I have a new Rigol DS1054z that isn't correctly displaying / reading voltage.  Am I doing something wrong?

I hooked up the output of a TEMS device (one of those zap zap medical things) to my DS1054z input.  For some reason it says the channel is receiving more than 30 volts.  However, my Fluke 87V says:  Mean voltage = 2.5 VAC, Min voltage = 1.560 VAC, Max voltage = 3.457 VAC.  The current is very low (on the order of 10^-2 A).  The oscilloscope leads were on the 1X setting.

I trust the readings from my Fluke.  I literally trust my life to that meter. 

Is there a setting in the Rigol that is scaling the potential?  I've scaled the *.csv data set to work with it in R - but I'd much rather just get good information without any scaling!  In fact the *.csv output has a high value of "284," which is really confusing.

Thank you very much for your input! 

Yay EEVBlog!!

E   
:scared:
 

Online coppice

  • Super Contributor
  • ***
  • Posts: 8646
  • Country: gb
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #1 on: October 06, 2015, 03:30:11 am »
Since the scope is saying about 10x what you expect it to, I guess you only changed the switch in the probe to x1, and didn't change the software setting on the scope to match this.

You would be better off changing the probe to x10. Unless you want to look at a signal so small you need the added gain of a x1 probe, x10 is almost always the better choice. To see why, look through Dave's videos. He has one on this topic.
 

Offline EjohnsonTopic starter

  • Newbie
  • Posts: 5
  • Country: us
  • Mathematician / Musician / EE Geek : )
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #2 on: October 06, 2015, 03:31:13 am »
Ah!  Fantastic!  I'll try that. 

Thank you for the reply.  : )
 

Offline alsetalokin4017

  • Super Contributor
  • ***
  • Posts: 2055
  • Country: us
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #3 on: October 06, 2015, 04:01:46 am »
That's right. The DS1054Z cannot auto-sense the probe's attenuation setting, and defaults to the 10x setting in the Channel menu. For most normal use, one would use the Probe's switch set to 10x and also in the Channel menu, one selects 10x for the Probe attenuation.

The Rigol's Channel>Probe menu has a big selection of possible probe settings.
The easiest person to fool is yourself. -- Richard Feynman
 

Online coppice

  • Super Contributor
  • ***
  • Posts: 8646
  • Country: gb
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #4 on: October 06, 2015, 04:30:14 am »
The DS1054Z cannot auto-sense the probe's attenuation setting, and defaults to the 10x setting in the Channel menu.
Even if a scope does support auto-sense you need to check very carefully if its working. Typically you need to be using exactly the right model of probe with the right model of scope for it to function.
 

Offline SteveLy

  • Regular Contributor
  • *
  • Posts: 220
  • Country: au
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #5 on: October 06, 2015, 07:05:45 am »
In addition to getting the 1x vs 10x probe setting right on the scope, keep in mind that scopes typically report peak or peak-to-peak voltage, while DMMs show RMS; 2*V(RMS) = 1.414 * V(peak) = 0.707 * V(peak-to-peak).
« Last Edit: October 06, 2015, 08:46:27 am by SteveLy »
 

Offline nugglix

  • Regular Contributor
  • *
  • Posts: 209
  • Country: de
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #6 on: October 06, 2015, 07:12:39 am »
In addition to getting the 1x vs 10x probe setting right on the scope, keep in mind that scopes typically report peak or peak-to-peak voltage, while DMMs show RMS; V(RMS) = 1.414 * V(peak) = 0.707 * V(peak-to-peak).

*cough*

See https://en.wikipedia.org/wiki/Root_mean_square
 

Online Performa01

  • Super Contributor
  • ***
  • Posts: 1644
  • Country: at
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #7 on: October 06, 2015, 07:33:24 am »
In addition to getting the 1x vs 10x probe setting right on the scope, keep in mind that scopes typically report peak or peak-to-peak voltage ...

Scopes typically 'report' nothing at all by default. The user deliberately adds the measurements of interest and in case of voltage has the opportunity to select Vpp, Vmean and/or Vrms from the measurement menu.

One should keep in mind that the ADC on a typical scope is only 8 bits. Therefore the resolution and in turn the accuracy of measurements is rather poor, particularly when compared to a decent DMM. On the other hand, a DSO allows measurements at frequencies, where any DMM has long given up...
 

Offline SteveLy

  • Regular Contributor
  • *
  • Posts: 220
  • Country: au
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #8 on: October 06, 2015, 08:44:44 am »
In addition to getting the 1x vs 10x probe setting right on the scope, keep in mind that scopes typically report peak or peak-to-peak voltage, while DMMs show RMS; V(RMS) = 1.414 * V(peak) = 0.707 * V(peak-to-peak).

*cough*

See https://en.wikipedia.org/wiki/Root_mean_square
Good catch. 2*V(pp) = V(p) = sqrt(2)*V(rms)
 

Online Performa01

  • Super Contributor
  • ***
  • Posts: 1644
  • Country: at
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #9 on: October 06, 2015, 09:14:15 am »
Good catch. 2*V(pp) = V(p) = sqrt(2)*V(rms)

Not quite correct, but you slowly get there...

Try it this way:

Vpp = 2 * Vp;
Vp   = sqrt(2) * Vrms;  --> FOR PURE SINE WAVES ONLY!
 

Offline SteveLy

  • Regular Contributor
  • *
  • Posts: 220
  • Country: au
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #10 on: October 06, 2015, 10:52:11 am »
LOL! Clearly not my day for tackling advanced mathematics like multiplying and dividing by 2, let alone square root of 2.
« Last Edit: October 06, 2015, 10:54:25 am by SteveLy »
 

Online Performa01

  • Super Contributor
  • ***
  • Posts: 1644
  • Country: at
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #11 on: October 06, 2015, 11:09:41 am »
LOL! Clearly not my day for tackling advanced mathematics like multiplying and dividing by 2, let alone square root of 2.

That's the beauty of the community - with combined efforts we're getting it right, eventually! ;)
 

Offline jwm_

  • Frequent Contributor
  • **
  • Posts: 319
  • Country: us
    • Not A Number
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #12 on: October 06, 2015, 11:14:11 am »
That's right. The DS1054Z cannot auto-sense the probe's attenuation setting, and defaults to the 10x setting in the Channel menu. For most normal use, one would use the Probe's switch set to 10x and also in the Channel menu, one selects 10x for the Probe attenuation.

The Rigol's Channel>Probe menu has a big selection of possible probe settings.

I always thought a good feature for a scope would be to use the probe calibration port to set the probe multiplier properly. so it would be probe independent. Just take the probe, stick it to the calibration point and press and the scope will find the channel with the cal signal on it and see exactly what it is being attenuated by and set the multiplier accordingly.

Online Fungus

  • Super Contributor
  • ***
  • Posts: 16659
  • Country: 00
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #13 on: October 06, 2015, 11:31:43 am »
I always thought a good feature for a scope would be to use the probe calibration port to set the probe multiplier properly. so it would be probe independent. Just take the probe, stick it to the calibration point and press and the scope will find the channel with the cal signal on it and see exactly what it is being attenuated by and set the multiplier accordingly.

??

Lots of 'scopes already have probes that tell the 'scope what the attenuation is automatically. Just not the DS1054Z.

 

Offline jwm_

  • Frequent Contributor
  • **
  • Posts: 319
  • Country: us
    • Not A Number
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #14 on: October 06, 2015, 11:42:50 am »
I always thought a good feature for a scope would be to use the probe calibration port to set the probe multiplier properly. so it would be probe independent. Just take the probe, stick it to the calibration point and press and the scope will find the channel with the cal signal on it and see exactly what it is being attenuated by and set the multiplier accordingly.

??

Lots of 'scopes already have probes that tell the 'scope what the attenuation is automatically. Just not the DS1054Z.

Yes, but only with probes specifically designed for that scope generally from the same manufacturer.

This method would work with any probes (including homemade ones) and be a pure software thing so no added expense in fancy BNC plugs that have extra sensing contacts and can be added to inexpensive scopes.


Offline Salas

  • Frequent Contributor
  • **
  • Posts: 292
  • Country: gr
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #15 on: October 06, 2015, 11:55:46 am »
As an aside lets mention that a scope's probe Cal signal is usually a non zero crossing pulse i.e. it has positive DC component. A scope set to DC coupling will measure its RMS correctly when even a True RMS DMM without DC+AC True RMS ability will measure it wrongly.
 

Offline EjohnsonTopic starter

  • Newbie
  • Posts: 5
  • Country: us
  • Mathematician / Musician / EE Geek : )
Re: Incorrect voltage readings -- Rigol DS1054z Oscilloscope
« Reply #16 on: October 11, 2015, 05:30:10 am »
This is a great site and community.  I really appreciate all of the great comments that came from my question!

Thanks again everyone!

 :-+
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf