Author Topic: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules  (Read 10410 times)

0 Members and 1 Guest are viewing this topic.

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« on: October 04, 2015, 03:44:17 am »
You can watch it here: [32 Minutes]
youtu.be/dAnMCgdRAHk

More vidoes at The Signal Path:
http://www.TheSignalPath.com

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3743
  • Country: ca
  • Living the Dream
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #1 on: October 04, 2015, 05:36:50 am »
Another great video Shahriar - thank you for posting it. Do you have Pooch helping when you solder stuff, an extra set of paws can be pretty useful.

Shame the DMM didn't come back to life. The memory chip you replaced is ram I assume. Have you spent any time to see if the module is killing the power supply or is it the data bus itself? You've likely done more troubleshooting off camera - maybe walk us through what you have all done.
VE7FM
 

Offline bktemp

  • Super Contributor
  • ***
  • Posts: 1616
  • Country: de
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #2 on: October 04, 2015, 06:22:33 am »
In the video you said the missing component looked similar to the one on the other side. And this looks like a Flash memory, most likely containing some firmware, or the FPGA bitstream.
If you replace it with a new one, the firmware is still missing. If this firmware controls and initializes the PCI controller this could explain why the bus is not workling.
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #3 on: October 04, 2015, 04:12:18 pm »
In the video you said the missing component looked similar to the one on the other side. And this looks like a Flash memory, most likely containing some firmware, or the FPGA bitstream.
If you replace it with a new one, the firmware is still missing. If this firmware controls and initializes the PCI controller this could explain why the bus is not workling.

It is possible that some firmware is missing, unfortunately I don't have the original firmware...

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #4 on: October 04, 2015, 04:33:21 pm »
Perhaps you can read firmware from your other module and try flash that one. There is little chance PXI side of things is unified, so firmware from other module would work.
Or maybe NI can send you the correct one?
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline lukier

  • Supporter
  • ****
  • Posts: 634
  • Country: pl
    • Homepage
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #5 on: October 04, 2015, 04:45:50 pm »
Do you have high-res photos of these cards by any chance?

I'm curious what ADC and DACs they use on the PXI-4461. 24 bit DAC is actually something spectacular, so I strongly suspect they use cheapo audio AD/DA! These things have horrible DC specs, so often they don't even mention INL/DNL and ENOB in the datasheets (not to mention gain/offset temperature drifts).
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #6 on: October 04, 2015, 10:35:19 pm »
Perhaps you can read firmware from your other module and try flash that one. There is little chance PXI side of things is unified, so firmware from other module would work.
Or maybe NI can send you the correct one?

I am not sure if this is easy to do... My other PXI modules don't use the same PCI bus IC...

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3743
  • Country: ca
  • Living the Dream
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #7 on: October 05, 2015, 02:27:14 am »
Well if it is a flash chip I'd say you're much stuck unless you can find another DMM module and remove/read the flash chips to copy the images. I would have thought a card with a blank flash would not cause problems on the bus. Perhaps it reads some code from the first flash and then gets stuck looking for the second one.
VE7FM
 

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2710
  • Country: us
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #8 on: October 05, 2015, 03:19:18 am »
What is the voltage reference in the 4071?
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #9 on: October 05, 2015, 05:28:15 pm »
What is the voltage reference in the 4071?

LTZ1000

Offline guido

  • Regular Contributor
  • *
  • Posts: 207
  • Country: nl
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #10 on: October 05, 2015, 05:32:11 pm »
I would have done the repair of the relay module differently:
  • put the connector on the third, unused, section
  • connect the transistors (or FETs, can't remember) to the equivalent transistor pads going to the connector
Connector mounted sturdy on the pcb and just six (or seven with gnd?) easy connections.

Just my 2 cents  :blah: :P

Edit: too bad that was not possible. Looked that way but i guess you checked.
« Last Edit: October 05, 2015, 08:29:01 pm by guido »
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #11 on: October 05, 2015, 06:01:57 pm »
I would have done the repair of the relay module differently:
  • put the connector on the third, unused, section
  • connect the transistors (or FETs, can't remember) to the equivalent transistor pads going to the connector
Connector mounted sturdy on the pcb and just six (or seven with gnd?) easy connections.
Just my 2 cents  :blah: :P

That was not possible. The empty connectors aren't actually empty. The connections go to the other MOS switches which are already occupied. The connections are designed with some other type of microwave switches in mind.

Offline Vgkid

  • Super Contributor
  • ***
  • Posts: 2710
  • Country: us
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #12 on: October 05, 2015, 07:10:52 pm »
Snip
Thanks, that is what I assumed it was.
If you own any North Hills Electronics gear, message me. L&N Fan
 

Offline AF6LJ

  • Supporter
  • ****
  • Posts: 2902
  • Country: us
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #13 on: October 05, 2015, 07:52:40 pm »
Good video.
Good luck with the DMM module, but I am not holding my breath.
Sue AF6LJ
 

Offline BFX

  • Frequent Contributor
  • **
  • Posts: 376
  • Country: sk
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #14 on: October 05, 2015, 10:39:41 pm »
Thank you Shahriar for another instructive video. It's very helpful for lot of people here I think  :-+
Did you try thermal camera scanning with power of this board ? Only to be sure if isn't bug somewhere in power lines.
 

Offline free_electron

  • Super Contributor
  • ***
  • Posts: 8515
  • Country: us
    • SiliconValleyGarage
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #15 on: October 05, 2015, 11:05:24 pm »
PXI ? SET IT ON FIRE !!

