Just a bit surprised to see such, err 'dedication' in the Test Equipment section rather than Metrology.
Quote from: Gyro on March 17, 2019, 11:31:07 amJust a bit surprised to see such, err 'dedication' in the Test Equipment section rather than Metrology. The problem only exists because he has "something with higher resolution and accuracy".If he didn't have that fancy HP meter he'd be perfectly happy with the handheld.Moral: Ignorance is bliss.
Quote from: Fungus on March 17, 2019, 11:54:22 amThe problem only exists because he has "something with higher resolution and accuracy".But is it 'actually' more accurate and to what references. A better resolved inaccurate reading is useless
The problem only exists because he has "something with higher resolution and accuracy".
Quote from: beanflying on March 17, 2019, 11:59:39 amQuote from: Fungus on March 17, 2019, 11:54:22 amThe problem only exists because he has "something with higher resolution and accuracy".But is it 'actually' more accurate and to what references. A better resolved inaccurate reading is useless Joy comes from creating a closed system where everything reads the same value.
Suggesting anyone should re calibrate Meters for the sake of having them agree with each other is pointless
If you only have calibration points at 50 and 500mV and if, as Dave says, the other ranges are taken care of by the internal dividers, then the calibration lab isn't going to be able to do anything about the tolerance of those dividers. It sounds as if they are within the manufacturer's spec anyway.