« Reply #1225 on: July 26, 2020, 01:29:28 pm »
Thread restored after being accidentally deleted. A bug in SMF means big threads can't be restored using the Restore Topic admin option. But gnif worked his magic on the database manually and it's back!  :clap:
 
The following users thanked this post: newbrain, Andrew McNamara, jack-daniels

Online newbrain

  • Super Contributor
  • ***
  • Posts: 1719
  • Country: se
Re: EEVblog 121GW Multimeter Issues
« Reply #1226 on: July 26, 2020, 02:48:59 pm »
Great!
And please, remember to make it sticky!
Nandemo wa shiranai wa yo, shitteru koto dake.
 

Offline dcac

  • Frequent Contributor
  • **
  • Posts: 339
Re: EEVblog 121GW Multimeter Issues
« Reply #1227 on: July 26, 2020, 03:38:36 pm »
Thanks for getting the thread restored.

One issue I've noticed is my post count seems to have dropped 40 odd posts - but I just checked my posting list and all posts are there and I can access them from the list. So no big deal really.

 

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7992
  • Country: gb
Re: EEVblog 121GW Multimeter Issues
« Reply #1228 on: July 26, 2020, 03:42:41 pm »
Post counters wouldn't have been incremented by manually restoring just the topic.
 

Offline dcac

  • Frequent Contributor
  • **
  • Posts: 339
Re: EEVblog 121GW Multimeter Issues
« Reply #1229 on: July 26, 2020, 03:54:38 pm »
Yeah I figured it was something like that - and like I said all my posts seems intact now - it’s just the counter that’s of.

 

Offline steamfire

  • Newbie
  • Posts: 1
  • Country: us
Re: EEVblog 121GW Multimeter Issues
« Reply #1230 on: July 29, 2020, 02:02:06 am »
Hi,  I have found a bug in the iOS app.  App version 1.1 The date and time don't display properly in the upper right of the app.  See attached pic.  Meter firmware 2.04. 

This is also present in the 3rd party app Meteor version 0.2. 
@DavidWC

1035516-0
« Last Edit: July 29, 2020, 02:07:29 am by steamfire »
 

Offline Kaziq

  • Supporter
  • ****
  • Posts: 24
  • Country: pl
Re: EEVblog 121GW Multimeter Issues
« Reply #1231 on: July 30, 2020, 02:25:23 pm »
No.  The official firmware is proprietary.

The reason given is that there are some "secret sauce" routines that UEI consider commercially valuable.  These would have been developed for their own use and would have been a significant investment - and they are not prepared to make those known to the world.
They could release the part of the source that is not secret (blanking out the valuable parts, and leaving some markers in the code, so people would know what the missing code should be doing).
Some companies do this, and the community makes great use of it (one example I know is Descent 2 game).

For example I would be interested in writing a better software for the meter if I knew I wouldn't screw the calibration data.
 

Offline dcac

  • Frequent Contributor
  • **
  • Posts: 339
Re: EEVblog 121GW Multimeter Issues
« Reply #1232 on: July 31, 2020, 01:47:24 pm »
No.  The official firmware is proprietary.

The reason given is that there are some "secret sauce" routines that UEI consider commercially valuable.  These would have been developed for their own use and would have been a significant investment - and they are not prepared to make those known to the world.
They could release the part of the source that is not secret (blanking out the valuable parts, and leaving some markers in the code, so people would know what the missing code should be doing).
Some companies do this, and the community makes great use of it (one example I know is Descent 2 game).

For example I would be interested in writing a better software for the meter if I knew I wouldn't screw the calibration data.

After messing quite a bit with the disassembled version of the FW - I’m really not sure if any part looks like "secret sauce" or appear particularly ‘developed’. It seems more like an added (often unnecessary) layer of complexity in many routines - even i.e. in the more straight forward ones like setting up the stm32 HAL. So if you were to exclude all of these - I’m afraid not much would be left.

Still the official FW must have taken some man hours to put together. And then there’s all the testing needed to be done. I can understand that they are reluctant to release the source code.

 

Online BU508A

  • Super Contributor
  • ***
  • Posts: 4526
  • Country: de
  • Per aspera ad astra
Re: EEVblog 121GW Multimeter Issues
« Reply #1233 on: August 23, 2020, 10:39:47 am »
Hello,

I'm not sure if this is a known issue to the 121GW. I discovered this issue recently.
I've looked in the 121GW issues thread in the forum but I didn't find anything related. Perhaps I've overlooked something.

What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

Firmware on the 121GW: 2.04
The 121GW is one from the kickstarter campaign (not early bird).

This is my setup:


This is the schematics:
https://www.mikrocontroller.net/attachment/133498/nixie-wandler.pdf

Here is a video which I did to show the issue (ca. 2.5MByte, just remove .zip after downloading):
https://www.eevblog.com/forum/testgear/eevblog-121gw-multimeter-issues/?action=dlattach;attach=1052136

What do you think? Did I do something stupid?

Thanks for your comments etc.
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline HKJ

  • Super Contributor
  • ***
  • Posts: 2904
  • Country: dk
    • Tests
Re: EEVblog 121GW Multimeter Issues
« Reply #1234 on: August 23, 2020, 11:29:26 am »
What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

This may be because you have a DC offset on the signal.
 

Online BU508A

  • Super Contributor
  • ***
  • Posts: 4526
  • Country: de
  • Per aspera ad astra
Re: EEVblog 121GW Multimeter Issues
« Reply #1235 on: August 23, 2020, 04:33:06 pm »
What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

This may be because you have a DC offset on the signal.

Thanks, you are right. I've putted a 4.7µF capacitor between
and the display stays at 17.588kHz without any blinking or something.

Is this a behaviour which should to be expected?
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: EEVblog 121GW Multimeter Issues
« Reply #1236 on: August 23, 2020, 04:35:36 pm »
A lot of meters require zero crossing for accurate frequency measurements.
 
The following users thanked this post: BU508A

Offline HKJ

  • Super Contributor
  • ***
  • Posts: 2904
  • Country: dk
    • Tests
Re: EEVblog 121GW Multimeter Issues
« Reply #1237 on: August 23, 2020, 05:00:57 pm »
Thanks, you are right. I've putted a 4.7µF capacitor between
and the display stays at 17.588kHz without any blinking or something.

Is this a behaviour which should to be expected?

Meters basically comes in 3 variations:
1) Capacitor coupled input where they do not care about DC voltage
2) A circuit that can compensate for a "small" DC offset
3) A circuit that requires the frequency swings around zero.

The 121GW is 2)
Also be aware that frequency input with the selector on AC voltage and on frequency is very different, the later works for low voltage (Overload protection kicks in at a few volts and reduces the input impedance dramatically) and often up into the MHz range. The AC voltage input can maintains high input impedance, but cannot handle high frequencies.
 

Offline Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2384
  • Country: de
Re: EEVblog 121GW Multimeter Issues
« Reply #1238 on: August 23, 2020, 05:16:07 pm »
What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

This may be because you have a DC offset on the signal.

Thanks, you are right. I've put a 4.7µF capacitor between
and the display stays at 17.588kHz without any blinking or something.

Is this a behaviour which should to be expected?

Yes.
That seems to be  the  very same overload problem (AC +DC) we're discussing in the other thread (new eevBlog DMM).
Op1 inside the HY3131 goes in saturation with DC offset, therefore the comparator which produces the count signal, stays at high level.
Overload condition is not detected, because comparator is used for frequency detection.

Frank 
« Last Edit: August 24, 2020, 07:44:44 am by Dr. Frank »
 
The following users thanked this post: BU508A

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37740
  • Country: au
    • EEVblog
Re: EEVblog 121GW Multimeter Issues
« Reply #1239 on: August 24, 2020, 12:04:52 am »
Hello,

I'm not sure if this is a known issue to the 121GW. I discovered this issue recently.
I've looked in the 121GW issues thread in the forum but I didn't find anything related. Perhaps I've overlooked something.

What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

Firmware on the 121GW: 2.04
The 121GW is one from the kickstarter campaign (not early bird).

This is my setup:


This is the schematics:
https://www.mikrocontroller.net/attachment/133498/nixie-wandler.pdf

Here is a video which I did to show the issue (ca. 2.5MByte, just remove .zip after downloading):
https://www.eevblog.com/forum/testgear/eevblog-121gw-multimeter-issues/?action=dlattach;attach=1052136

What do you think? Did I do something stupid?

Thanks for your comments etc.

As others have said, this is likely a DC offset issue.
I just tested the 121GW with a 5Vp-p sine wave and it stop displaying the frequency when the DC offset gets to about 4.8V.
The BM235 has the same issue, and actually stops at about 2.8V offset.
So you should be able to confirm this by swapping the meters you have.
For comparison I tried the new 500 series design Kane meter (see other thread) and it worked fine with up to 7.5V DC offset (the limit of my function gen).
The new $150 class meter also has no problem.
 
The following users thanked this post: BU508A, Fuzzy Star

Offline dcac

  • Frequent Contributor
  • **
  • Posts: 339
Re: EEVblog 121GW Multimeter Issues
« Reply #1240 on: August 25, 2020, 10:05:08 pm »
What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

This may be because you have a DC offset on the signal.

Thanks, you are right. I've put a 4.7µF capacitor between
and the display stays at 17.588kHz without any blinking or something.

Is this a behaviour which should to be expected?

Yes.
That seems to be  the  very same overload problem (AC +DC) we're discussing in the other thread (new eevBlog DMM).
Op1 inside the HY3131 goes in saturation with DC offset, therefore the comparator which produces the count signal, stays at high level.
Overload condition is not detected, because comparator is used for frequency detection.

Frank

I think the problem here might be slightly different. OP1 does not seem to be involved in the Hz mode. And what puzzled me is in BU508A’s video the frequency is actually measured correctly for 5 sec or so.

If the 121gw schematics is correct - the input signal comes in at R RLD and through the U9 switches to RLD1 and then AC coupled with C32 into the CNT counter pin on hy3131.




While U9 will likely clip the signal if it goes beyond 0 - 3.6V. Else D7-D8 will certainly clip it at about -0.6 and 4.2V.

But I’m thinking this is probably a problem also related to non symmetrical pulse shaped signals - which in this case causes C32 to slowly charge to a DC offset level where the signal no longer can trigger the CNT input. According to hy3131 datasheet it requires a high level >2.4V and low level <1.3V, so 1.1V peak to peak for a symmetrical signal.

The 121gw manual does not seem to state minimum voltage required for the Hz mode - but I measured it quite exactly to those 1.1V p-p. And for an asymmetrical signal i.e. square wave with 10% duty cycle it’s about 2.8V p-p.

I would actually recommend using ACV mode if you only want to measure frequency and not the duty cycle and such. In the lowest 5.0000V range the 2nd display for frequency only require about 0.4V p-p to trigger. Though also here a DC offset can/will cause problems.

« Last Edit: August 25, 2020, 10:15:50 pm by dcac »
 
The following users thanked this post: EEVblog

Online BU508A

  • Super Contributor
  • ***
  • Posts: 4526
  • Country: de
  • Per aspera ad astra
Re: EEVblog 121GW Multimeter Issues
« Reply #1241 on: August 25, 2020, 10:58:40 pm »
What I've found:
I'm measuring the frequency in a step-up converter (around 17.5kHz which is to be expected, C3=2.2nF) and after a few seconds the display is showing only zeros. I'm measuring between R4 and the bases of T3 and T4.
On the other side, when I'm measuring the 50Hz hum with the probe in my hand, the digits will stay.

This may be because you have a DC offset on the signal.

Thanks, you are right. I've put a 4.7µF capacitor between
and the display stays at 17.588kHz without any blinking or something.

Is this a behaviour which should to be expected?

Yes.
That seems to be  the  very same overload problem (AC +DC) we're discussing in the other thread (new eevBlog DMM).
Op1 inside the HY3131 goes in saturation with DC offset, therefore the comparator which produces the count signal, stays at high level.
Overload condition is not detected, because comparator is used for frequency detection.

Frank

I think the problem here might be slightly different. OP1 does not seem to be involved in the Hz mode. And what puzzled me is in BU508A’s video the frequency is actually measured correctly for 5 sec or so.

If the 121gw schematics is correct - the input signal comes in at R RLD and through the U9 switches to RLD1 and then AC coupled with C32 into the CNT counter pin on hy3131.




While U9 will likely clip the signal if it goes beyond 0 - 3.6V. Else D7-D8 will certainly clip it at about -0.6 and 4.2V.

But I’m thinking this is probably a problem also related to non symmetrical pulse shaped signals - which in this case causes C32 to slowly charge to a DC offset level where the signal no longer can trigger the CNT input. According to hy3131 datasheet it requires a high level >2.4V and low level <1.3V, so 1.1V peak to peak for a symmetrical signal.

The 121gw manual does not seem to state minimum voltage required for the Hz mode - but I measured it quite exactly to those 1.1V p-p. And for an asymmetrical signal i.e. square wave with 10% duty cycle it’s about 2.8V p-p.

I would actually recommend using ACV mode if you only want to measure frequency and not the duty cycle and such. In the lowest 5.0000V range the 2nd display for frequency only require about 0.4V p-p to trigger. Though also here a DC offset can/will cause problems.

Thanks for looking into it. I did a screenshot of the signal at the exact same point. Looks like it comes with a reasonable DC amount.



« Last Edit: August 25, 2020, 11:01:01 pm by BU508A »
“Chaos is found in greatest abundance wherever order is being sought. It always defeats order, because it is better organized.”            - Terry Pratchett -
 

Offline Kal

  • Newbie
  • Posts: 4
  • Country: ee
Re: EEVblog 121GW Multimeter Issues
« Reply #1242 on: October 12, 2020, 11:43:09 pm »
Has anyone encountered DC VA measurement bug where current and VA values are wrong when current is larger than 2.500A?
Basically few mA over 2.500A makes multimeter value jump from 2.500 straight to 2.657A, display and mem card logging shows the same problem.
Firmware is 2.04
Voltage was about 12.725V and VA reading oscillates between 31,81 and 33,81 when keeping current close to 2.500A.

Seems that all values larger have this offset, so datalogging data is maybe still usable when subracting 157mA from every value bigger than 2.5A. Haven't measured if the offset is indeed fixed or changes over bigger current values.
Error is repeatable after multimeter off/on.
 

Offline movax

  • Contributor
  • Posts: 21
  • Country: aq
Re: EEVblog 121GW Multimeter Issues
« Reply #1243 on: October 15, 2020, 10:08:38 am »
Has anyone encountered DC VA measurement bug where current and VA values are wrong when current is larger than 2.500A?
Basically few mA over 2.500A makes multimeter value jump from 2.500 straight to 2.657A, display and mem card logging shows the same problem.
Firmware is 2.04
Voltage was about 12.725V and VA reading oscillates between 31,81 and 33,81 when keeping current close to 2.500A.

Seems that all values larger have this offset, so datalogging data is maybe still usable when subracting 157mA from every value bigger than 2.5A. Haven't measured if the offset is indeed fixed or changes over bigger current values.
Error is repeatable after multimeter off/on.

Holly shit. You are right.

My multimeter is also doing this:

Code: [Select]
1361,2020/10/15,12:02:37,DCVA,0022.32,VA,DCV,008.401,V,DCA,002.657,A,,
1362,2020/10/15,12:02:38,DCVA,0021.00,VA,DCV,008.401,V,DCA,002.500,A,,

I was testing the 12V 35W halogen bulb, using my Agilent E3642A power supply, using 0-8V/5A output.

I was varying the output voltage on the PSU in few mV increments. The PSU would show change from something like 2.500A to 2.503A, or even less. The display of the multimeter will then jump from 20.99W or 21.00W to about 22.0W, or 22.3W.

The 2.657A figure is absolutely wrong


This puts into question months of my data gathering on one of my project.
« Last Edit: October 15, 2020, 10:18:11 am by movax »
 

Offline e0ne199

  • Regular Contributor
  • *
  • Posts: 131
  • Country: id
Re: EEVblog 121GW Multimeter Issues
« Reply #1244 on: October 15, 2020, 04:21:30 pm »
Has anyone encountered DC VA measurement bug where current and VA values are wrong when current is larger than 2.500A?
Basically few mA over 2.500A makes multimeter value jump from 2.500 straight to 2.657A, display and mem card logging shows the same problem.
Firmware is 2.04
Voltage was about 12.725V and VA reading oscillates between 31,81 and 33,81 when keeping current close to 2.500A.

Seems that all values larger have this offset, so datalogging data is maybe still usable when subracting 157mA from every value bigger than 2.5A. Haven't measured if the offset is indeed fixed or changes over bigger current values.
Error is repeatable after multimeter off/on.

could it be another hardware bug?
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11743
  • Country: us
Re: EEVblog 121GW Multimeter Issues
« Reply #1245 on: October 15, 2020, 05:10:23 pm »
Maybe you followed the instructions in the manual?  If so, that would give some poor results. 


Offline movax

  • Contributor
  • Posts: 21
  • Country: aq
Re: EEVblog 121GW Multimeter Issues
« Reply #1246 on: October 15, 2020, 05:48:48 pm »
Maybe you followed the instructions in the manual?  If so, that would give some poor results. 



This is totally unrelated.
 

Offline J-R

  • Frequent Contributor
  • **
  • Posts: 975
  • Country: us
Re: EEVblog 121GW Multimeter Issues
« Reply #1247 on: October 16, 2020, 02:48:06 am »
I was able to reproduce the issue as well.  I ran some quick tests with a 66332A at around 2.5V, 4V, 5V, 12V.  I did not notice the issue at 2.5V?
 

Offline movax

  • Contributor
  • Posts: 21
  • Country: aq
Re: EEVblog 121GW Multimeter Issues
« Reply #1248 on: October 16, 2020, 09:09:57 am »
I was able to reproduce the issue as well.  I ran some quick tests with a 66332A at around 2.5V, 4V, 5V, 12V.  I did not notice the issue at 2.5V?


2.5V volts? We got issues with current reading, not voltage. It only shows up in VA mode - both on display without data logging and with data logging. The current reads and data logs correctly in current (A) mode.
 

Offline bicycleguy

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: EEVblog 121GW Multimeter Issues
« Reply #1249 on: October 16, 2020, 04:13:47 pm »
I was able to reproduce the issue as well.  I ran some quick tests with a 66332A at around 2.5V, 4V, 5V, 12V.  I did not notice the issue at 2.5V?


2.5V volts? We got issues with current reading, not voltage. It only shows up in VA mode - both on display without data logging and with data logging. The current reads and data logs correctly in current (A) mode.
I think that J-R is pointing out that the current reading problem in VA mode also depends on the Voltage your around, even a  constant voltage.  Haven't spent the time to look at it enough but it also may have something to do with what range (voltage & current) you are on.  Definitely another small but irritating bug.
 