Hello everyone, thank you for taking interest in this issue.

This is an older E3631A, from '96 with 1.5 firmware version.
After power on the fan tuns on, there is a beep and the VFD lights up then it goes blank. When pressing the recall button, the unreg indicator is shown.
All voltages on the power rails appear good.

From the discussion in various topics on this forum, it appears that this indicates that the communication between the different CPUs doesn't work properly. So the signals from the main CPU was checked. On J4 to the front panel the supply voltages are present but all the signal lines are static. On J3 to the bottom panel some pulses can be seen from the IO CPU (which is passed on from the ASIC to the main CPU) but there is not response to that from the main CPU.

The oscillator of the main CPU is running, there's activity on the address/data lines.
On the reset pin (from the 5V regulator to the main CPU and the ASIC) however, there is periodic reset signal after ~200us. The reset pin of the regulator was lifted and it is stable, so the reset signal must come from the CPU or the ASIC.
It looks like something very basic is wrong here, that prevents the main CPU from properly starting.

Did anyone see this kind of failure? Any tip is welcome on what to check, how to proceed.
The contents of the EPROM was read, it looks valid but I cannot be sure that it is not corrupt. If someone had this firmware and could post it, it would be highly appreciated.
Thanks!
 

Offline vtp

  • Regular Contributor
  • *
  • Posts: 75
Re: Faulty HP E3631A
« Reply #1 on: April 03, 2017, 10:24:10 am »
Did anyone see this kind of failure?

Yes, exactly like that. Change the SRAM.
 
The following users thanked this post: z01z

Offline Samogon

  • Frequent Contributor
  • **
  • Posts: 457
  • Country: us
Re: Faulty HP E3631A
« Reply #2 on: April 04, 2017, 05:51:38 am »
Did anyone see this kind of failure?

Yes, exactly like that. Change the SRAM.
Yeah same with me. U15 was dead.
 
The following users thanked this post: z01z

Offline z01zTopic starter

  • Regular Contributor
  • *
  • Posts: 151
Re: Faulty HP E3631A
« Reply #3 on: April 04, 2017, 05:58:19 am »
I'll get a replacement and report back.
Thanks!
 

Offline z01zTopic starter

  • Regular Contributor
  • *
  • Posts: 151
Re: Faulty HP E3631A
« Reply #4 on: April 04, 2017, 06:46:58 pm »
It was the RAM indeed. Now the diag test is passed without an error.
The +6V output doesn't work properly though, it is stuck at -0.29V and 33mA in CC mode. Back to making measurements...

@vtp
Funnily I've just found your topic about the 1.6 fw, in which you described the faulty RAM symptom.
Drop me a PM with your email address, if you need the 1.5 fw.
 

Offline z01zTopic starter

  • Regular Contributor
  • *
  • Posts: 151
Re: Faulty HP E3631A
« Reply #5 on: May 08, 2017, 05:22:14 pm »
For the record, the 6V channel malfunction was caused by a faulty HCNR200 optocoupler, the LED measured 90 Ohms.
 