« Reply #25 on: February 14, 2018, 11:03:47 am »
From memory I think it's 9600 baud
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #26 on: February 14, 2018, 11:20:40 am »
This is serial to original controller ;)

Can I ask where you were providing this data, and at what baud rate (9600, as Mike says?) 

The MAX202 that's next to the 10-pin port labeled "RS232" doesn't have +5 on its VCC pin by default.  I tried applying +5 externally and sending a couple of commands (!MR<cr> and !T1230<cr>) at rates from 110 to 115200, but nothing happened.

 

Online mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13748
  • Country: gb
    • Mike's Electric Stuff
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #27 on: February 14, 2018, 11:34:04 am »
This is serial to original controller ;)

Can I ask where you were providing this data, and at what baud rate (9600, as Mike says?) 

The MAX202 that's next to the 10-pin port labeled "RS232" doesn't have +5 on its VCC pin by default.  I tried applying +5 externally and sending a couple of commands (!MR<cr> and !T1230<cr>) at rates from 110 to 115200, but nothing happened.


Can you post a photo of your controller to see if it's the same as mine - may be some link options or different firmware
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #28 on: February 14, 2018, 12:00:11 pm »
Can you post a photo of your controller to see if it's the same as mine - may be some link options or different firmware

Here's a link to the photo (too big to post in the thread directly):

http://www.ke5fx.com/fx_ctrl.jpg

The 27C512 EPROM is labeled "V3.1 MOD", with the checksum 317E.  Binary image is here.

There is another chip that looks like a mask-programmed ROM, which is labeled 43805-65202-F SN 4312E-082 MX20 MICROPROS.
 

Online mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13748
  • Country: gb
    • Mike's Electric Stuff
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #29 on: February 14, 2018, 12:15:07 pm »
Can you post a photo of your controller to see if it's the same as mine - may be some link options or different firmware

Here's a link to the photo (too big to post in the thread directly):

http://www.ke5fx.com/fx_ctrl.jpg

The 27C512 EPROM is labeled "V3.1 MOD", with the checksum 317E.  Binary image is here.

There is another chip that looks like a mask-programmed ROM, which is labeled 43805-65202-F SN 4312E-082 MX20 MICROPROS.
Strange about the unconnected supply to the 232 transceiver - may be worth seeing if you can trace where it goes to get a clue as to why it's disabled ( a fine wire brush can be useful for this).
 

Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #30 on: February 15, 2018, 08:15:39 am »
Strange about the unconnected supply to the 232 transceiver - may be worth seeing if you can trace where it goes to get a clue as to why it's disabled ( a fine wire brush can be useful for this).

I suspect that it's some kind of factory diagnostic port, given that nothing was originally plugged into it.   Ordinarily I would X-ray it to find out where the pins go, but...  ;D

Contrary to what I said earlier, there actually was a 9-pin connector in the digitizer area, connected to J5 on the controller board, but it doesn't have a MAX202 nearby and it doesn't act like an RS-232 port.  Pins 3, 4, and 6 are pulled up to +5, with all others low.  It appears on the block diagram in the manual as "Cable A." 
 

Offline kilohercasTopic starter

  • Regular Contributor
  • *
  • Posts: 60
  • Country: 00
  • Engineer
« Last Edit: February 16, 2018, 02:25:16 pm by kilohercas »
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #32 on: February 26, 2018, 04:45:27 pm »
KE5FX,

The 9 pin connector (cable A) located low down on the rear of the cabinet is the standard RS232 serial communications link to a PC serial port. I have various versions of MX-20 control software but they all expect to find a camera module connected so sadly they will not work for you :(

Fraser
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #33 on: February 26, 2018, 11:17:12 pm »
If it's a standard port, it certainly doesn't act like one.   :-// 

The path of least resistance in my case will probably be to run some spare I/O lines from the FPGA board to optocouplers across the up/down and time/voltage keys.  Just seems like unnecessary hackery, that's all.
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #34 on: February 26, 2018, 11:44:29 pm »
I think I have the schematics for the control board. If so I will check the serial port design for you.

It will have to be tomorrow though.

Fraser
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #35 on: February 27, 2018, 12:00:27 am »
I just had a thought........

The early MX-20 Digital with Bioptics camera array is not as highly integrated with the OC in terms of control.

From memory the PC controls the camera side of things but only provides monitoring of the safety interlocks, X-Ray on condition and initiatedps X-Ray off. It does not need a complex RS232C control set to achieve such limited monitoring and control. It could even be achieved with just the handshake lines !

The later MX-20 design provides much more control over the cabinet from the PC. The PC is able to monitor and set Exposure and Kv as well as control the X-Ray generator. I am uncertain what changes were made to the control PCB but IIRC it had an additional 7 segment LED for the time window for tenths of a second. I both versions of the MX-20 X-Ray head so have the EPROM's if that is all that needs to be changed.

Fraser
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #36 on: February 27, 2018, 01:44:43 am »
I think I have the schematics for the control board. If so I will check the serial port design for you.

It will have to be tomorrow though.

Fraser

 :-+ That would rock, if you can find schematics.  Even if your board is different from mine (and it probably is, if you have 0.1-second exposure time resolution) it could still come in very handy in the future.

To date, the only connection I've added is a wire from the "X-ray on" LED to an input line on my digitizer board, so that the software can tell when the exposure starts and stops.  If I eventually decide to add remote controls to some of the keypad buttons, a schematic could save some hunting and poking, even if it's not the right version.
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #37 on: February 27, 2018, 06:59:58 pm »
OK, I do have schematics for the control board...... of the original film type MX-20 circa 1997.

Whilst the controller does have a RS232 port it does not seem to be for remote control as the film cabinet had no need for such (having no PC).

Looking at the schematics I can also seen references to a 9 pin AEC sensor D type connector. I will upload the controller schematics but I am wondering........ do you have a Digital era MX-20 or the original film type cabinet that does not have the remote control capability or digital camera in the extended base area ?

The film type cabinet did have the uptight of an auto exposure X-Ray sensor that sat under the cabinet chamber and helped the user achieve good exposures of the film.

A digital era MX-20 would have the camera option detailed on the cabinets ID label on its rear. There would be a "DC" option type shown.

Fraser
« Last Edit: February 27, 2018, 08:40:25 pm by Fraser »
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #38 on: February 27, 2018, 07:32:26 pm »
Schematics for the non digital 1997 Faxitron MX-20......
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 
The following users thanked this post: KE5FX

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #39 on: February 27, 2018, 07:34:12 pm »
Continued
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #40 on: February 27, 2018, 07:36:00 pm »
Continued
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #41 on: February 27, 2018, 07:38:20 pm »
Last pictures for now :)
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 
The following users thanked this post: eliocor

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #42 on: February 27, 2018, 08:31:12 pm »
Schematics for the non digital 1997 Faxitron MX-20......

Much appreciated.  Here's a tidied-up .PDF version: Faxitron_MX20_43805_65201D_65202C_schematics.pdf (6.6 MB)

These drawings apply to rev C of the 43805-65202 controller, while my PCB has revision F of the same part #.  So they should be a good match.  U7 is the mystery MAX202 with no Vcc connection, for instance. 

My 9-pin connector goes to the "Medoptics interface" at J5, which is definitely some sort of proprietary interface.  One difference between revs C and F may be pin 3 of this connector.  It's grounded on your rev-C schematic, while it seems to be connected to something else on the rev-F PCB.  That may have been done to add proprietary control support to film-only machines.
« Last Edit: February 27, 2018, 09:04:37 pm by KE5FX »
 
The following users thanked this post: eliocor

Offline danielbriggs

  • Regular Contributor
  • *
  • Posts: 127
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #43 on: March 15, 2018, 11:21:47 pm »
This is an interesting thread, and much along the lines of what I'm slowly heading towards with my machine.

Great work kilohercas!  8)

Do you have any more details on the CBCT MATLAB route? I tried several algorithms myself, but am curious to what you used in the end as your results look really sharp.
[I found while most algorithms worked, some were ultra efficient, and some took a huge amount of computational effort for mixed results].

Even better... care to share you MATLAB .m files // or post an example of what you did?


@KE5FX + Mike...
Attached are my findings from when I was having a sniff around the RS232. It's not an exhaustive list, but contains pretty much all you'd need to run the cabinet from RS232.
[Excuse the notes too, it's just what I noticed the bus doing and scribbled down the findings].
 
The following users thanked this post: Fraser, eliocor, KE5FX

Offline Manawyrm

  • Contributor
  • Posts: 20
  • Country: de
    • tbspace.de
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #44 on: January 21, 2022, 08:33:37 pm »
@KE5FX:
Almost 4 years later, I now own one of these beautiful machines myself.
I'm at the same point as you, with the same Rev. F PCB and also missing the +5V supply on the MAX202.

Have you done any further work on your machine? How did it turn out?

@Fraser:
You seemed to have a different/newer firmware. Would you happen to have a backup copy of the EEPROM?
It might be a good idea to start a little collection.
My machine also has the "v3.1 MOD" ROM.

Thanks,
Tobi
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #45 on: January 21, 2022, 08:59:10 pm »
I haven't had time to do any more tinkering with it, unfortunately.  It works well enough with the keypad, although I'd still prefer to have full remote control if it could be achieved without a lot of modification work. 

One thing I'm still hoping to learn is the nature of the "seasoning" requirement, in which the manual says to perform a 5-minute warmup run at 20 kV before doing anything else.  Exactly how often this should be done, and exactly why it would be necessary at such low voltages, has never been explained.  That's the sort of thing I'd automate if I could, but would much prefer to skip altogether.
 

Offline Manawyrm

  • Contributor
  • Posts: 20
  • Country: de
    • tbspace.de
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #46 on: January 21, 2022, 09:07:14 pm »
Thanks for the reply!

Have you figured out how to remove/access the sensor yet?
It looks like the chassis is slightly different than described in my (slightly too new) 2009 service manual.
 

Offline Fraser

  • Super Contributor
  • ***
  • Posts: 13170
  • Country: gb
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #47 on: January 21, 2022, 09:30:06 pm »
I sold all of my MX-20 machines so regret I cannot help with firmware images.

Fraser
If I have helped you please consider a donation : https://gofund.me/c86b0a2c
 

Offline Manawyrm

  • Contributor
  • Posts: 20
  • Country: de
    • tbspace.de
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #48 on: January 21, 2022, 10:16:15 pm »
... I'd love to talk more about my reverse engineering effort, should I do it here in this thread or open a new one?

Anyway:
I've taken a closer look at the "3.1 MOD" firmware.
It's pretty straight-forward to read, but there are no references to the serial I/O ports anywhere in the code.
That pretty much means that the 3.1 MOD firmware does not have any serial capabilities at all.
 

Offline KE5FX

  • Super Contributor
  • ***
  • Posts: 1894
  • Country: us
    • KE5FX.COM
Re: Faxitron MX-20 Reverse engineering (with some good results)
« Reply #49 on: January 21, 2022, 10:26:00 pm »
... I'd love to talk more about my reverse engineering effort, should I do it here in this thread or open a new one?

Here's good!   :-+

Thanks for the reply!

Have you figured out how to remove/access the sensor yet?
It looks like the chassis is slightly different than described in my (slightly too new) 2009 service manual.

I had to build my own digitizer, since my MX-20 didn't come with its original PC.  I haven't made any modifications to the upper part of the machine, though, except for connecting the 'X-RAY ON' LED to the FPGA board. 
 