  • Posts: 94
  • Country: at
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #3 on: September 21, 2016, 02:51:02 pm »
 

Offline singapol

  • Frequent Contributor
  • **
  • Posts: 420
  • Country: sg
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #4 on: September 21, 2016, 04:36:29 pm »
The broken component is an NTC.Hope someone with the same model will be kind enough to give you the resistance value. Just be sure to get the right size or diameter, of course you can fit in a bigger one if the hole fits the lead. Replace the big cap and check if other components have failed. The power supply should be repairable and work again. ;)
 

Online Kleinstein

  • Super Contributor
  • ***
  • Posts: 14072
  • Country: de
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #5 on: September 21, 2016, 05:03:36 pm »
One should check the input side rectifier too. Chances are that it is only that poor little NTC that worked as a fuse.
 
The following users thanked this post: Paciente8159

Online PA0PBZ

  • Super Contributor
  • ***
  • Posts: 5121
  • Country: nl
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #6 on: September 21, 2016, 05:05:55 pm »
One should check the input side rectifier too. Chances are that it is only that poor little NTC that worked as a fuse.

I was just going to post that. Just replace it, even if it measures ok you can be sure it's not what it was before the accident.
The cap will be ok.
Keyboard error: Press F1 to continue.
 
The following users thanked this post: Paciente8159

Offline Paciente8159Topic starter

  • Newbie
  • Posts: 4
  • Country: pt
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #7 on: September 21, 2016, 05:12:48 pm »
Thanks guys.
After reopening scope my suspicions are that it was the NTC :P. The piece of plastic that felt from the had the letters "D-" on it like these.

If anyone could supply the correct value/reference it would be sweet.

Here is a better picture of what used to be an NTC.
There is the back picture of the point of contact too. I clipped the cap pins to give it more room.
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16272
  • Country: za
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #8 on: September 21, 2016, 05:27:59 pm »
Just grab the one out of the nearest dead ATX power supply, it will be around 10R cold resistance ( generally) and will work fine there. Also change out that bridge rectifier, it will have had a massive stress event from the single cycle of current that blew the side off the NTC, and the donor PC power supply one likely will still be fine ( normally the caps die on those cheap ones) do is a good fit in item.
 

Offline Assafl

  • Frequent Contributor
  • **
  • Posts: 600
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #9 on: September 21, 2016, 06:36:11 pm »
The component is an NTC.

It Reads NTC 10D-11.

I can't see why it would have shorted even if dropped... Perhaps a nasty angle?
« Last Edit: September 21, 2016, 06:38:28 pm by Assafl »
 
The following users thanked this post: Paciente8159

Offline Paciente8159Topic starter

  • Newbie
  • Posts: 4
  • Country: pt
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #10 on: September 21, 2016, 08:39:22 pm »
The component is an NTC.

It Reads NTC 10D-11.

I can't see why it would have shorted even if dropped... Perhaps a nasty angle?

Thank you Assafl.
Has I said the legs on that cap were left too long. Probably they were almost touching the metal frame. The oscillation from the fall was enough to allow an electric arch.

Thanks you all. I'll try to fix it and post the results after.
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #11 on: September 21, 2016, 10:51:15 pm »
Wow, that was quite a fall to bend the PCB enough to short the filter cap. I guess the leads were pretty long. Hopefully, the screen and encoders on the front panel are still OK.

Looking forward to your update once you get the power supply working again.
TEA is the way. | TEA Time channel
 

Offline Paciente8159Topic starter

  • Newbie
  • Posts: 4
  • Country: pt
Re: Broken Rigol DS1054Z - Blown PSU
« Reply #12 on: September 27, 2016, 11:14:06 pm »
One should check the input side rectifier too. Chances are that it is only that poor little NTC that worked as a fuse.

You were right.
The bridge rectifier was bad. Actually I could see it in circuit with the DMM.
I replaced the original (GBU 408) with a KBL406G salvaged from a dead PC PSU. I got a almost identical NTC (a 10D-9) from a broken laptop transformer.
The cap measured 136uF (~-10%) (I don't have an ESR meter to check the cap ESR), so I left it in place.
Replaced the fuse and the Rigol is now alive.

There is some noise/distortion that changes when I change the DC offset.
I haven't changed the 150uF cap. I'll probably order the correct components to see if it fixes the distortion, but giving that the scope is alive I'm happy. ;D
 