Calibration of Fluke 15/17 B+ ("PLUS")
« on: April 07, 2015, 02:06:41 pm »
Hi,
has anyone an idea to calibrate a Fluke 15 or 17 B+?

Not the "B"-series without the "+", I mean the "B+"!

They are completely different with the "B" (without "+"), there are NO trimpot inside, but contacts to connect a software based calibration unit. One of these contacts are labeled "CAL".

Ralf.

 

Offline DG5SAYTopic starter

  • Contributor
  • Posts: 42
  • Country: de
Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #1 on: July 10, 2016, 07:34:11 pm »
Yes!  :-+

I can´t believe it!

FLUKE published the calibration manual for the "+" models:

http://assets.fluke.com/manuals/15BPlus_cmeng0000.pdf

 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #2 on: July 10, 2016, 09:48:06 pm »
« Last Edit: July 10, 2016, 09:50:20 pm by retiredcaps »
 

Offline DG5SAYTopic starter

  • Contributor
  • Posts: 42
  • Country: de
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #3 on: July 11, 2016, 11:01:52 am »

[/quote]
Do you have the equipment to calibrate/adjust your 15B+ now?  History at

https://www.eevblog.com/forum/testgear/hear-kitty-kitty-kitty-nope-not-that-kind-of-cat/msg809723/#msg809723
[/quote]

You don´t need a "special Fluke calibrator". A stable voltage source, checked with an other (calibrated?) DMM is good enough for a 15B+.
 

Offline boggis the cat

  • Regular Contributor
  • *
  • Posts: 218
  • Country: nz
Re: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #4 on: July 11, 2016, 12:02:06 pm »
If you want to adjust the multimeter, then you'll have to be able to source the exact specified voltage, current, and resistance values.

I haven't checked the 17B+, but for AC voltage it is typical for Fluke to specify 60 Hz and a high frequency (e.g. 20 kHz to 10 kHz for the Fluke 87 V) as parameters.  If those aren't very accurate (either voltage or frequency) then you'll pass the inaccuracy to the multimeter -- i.e. mis-calibrate it.  The signal also has to be sinusoidal and fairly clean from noise.  Purpose-built calibrators (from Fluke or other providers) are the usual approach.

On the other hand, just checking the accuracy against a second known multimeter (ideally, recently calibrated with full data) only requires reasonably stable signals from any convenient source.
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16560
  • Country: 00
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #5 on: July 11, 2016, 10:32:11 pm »
You don´t need a "special Fluke calibrator". A stable voltage source, checked with an other (calibrated?) DMM is good enough for a 15B+.

Maybe you'd like to walk us through the procedure after reading that document.

I've got a super-stable voltage source. It's connected to the meter. Now what...?

 

Offline DG5SAYTopic starter

  • Contributor
  • Posts: 42
  • Country: de
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #6 on: July 12, 2016, 01:50:18 pm »
You don´t need a "special Fluke calibrator". A stable voltage source, checked with an other (calibrated?) DMM is good enough for a 15B+.

Maybe you'd like to walk us through the procedure after reading that document.

I've got a super-stable voltage source. It's connected to the meter. Now what...?

OK, no problem:

1. remove the calibration sticker
2. connect a external 3 V battery to the battery contacts
3. turn the rotary switch to "VDC" (if you wish to recalibrate VDC, of course)
4. connect the meter to your "super-stable voltage source" with exactly 3.000 VDC and wait a few seconds to stabilize
5. short on the PCB the contacts "CAL1" and "WP6" (there are under the cal-sticker you have removed) until you hear a short "beep" and the high voltage mark appears on the display
6. push the [HOLD]-button to confirm
7. turn off the multimeter

That´s it!
Repeat this steps for VAC (or other functions, see table 7 in the calibration manual) similar.

It works fine on my 15B+


 
The following users thanked this post: invzim, tautech

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16560
  • Country: 00
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #7 on: July 12, 2016, 02:01:51 pm »
It works fine on my 15B+

Ah, I mes-read that bit.  :-[

Yeah, it should work. All you need is the super-stable references.
 

Offline AlxDroidDev

  • Frequent Contributor
  • **
  • Posts: 471
  • Country: br
    • Arduino Web Brasil
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #8 on: December 15, 2016, 07:09:50 pm »
It works fine on my 15B+

Ah, I mes-read that bit.  :-[

Yeah, it should work. All you need is the super-stable references.

Would a MAX6126AASA41+  and a 12-bit DAC controlled by an Arduino be good enough to create a 3.000V reference for calibrating a 17B+ (VDC only) ? Or should there be a much more precise DAC (>= 16-bit) ?

I just found out that my 17B+ is about 2% off, which has been enough to steer me on the wrong direction when working with it.

I discovered that because I couldn't figure out what was wrong with my project and decided to double check everything with a Fluke 117 and a Mastech MS8260E, both of which gave me the same reading, confirming that my circuit and software were correct. The Fluke 117 came with a factory calibration certificate, while the 17B+ didn't. Both were bought in Sept. 2015.
"The nice thing about standards is that you have so many to choose from." (Andrew S. Tanenbaum)
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16560
  • Country: 00
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #9 on: December 15, 2016, 07:21:56 pm »
Would a MAX6126AASA41+  and a 12-bit DAC controlled by an Arduino be good enough to create a 3.000V reference for calibrating a 17B+ (VDC only) ? Or should there be a much more precise DAC (>= 16-bit) ?
What's the DAC for?

Just get a MAX6126AASA30+ (3V reference).

5 bucks from mouser.

(or ask MAXIM for a free sample, you never know...)
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #10 on: December 15, 2016, 07:27:41 pm »
I just found out that my 17B+ is about 2% off, which has been enough to steer me on the wrong direction when working with it.
I think one other member reported that his 17B+ was out of calibration right from the start, but never sent it back to the ebay seller due to high shipping costs.
 

Offline AlxDroidDev

  • Frequent Contributor
  • **
  • Posts: 471
  • Country: br
    • Arduino Web Brasil
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #11 on: December 16, 2016, 01:14:38 am »
What's the DAC for?

To trim the 4.096V of the MAX6126xx41 to 3.000V.

I already have a MAX6126AASA41+, that I got for my programmable load neverending project, and a few DACs of different precisions.

I don't work with VAC, so I have no way of telling if my 17B+ is calibrated on that. I can tell, however, that it's temperature measurement are also quite off: I just built a breadboard circuit to with a LM335AZ and a DS1631+ to compare against the it, and the DMM was reading >4 degress above ambiente temperature (actual: 18.5C, read: 22.7C).

A couple hours ago I wrote Fluke asking for quote on a calibration for the 17B+, because I suspect all of its readings are out of spec.
"The nice thing about standards is that you have so many to choose from." (Andrew S. Tanenbaum)
 

Offline Fungus

  • Super Contributor
  • ***
  • Posts: 16560
  • Country: 00
Re: Solved: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #12 on: December 16, 2016, 09:47:56 am »
What's the DAC for?

To trim the 4.096V of the MAX6126xx41 to 3.000V.

Seems like a weird way to do it.

Still, anything that's better than 2% accuracy will be an improvement right now.  :popcorn:

In theory you could do it with an LM317 and a potentiometer. Connect it to both the calibration input of the 17B+ and the voltage measurement of the Fluke 117 (and the Mastech if you want). Dial 3.00V and press 'calibrate'.
« Last Edit: December 16, 2016, 09:55:33 am by Fungus »
 

Offline AndyBollo

  • Newbie
  • Posts: 5
  • Country: mx
Re: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #13 on: November 19, 2017, 09:46:34 pm »
Even an $8 USD multimeter just out of the box performs better than the FLUKE 17B+. Shame on you Fluke!  Respect to the fuke 17B+, Amprobe was very superior in all tests! Does not matter if the 17B+ was made for the asian market, carry the brand "Fluke" and should shows quality and accuracy!!!!!
« Last Edit: November 19, 2017, 09:59:45 pm by AndyBollo »
 

Offline HalFET

  • Frequent Contributor
  • **
  • Posts: 512
  • Country: 00
Re: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #14 on: November 19, 2017, 10:19:25 pm »
Even an $8 USD multimeter just out of the box performs better than the FLUKE 17B+. Shame on you Fluke!  Respect to the fuke 17B+, Amprobe was very superior in all tests! Does not matter if the 17B+ was made for the asian market, carry the brand "Fluke" and should shows quality and accuracy!!!!!

Actually that one is still within call if I recall the 17B+ specs: 0.5% on DC volts, so 20 mV error on 5V is quite acceptable...  :palm: 

Additionally, there's a reason we say precision instrument, and not accurate instrument! Precision means it's repeatable and the measurements are tightly grouped, offsets are acceptable as long as they are repeatable. It's a common mistake to assume that a good multimeter will always indicate the exact value, if you want that you better pull out the big bucks and invest in a high-end HP or Keithley with a LTZ1000 or LM399 reference (each of which will cost more than your Chinese multimeter), and be prepared to leave them on 24/7, and dish out the 800 bucks it costs each year to get it calibrated. And even those will drift and give an error, but instead of ppt it'll be ppm there. I suppose you could always get yourself a Josephson standard and a Caesium beam standard to drive it.  |O I check my 17B+ against my calibrated Keithley 2001 every time I start using it on the bench, and haven't noticed any real drift on the 17B+ yet.
 

Offline AndyBollo

  • Newbie
  • Posts: 5
  • Country: mx
Re: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #15 on: November 19, 2017, 11:59:47 pm »
Even an $8 USD multimeter just out of the box performs better than the FLUKE 17B+. Shame on you Fluke!  Respect to the fuke 17B+, Amprobe was very superior in all tests! Does not matter if the 17B+ was made for the asian market, carry the brand "Fluke" and should shows quality and accuracy!!!!!

Actually that one is still within call if I recall the 17B+ specs: 0.5% on DC volts, so 20 mV error on 5V is quite acceptable...  :palm: 

Additionally, there's a reason we say precision instrument, and not accurate instrument! Precision means it's repeatable and the measurements are tightly grouped, offsets are acceptable as long as they are repeatable. It's a common mistake to assume that a good multimeter will always indicate the exact value, if you want that you better pull out the big bucks and invest in a high-end HP or Keithley with a LTZ1000 or LM399 reference (each of which will cost more than your Chinese multimeter), and be prepared to leave them on 24/7, and dish out the 800 bucks it costs each year to get it calibrated. And even those will drift and give an error, but instead of ppt it'll be ppm there. I suppose you could always get yourself a Josephson standard and a Caesium beam standard to drive it.  |O I check my 17B+ against my calibrated Keithley 2001 every time I start using it on the bench, and haven't noticed any real drift on the 17B+ yet.


Sorry for expect a better accuracy out-of-the-box than cheaper models of multimeters, at least I would expect that from a  Fluke (It does not matter if it is a low-end product or where it was manufactured)
 

Offline HalFET

  • Frequent Contributor
  • **
  • Posts: 512
  • Country: 00
Re: Calibration of Fluke 15/17 B+ ("PLUS")
« Reply #16 on: November 20, 2017, 06:11:02 am »
The accuracy is within tolerance, difference is that the Fluke has input protection, will stay within cal if you throw it in your toolbox for a year, and can survive a drop left and right.
 