#$%(@*#$% overpriced craptional instrucrap crap.
Professional Electron Wrangler.
Any comments, or points of view expressed, are my own and not endorsed , induced or compensated by my employer(s).
 

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7990
  • Country: gb
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #16 on: October 06, 2015, 12:07:35 pm »
Another showing of the rare and underappreciated oscillocat ;D
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 955
  • Country: us
    • The Signal Path Video Blog
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #17 on: October 06, 2015, 08:32:30 pm »
PXI ? SET IT ON FIRE !!

#$%(@*#$% overpriced craptional instrucrap crap.

Haha! Yeah... that DMM is listed for $3.3k.

Offline rx8pilot

  • Super Contributor
  • ***
  • Posts: 3634
  • Country: us
  • If you want more money, be more valuable.
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #18 on: October 06, 2015, 08:39:46 pm »
PXI ? SET IT ON FIRE !!

#$%(@*#$% overpriced craptional instrucrap crap.
I like PXI and LabVIEW.

Sent from my SAMSUNG-SM-G890A using Tapatalk

Factory400 - the worlds smallest factory. https://www.youtube.com/c/Factory400
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #19 on: October 06, 2015, 08:41:18 pm »
It's so overengineered! That microwave switch module had what - an FPGA, a PCI interface chip, like three memories, some unidentified SSOP, some microcontroller-looking thing... did I miss anything? All to switch relays! Even if they want it to be PCI only, it doesn't take that much to slap something onto a PCI bus, pretty much a lone FPGA with no external memory other than the main flash can do that. The FPGA can also generate the relay control signals directly, store the actuation counts back into the flash, etc...
« Last Edit: October 06, 2015, 08:44:53 pm by c4757p »
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline rx8pilot

  • Super Contributor
  • ***
  • Posts: 3634
  • Country: us
  • If you want more money, be more valuable.
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #20 on: October 06, 2015, 11:19:31 pm »
It's so overengineered!

I won't argue that. I will say that as I have been researching and learning a lot about automated T&M for design validation and production and National Instruments has been a great thing. If you look at it from the perspective of someone like myself that will use LabView with PXI hardware and stand-alone T&M - the integration is hard to beat. I can rather quickly setup and automate a test that was taking 30min to do and required me or some other qualified engineer to do and smash it down to 1 minute and require next to no skill at all. The time savings of automated and integrated tests that I can have housed in a single chassis is enough to justify the cost of a PXI system.

In my previous careers, I have always 'over purchased' what many called over engineered. It made me and my companies the quickest most reliable in our field - and the most profitable. I am a relative newbie in electronics design and manufacturing, but I can already tell that buying the high-end pays for itself. I have not yet purchased my PXI system, but I am confident the $ that goes in will be $$$ that comes out.

It may be never that I need a 26Ghz relay on the PXI system.....
Factory400 - the worlds smallest factory. https://www.youtube.com/c/Factory400
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #21 on: October 06, 2015, 11:23:01 pm »
And yet the exact same functionality can be implemented with significantly less faff. What does any of that have to do with whether or not the modules are overengineered?

I'm not saying anything against the PXI system existing. It looks quite nice, I'd love to have one. It just looks like they made it cost about 10x what it has to.
« Last Edit: October 06, 2015, 11:24:46 pm by c4757p »
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline rx8pilot

  • Super Contributor
  • ***
  • Posts: 3634
  • Country: us
  • If you want more money, be more valuable.
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #22 on: October 06, 2015, 11:49:02 pm »
The over-engineered part raises the price of the modules. While I wish they were less expensive - I voted to pay extra to have a single, compact, totally integrated test system that is easily operable by a low-level  technician. Having an immobile bench with a stack if 8 instruments all connected to a laptop is not nearly as friendly.

You say....
It's so overengineered!

The most glaring problem with over-engineered T&M is that the price makes it scary. My point was - who cares if it is over engineered if it solves all of my problems and I don't have to worry about it. The problem with the traditional route is the significant increase in labor costs since it needs a much better technician or engineer to operate. The over-engineering buys me a solution that cost very little to operate every day. I can put a PXI chassis on a cart with a bed of nails, roll it over to the end of the P&P line and have the operator running testing and programming of the PCB's as they come out. At that point I don't care if the relay card is controlled by FPGA's  - I am happy that I don't have a $75k/year technician/engineer testing PCB's.

Hopefully that makes sense.
Factory400 - the worlds smallest factory. https://www.youtube.com/c/Factory400
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #23 on: October 06, 2015, 11:51:53 pm »
Sure, of course, it can be overlooked, and clearly is, by many people. Just saying, all that stuff to drive some relays is a bit comical.
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline rx8pilot

  • Super Contributor
  • ***
  • Posts: 3634
  • Country: us
  • If you want more money, be more valuable.
Re: Video Repair of PXI-2596, PXI-4461 & PXI-4071 Modules
« Reply #24 on: October 06, 2015, 11:58:27 pm »
I agree.

There is a company that I used to buy very expensive video terminal equipment from. They had many general purpose designs that could be used for many purposes. What you ended up with was a product that was totally appropriate if it was the most sophisticated in the series. When you look at the entry level configuration - it was hilarious. Sometimes a few FPGA's were doing what could be done with a single microcontroller. The I/O architecture required all three FPGA to be doing a teeny tiny bit of work.

The justification was the the cost of the extra components was less than the cost of whole new dedicated and optimized design. Volume was rather low, so I believed that explanation.
Factory400 - the worlds smallest factory. https://www.youtube.com/c/Factory400
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo