I see firmware V2.02 is available!
https://www.eevblog.com/product/121gw/
 
The following users thanked this post: Marco1971

Offline amspire

  • Super Contributor
  • ***
  • Posts: 3802
  • Country: au
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #276 on: September 28, 2019, 07:18:02 am »
I see firmware V2.02 is available!
https://www.eevblog.com/product/121gw/
And a new manual update.

Looks like they added a check of the battery voltage before doing a firmware update in the latest release. Sounds like a good idea.
 

Offline daveyk

  • Frequent Contributor
  • **
  • Posts: 413
  • Country: us
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #277 on: November 13, 2019, 06:15:23 pm »
I'm still not thrilled with resistance mode.  It is still somewhat slow, but putting the two leads together, it takes 2.5 seconds and then readings float between .018 to .103 ohms.  The longer you hold the leads together the more the resistance reading goes up.  V2.02, I believe is better than v2.00 in resistance.  I've swapped leads and same thing.  Perhaps the leads I have are not ideal, but resistance readings so sluggish make it hard to use for trouble-shooting.  Its hard to know when to press the REL button as it takes a while for the readings to settle down when shorting the leads together.

On another subject, has anyone thought to do some customer firmware for working on older automobiles?   I want to read RPM with a clamp-on probe (already have) and then drell angle (not real familiar with that).  I am fixing-up (not down, but it is currently - lol) a 1974 Beetle as a hobby.  I bought an automotive meter that is supposed to do this, but what a piece of fragile crap.  RPM is basically reading frequency.  I couldn't get that meter to read the RPM consistently.  I have no idea what dwell angle is, but when I get to tuning the engine they say at a certain RPM set the timing for a dwell angle of xx.  Anywho, I don't expect that, but did wonder if people have been doing customer firmware for this meter.

Dave
 

Offline npelov

  • Frequent Contributor
  • **
  • Posts: 330
  • Country: bg
    • Microlab.info
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #278 on: November 29, 2019, 12:25:06 pm »
Since v2.0 the SD card logging stops working properly after a while. I haven't tried that in 2.02. The last time I recorded voltage for 19.5-20 hours with interval 2 seconds, fixed range. I monitored the display - the battery dropped gradually with cut off at about 10.8V. At about 19.5 hours the voltage was 11. At 19.0 hours - 11.8V. However since about 2h 40min (about 4800 recods) the value in csv is  12.529V +/- 3 LSD. The last digit keeps moving around like it's recording real data, but it has nothing to do with what's on the display.

I've reported this before, but I don't see it in firmware change log. Is this fixed? Does anyone else have similar problems?

Edit: Also since v2.0 the multimeter will occasionally freeze when I start logging - switching ranges, pressing buttons have no effect, until I turn it off.

P.S. Yes, I know that bluetooth exists. I don't want to rely on wireless for long term logging.
« Last Edit: November 29, 2019, 12:34:52 pm by npelov »
 

Offline eV1Te

  • Regular Contributor
  • *
  • Posts: 186
  • Country: se
  • Your trusted friend in science!
    • richardandersson.net
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #279 on: December 02, 2019, 03:19:34 pm »
Since v2.0 the SD card logging stops working properly after a while. I haven't tried that in 2.02. The last time I recorded voltage for 19.5-20 hours with interval 2 seconds, fixed range. I monitored the display - the battery dropped gradually with cut off at about 10.8V. At about 19.5 hours the voltage was 11. At 19.0 hours - 11.8V. However since about 2h 40min (about 4800 recods) the value in csv is  12.529V +/- 3 LSD. The last digit keeps moving around like it's recording real data, but it has nothing to do with what's on the display.

I've reported this before, but I don't see it in firmware change log. Is this fixed? Does anyone else have similar problems?

Edit: Also since v2.0 the multimeter will occasionally freeze when I start logging - switching ranges, pressing buttons have no effect, until I turn it off.

P.S. Yes, I know that bluetooth exists. I don't want to rely on wireless for long term logging.

This sounds like a serious issue, however I am confused how to reproduce this...

The total battery voltage in the 121GW should be about 6 V, not 10.8 to 11.8 as you mention. There are 4 batteries, not 8.

What was the voltage source that you measured? Was it a constant voltage of 12.529 or did it change during the measurement? Did you check the voltage with a different meter during the test to confirm that the 121GW indeed measured/logged incorrect values?



 

Offline fabrizio_fabrice

  • Supporter
  • ****
  • Posts: 21
  • Country: us
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #280 on: December 14, 2019, 04:03:24 am »
Can  anyone who has updated to v 2.2 say how long to expect the process to remain in the "douun" mode?
 

Offline Scottjd

  • Frequent Contributor
  • **
  • Posts: 436
  • Country: us
    • YouTube Gadget Reveiws
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #281 on: December 14, 2019, 05:47:44 am »
Can  anyone who has updated to v 2.2 say how long to expect the process to remain in the "douun" mode?
I am assuming you mean firmware version 2.02?
First make sure the new firmware file on the SD card is named “EEVBlog.bin” with no spaces in the file name or the meter might hang and just display down. If the current installed firmware is old then you might want to make sure you have a fresh set of batteries installed before running the firmware update. I think I’ve read the newer firmware will check the battery level first.

1. Turn the meter off then hold down MEM and HOLD while turning on the meter.  You will see IAP on the display.
2. Press the “SETUP” button and it will display “DOWN / DOUUN” and after 5 seconds the meter bar should start rising from the left to right.
3. Once the bar hits the end the meter will restart displaying the new version of firmware installed.
This whole process after it displays “DOWN / DOUUN” shouldn’t take more then 15 seconds.
Please be sure to check out my YouTube channel and subscribe if you like the videos. https://www.youtube.com/c/GadgetReviewVideos

By people subscribing and giving thumbs up I know what I am doing is still wanted and adding value, then will continue to release new videos. Thank you for your support.
 
The following users thanked this post: fabrizio_fabrice

Offline fabrizio_fabrice

  • Supporter
  • ****
  • Posts: 21
  • Country: us
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #282 on: December 16, 2019, 04:51:27 am »
Can  anyone who has updated to v 2.2 say how long to expect the process to remain in the "douun" mode?
I am assuming you mean firmware version 2.02?
First make sure the new firmware file on the SD card is named “EEVBlog.bin” with no spaces in the file name or the meter might hang and just display down. If the current installed firmware is old then you might want to make sure you have a fresh set of batteries installed before running the firmware update. I think I’ve read the newer firmware will check the battery level first.

1. Turn the meter off then hold down MEM and HOLD while turning on the meter.  You will see IAP on the display.
2. Press the “SETUP” button and it will display “DOWN / DOUUN” and after 5 seconds the meter bar should start rising from the left to right.
3. Once the bar hits the end the meter will restart displaying the new version of firmware installed.
This whole process after it displays “DOWN / DOUUN” shouldn’t take more then 15 seconds.
Thanks, Scottjd, yes, I followed the instructions to a Tee, but it stayed in DOUUN mode for at least 6 hours.  It started at 9 PM, but I couldn't sleep well, so up at 3 am to check.  No change, still DOUUN, except the display contrast was changing as the batteries dropped from ~ 1.54 v at the start, to 1.47.

With much trepidation I turned the meter off, then back on and it was still v. 1.57.   I repeated the process with a different microSD using the same downloaded 2.02 .bin file, but this time it worked, and it was very quick.  It immediately showed % progress and the analog bar incremented, too.   5 seconds, maybe, and it was done!  A new set of batteries, and it's ready to go!  :-+   Thanks again!
 
The following users thanked this post: Scottjd

Offline Sparky Faraday

  • Regular Contributor
  • *
  • Posts: 59
  • Country: us
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #283 on: February 09, 2020, 10:16:14 pm »
Are all the functions reliable now? Anything crash?
Thank you,
Sparky
 

Offline Kibabalu

  • Regular Contributor
  • *
  • Posts: 106
  • Country: de
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #284 on: March 05, 2020, 07:12:32 am »
Just received my 121GW with firmware version 2.04. No issues discovered so far during first day of use.
 
The following users thanked this post: Marco1971

Offline pppopol

  • Newbie
  • Posts: 6
  • Country: fr
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #285 on: March 05, 2020, 06:32:17 pm »
Are you sure it's V2.04? On the website is V2.02?
 

Offline Brumby

  • Supporter
  • ****
  • Posts: 12298
  • Country: au
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #286 on: March 05, 2020, 09:13:42 pm »
Dave provided v2.04 a few weeks ago: HERE
 
The following users thanked this post: Xenawise, pppopol

Offline ceut

  • Regular Contributor
  • *
  • Posts: 124
  • Country: fr
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #287 on: March 06, 2020, 02:25:20 pm »
Just received my 121GW with firmware version 2.04. No issues discovered so far during first day of use.

Hello,
How fast is it in voltage mesurement with this new firmware for example with a lab power supply ?
Is there some lag when changing range ?

Thanks !
 

Offline dougg

  • Regular Contributor
  • *
  • Posts: 73
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #288 on: March 06, 2020, 04:02:29 pm »
These results were obtained with a recently delivered 121GW running upgraded 2.04 firmware** and an Agilent U8001A power supply. I was pressing the "Output On/Off" button with the voltage set to 20 Volts (1 Amp current limit). When pressing "On" the meter's digital readout blanks in most cases and just after 1 second displays 19.992 volts. The indicator bar goes fully to the right then drops back to the correct position in probably 1/3 second. Once in a while a voltage in the 8 to 9 volt range pops up momentarily before showing the correct voltage. Pretty well the same thing happens in reverse when I press the "Off" button, but it is more likely to show a lower voltage momentarily and the bar indicator drops to zero more or less immediately. The ambient temperature was around 17C .

** Somewhat worryingly the meter was delivered with 2.01 firmware which is listed in the manual as "do no use".
 
The following users thanked this post: ceut

Offline dougg

  • Regular Contributor
  • *
  • Posts: 73
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #289 on: March 06, 2020, 04:09:03 pm »
Plus I would guesstimate a similar 1 second delay while the digital readout was blank when pressing the Range button on the 121GW. One range indicated overflow (the press after showing "20.0"). The bar indication moved promptly.
 
The following users thanked this post: ceut

Offline ceut

  • Regular Contributor
  • *
  • Posts: 124
  • Country: fr
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #290 on: March 06, 2020, 04:54:00 pm »
These results were obtained with a recently delivered 121GW running upgraded 2.04 firmware** and an Agilent U8001A power supply. I was pressing the "Output On/Off" button with the voltage set to 20 Volts (1 Amp current limit). When pressing "On" the meter's digital readout blanks in most cases and just after 1 second displays 19.992 volts. The indicator bar goes fully to the right then drops back to the correct position in probably 1/3 second. Once in a while a voltage in the 8 to 9 volt range pops up momentarily before showing the correct voltage. Pretty well the same thing happens in reverse when I press the "Off" button, but it is more likely to show a lower voltage momentarily and the bar indicator drops to zero more or less immediately. The ambient temperature was around 17C .

** Somewhat worryingly the meter was delivered with 2.01 firmware which is listed in the manual as "do no use".

Plus I would guesstimate a similar 1 second delay while the digital readout was blank when pressing the Range button on the 121GW. One range indicated overflow (the press after showing "20.0"). The bar indication moved promptly.

Thank you !!
So as I have seen in many youtube reviews(I think I saw all  ;D) , the meter is really slow to change its ranges, even in DC Volt mode.
So I will check for future FW update before buying it, and I may buy the U1272A which I find really fantastic but very expensive (I have one which not mine for testing purpose).
I have a good BM257 too, very good but too low in diode&capacitance mode.
 

Offline pppopol

  • Newbie
  • Posts: 6
  • Country: fr
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #291 on: April 29, 2020, 06:41:42 pm »
Hi,

I ask myself this question, why the firmware version 2.04 is not included on Dave's EEVBlog site?
The latest firmware on EEVBlog being 2.02?

Regards
 

Offline Brumby

  • Supporter
  • ****
  • Posts: 12298
  • Country: au
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #292 on: April 30, 2020, 01:56:01 am »
It's on his "To Do" list.

Has been for a while ....  ;D
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #293 on: May 04, 2020, 01:22:28 am »
I ask myself this question, why the firmware version 2.04 is not included on Dave's EEVBlog site?
The latest firmware on EEVBlog being 2.02?

I'm a slack arse.
It's now updated.
 

Offline n3mmr

  • Regular Contributor
  • *
  • Posts: 121
  • Country: se
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #294 on: May 04, 2020, 02:37:34 pm »
I ask myself this question, why the firmware version 2.04 is not included on Dave's EEVBlog site?
The latest firmware on EEVBlog being 2.02?

I'm a slack arse.
It's now updated.

Great self-esteem!  :-DD
 

Offline ralphrmartin

  • Frequent Contributor
  • **
  • Posts: 480
  • Country: gb
    • Me
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #295 on: August 05, 2020, 07:32:12 pm »
The manual on the site doesn't seem to mention 2.04. Has it been updated (if only to tell us what's new in 2.04), and if so can it be made available please?
 

Offline jancumps

  • Supporter
  • ****
  • Posts: 1272
  • Country: be
  • New Low
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #296 on: August 05, 2020, 08:02:11 pm »
ow, c'mon
 

Offline ve2mrx

  • Regular Contributor
  • *
  • Posts: 50
  • Country: ca
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #297 on: August 05, 2020, 09:21:12 pm »
I ask myself this question, why the firmware version 2.04 is not included on Dave's EEVBlog site?
The latest firmware on EEVBlog being 2.02?

I'm a slack arse.
It's now updated.
The manual on the site doesn't seem to mention 2.04. Has it been updated (if only to tell us what's new in 2.04), and if so can it be made available please?
ow, c'mon

Well, I can confirm that on https://www.eevblog.com/product/121gw/, the documentation is NOT updated. The firmware 2.04 is there, however.

In addition, the "Report an issue" link is bad. Dave @EEVblog should look at this when he has a chance!

Martin
 

Offline dzenbo

  • Newbie
  • Posts: 2
  • Country: ua
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #298 on: June 06, 2021, 05:51:12 pm »
The message above is still actual. And now 2.05 firmware is not presented on the site.
Also, can someone present actual stage of further firmware features development? Because I little bit frustrated with my new device and waiting for a new firmware revision which makes this multimeter truly the best one!
 

Offline Kijo

  • Newbie
  • Posts: 6
  • Country: au
Re: EEVBlog 121GW Multimeter Firmware Details
« Reply #299 on: July 05, 2021, 10:11:21 am »
Just got a new 121GW and it is displaying 2.05 - not sure if its a beta release?
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf