In this episode Shahriar takes a close look at the Analog Arts USB based mixed-domain instrument. The SL987 offers an oscilloscope, arbitrary waveform generator, FFT-based spectrum analyzer, pattern generator and logic analyzer in a compact form factor.
The teardown of the unit is presented along with the datasheet of every major component on the PCB. Some design limitations and concerns are presented along with basic performance verifications.

Watch the video here: [46 Minutes]
http://youtu.be/F-1uA4Iwce8

More videos at The Signal Path:
http://www.TheSignalPath.com
« Last Edit: February 10, 2015, 12:52:12 am by Hugoneus »
 

Offline photon

  • Regular Contributor
  • *
  • Posts: 234
  • Country: us
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #1 on: February 09, 2015, 08:34:30 pm »
Thanks for the post and teardown. The USB connector with host ports on each end is not possilble, AFAIK. I wonder what happened there? It could be that their circuit board had mistakenly put a host port connector instead of a device port connector. If someone thinks they can use such a connector with 2 host ports to connect 2 PC's via USB, then this would not work.
 

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7992
  • Country: gb
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #2 on: February 09, 2015, 08:37:22 pm »
That.. is pretty horrific.
 

Offline photon

  • Regular Contributor
  • *
  • Posts: 234
  • Country: us
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #3 on: February 09, 2015, 09:44:57 pm »
I think it interesting that USB has spec'd a new connector, call TYPE-C, that will be universal, in the sense that there will be a single connector for both host and device. Also, it will have no up or down polarity in the plug, will support all versions of USB plus allow support for other bus protocols. It will negotiate all these issues, plus who is supplying power and how much, at plug-in time. The spec allows up to 100 watts of power!
 

Offline electr_peter

  • Supporter
  • ****
  • Posts: 1301
  • Country: lt
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #4 on: February 09, 2015, 10:13:17 pm »
Claimed "1GHz BW", wobbly BNC jacks, cross-talk all over the place, massive undersampling for BW, inadequate shielding, mysterious DC offsets, digital or analog modes only, ... :palm:
At least SW side is working without hiccups.

This device is well suited for lower (say 10-40 MHz) freq probing/logic probing/signal gen, but at this price point market is dense. Maybe manufacturer tried to amplify best possible scenario for specs to stand out. I think that built-in signal gen + scope is very useful for frequency response testing.
Comparison with bench scopes would be not be reasonable, because of different amount of processing/memory on main board. As mentioned in the video, DS1000Z series are hard to compete with.

As for USB connector, I think it is very reasonable choice. It is easy to connect, holds contacts well and gets good support from PCB. USB B is similar or a bit worse in some aspects, micro/mini are too weak in comparison. The only disadvantage - cable is harder to source.
« Last Edit: February 09, 2015, 10:17:45 pm by electr_peter »
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 956
  • Country: us
    • The Signal Path Video Blog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #5 on: February 09, 2015, 10:20:52 pm »
Unfortunately as you can imagine the company is not very happy with my review. As I mentioned I gave it the benefit of the doubt with regards to the board (assuming that it was a prototype). And I mentioned that for lower frequency inputs its performance is reasonable.

However, I would have betrayed my commitment to science and engineering if I didn't speak about its problems.  :-//

Offline Monkeh

  • Super Contributor
  • ***
  • Posts: 7992
  • Country: gb
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #6 on: February 09, 2015, 10:23:52 pm »
As for USB connector, I think it is very reasonable choice. It is easy to connect, holds contacts well and gets good support from PCB. USB B is similar or a bit worse in some aspects, micro/mini are too weak in comparison. The only disadvantage - cable is harder to source.

It's a stupid choice, it's non-compliant and the cables are, as a result, not standard.

Unfortunately as you can imagine the company is not very happy with my review. As I mentioned I gave it the benefit of the doubt with regards to the board (assuming that it was a prototype). And I mentioned that for lower frequency inputs its performance is reasonable.

However, I would have betrayed my commitment to science and engineering if I didn't speak about its problems.  :-//

You were very kind to it. Much kinder than it deserved. They should have known better than to send something like that to someone who actually knows what they're talking about.
 

Offline AF6LJ

  • Supporter
  • ****
  • Posts: 2902
  • Country: us
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #7 on: February 09, 2015, 10:35:07 pm »
 :-[ Very Scary.
Sue AF6LJ
 

Offline mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13736
  • Country: gb
    • Mike's Electric Stuff
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #8 on: February 09, 2015, 10:36:16 pm »
Unfortunately as you can imagine the company is not very happy with my review.
Well if they shipped you what looks like little more than a prototype what did they expect...?

I have seen USB A's abused like this on a couple of occasions, and USB A-A cables are fairly available, but it's an unnecessary annoyance.
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline electr_peter

  • Supporter
  • ****
  • Posts: 1301
  • Country: lt
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #9 on: February 09, 2015, 10:48:52 pm »
As for USB connector, I think it is very reasonable choice. It is easy to connect, holds contacts well and gets good support from PCB. USB B is similar or a bit worse in some aspects, micro/mini are too weak in comparison. The only disadvantage - cable is harder to source.

It's a stupid choice, it's non-compliant and the cables are, as a result, not standard.
It is non-compliant (what will happen when two hosts are connected with such cable? :-BROKE) and it should be marked with bright colors to inform general public about it. However, it is not that bad if only connector is concerned - one can do much worse. It is better than some non-standard connector for USB port or something crazy like plug/plug (male to male) IEC mains connector...

Thanks to Shahriar for such complete review of device as it is. Detailed explanations of schematics, comments on parts spec, fault finding, real performance testing and comparison to specs :-+ This is real journalism of technical topics.
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #10 on: February 09, 2015, 11:45:30 pm »
Damn you're a nice dude. This thing's a total heap of garbage, they're so lucky they got the kind review you gave it.

It needs to be reviewed with a hammer.
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37730
  • Country: au
    • EEVblog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #11 on: February 09, 2015, 11:47:21 pm »
There isn't much Art inside that! Yuck!
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 956
  • Country: us
    • The Signal Path Video Blog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #12 on: February 09, 2015, 11:48:31 pm »
There isn't much Art inside that! Yuck!

It would have been interesting to hear what you thought about it. As I said, I feel bad about the review and I have given them a chance to address the issues and get back to me.

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37730
  • Country: au
    • EEVblog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #13 on: February 09, 2015, 11:50:05 pm »
It needs to be reviewed with a hammer.

A Widlerizer  hammer!
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #14 on: February 09, 2015, 11:52:46 pm »
However, I would have betrayed my commitment to science and engineering if I didn't speak about its problems.  :-//

:clap:

I understand why it might disappoint you that they're Unimpressed, but dude, you scored a big one for integrity. You were honest about this thing, fair and balanced in every way - you gave it a chance when it deserved one, told the truth when the truth needed to be told, and as much as possible let the viewer reach his own decision. If they think the problem is your review and not their product, they have a serious disconnect with reality.

There's a reason your equipment reviews are my favorite of all of them. Analog Arts might not be happy, but I definitely thank you.
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 956
  • Country: us
    • The Signal Path Video Blog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #15 on: February 09, 2015, 11:55:34 pm »
However, I would have betrayed my commitment to science and engineering if I didn't speak about its problems.  :-//
:clap:
I understand why it might disappoint you that they're Unimpressed, but dude, you scored a big one for integrity. You were honest about this thing, fair and balanced in every way - you gave it a chance when it deserved one, told the truth when the truth needed to be told, and as much as possible let the viewer reach his own decision. If they think the problem is your review and not their product, they have a serious disconnect with reality.
There's a reason your equipment reviews are my favorite of all of them. Analog Arts might not be happy, but I definitely thank you.

I appreciate that.

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37730
  • Country: au
    • EEVblog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #16 on: February 09, 2015, 11:55:46 pm »
It would have been interesting to hear what you thought about it. As I said, I feel bad about the review and I have given them a chance to address the issues and get back to me.

I only briefly watch the inside teardown bit, but that was enough. Seeing that I doubt I would have even bothered continuing with any review.
For a company that claims:
Quote
Analog Arts is a private high-technology company headquartered in Massachusetts, USA. Formerly known as Edge Technology, the company was founded in 1992. It specializes in high performance digital and analog signal processing and advanced test measurement systems. Analog Arts products range from custom board level solutions to complete stand-alone equipment. Throughout its history, Analog Arts has been committed to offer reliable and cost effective solutions to challenging ideas. Relying on a pool of highly experienced and motivated scientists and engineers, Analog Arts has become a world resource for outstanding engineering. As a recognized leader in signal processing and measurement equipment, Analog Arts has provided designs, products, and services to scientific, military, and commercial companies throughout the world. The company designs, develops, tests, purchases parts, manufactures, and packages all products in the United States.
We customize our technology for particular applications. With features like, 140 dB of dynamic performance, 100 GHz sampling rates, 28 bits of resolution, and super high bandwidths, our solutions have satisfied a wide range of applications including radar, communication, automotive, medical, and satellite.
You have to wonder how they could be this incompetent?
$1260 and the best and most expensive product in their range? Yikes.
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37730
  • Country: au
    • EEVblog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #17 on: February 10, 2015, 12:00:02 am »
I understand why it might disappoint you that they're Unimpressed, but dude, you scored a big one for integrity. You were honest about this thing, fair and balanced in every way - you gave it a chance when it deserved one, told the truth when the truth needed to be told, and as much as possible let the viewer reach his own decision.

The funny thing is I've been criticised for doing that  ;D
I've found that people want a strong opinion one way or the other, Yay or Nay.
If I did anything less than toss it across the room, I'd get hate mail and accused of having sold out  ::)
 

Offline c4757p

  • Super Contributor
  • ***
  • Posts: 7799
  • Country: us
  • adieu
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #18 on: February 10, 2015, 12:02:53 am »
I understand why it might disappoint you that they're Unimpressed, but dude, you scored a big one for integrity. You were honest about this thing, fair and balanced in every way - you gave it a chance when it deserved one, told the truth when the truth needed to be told, and as much as possible let the viewer reach his own decision.

The funny thing is I've been criticised for doing that  ;D
I've found that people want a strong opinion one way or the other, Yay or Nay.
If I did anything less than toss it across the room, I'd get hate mail and accused of having sold out  ::)

People expect you to be mean. ;)
No longer active here - try the IRC channel if you just can't be without me :)
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 37730
  • Country: au
    • EEVblog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #19 on: February 10, 2015, 12:05:40 am »
People expect you to be mean. ;)

I suspect they confuse meanness with brutal untactful honesty  ;D

Ok, I'm mean  >:D
 

Offline Lukas

  • Frequent Contributor
  • **
  • Posts: 412
  • Country: de
    • carrotIndustries.net
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #20 on: February 10, 2015, 12:26:58 am »
1GHz with a 100MSa/s ADC is not completely impossible. You may use a comparator and a TDC to determine the exact triggering instant and then accumulate over multiple acquisitions. HP did that in the 54600 series and manged to get >100MHz bandwidth out of a 20MSa/s ADC.
 

Offline HugoneusTopic starter

  • Frequent Contributor
  • **
  • Posts: 956
  • Country: us
    • The Signal Path Video Blog
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #21 on: February 10, 2015, 12:39:41 am »
1GHz with a 100MSa/s ADC is not completely impossible. You may use a comparator and a TDC to determine the exact triggering instant and then accumulate over multiple acquisitions. HP did that in the 54600 series and manged to get >100MHz bandwidth out of a 20MSa/s ADC.

As I mentioned in the video, such a system requires either a coherent precision time-base signal or an entirely different architecture. The problem arises from the fact that the instrument makes mistakes on computing the actual frequency of the signal being applied once it is above Nyquist. And it can only do this on periodic signals (and if the period fits within the captured memory). It is essentially trying to reconstruct the aliased signal from multiple periods.
« Last Edit: February 10, 2015, 02:36:47 am by Hugoneus »
 

Offline photon

  • Regular Contributor
  • *
  • Posts: 234
  • Country: us
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #22 on: February 10, 2015, 02:37:20 am »
As for USB connector, I think it is very reasonable choice. It is easy to connect, holds contacts well and gets good support from PCB. USB B is similar or a bit worse in some aspects, micro/mini are too weak in comparison. The only disadvantage - cable is harder to source.

It's a stupid choice, it's non-compliant and the cables are, as a result, not standard.
It is non-compliant (what will happen when two hosts are connected with such cable? :-BROKE)

Yes, it is indeed broke. Since a USB hosts supplies power on VBUS, having 2 hosts connected together, type-A to type-A, will short their supplies and can cause damage. On the other hand, I agree that if one of the 2 ports is a device, then this problem does not occur since a device does not supply power but rather consumes it. The only type-A to type-A connector that I would trust would be one where the cable contains a small circuit board that bridges the two ports. Such cables exist and are handy to network 2 PC's using USB, but they require a special driver and are more expensive.
 

Offline Lukas

  • Frequent Contributor
  • **
  • Posts: 412
  • Country: de
    • carrotIndustries.net
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #23 on: February 10, 2015, 07:17:30 am »
1GHz with a 100MSa/s ADC is not completely impossible. You may use a comparator and a TDC to determine the exact triggering instant and then accumulate over multiple acquisitions. HP did that in the 54600 series and manged to get >100MHz bandwidth out of a 20MSa/s ADC.

As I mentioned in the video, such a system requires either a coherent precision time-base signal or an entirely different architecture. The problem arises from the fact that the instrument makes mistakes on computing the actual frequency of the signal being applied once it is above Nyquist. And it can only do this on periodic signals (and if the period fits within the captured memory). It is essentially trying to reconstruct the aliased signal from multiple periods.
I guess you're thinking of sampling oscilloscopes like the DCA from Keysight? Random sampling oscilloscopes like the 54600 don't require a coherent clock/trigger, since they determine the clock-to-trigger relationship on each trigger event using a TDC.
 

Offline nfmax

  • Super Contributor
  • ***
  • Posts: 1559
  • Country: gb
Re: Video Review & Teardown of the Analog Arts SL987 USB Combo Instrument
« Reply #24 on: February 10, 2015, 11:00:17 am »
The Intronix LogicPort USB logic analyser also uses a type A connector, and they supply a typeA/typeA cable with it. I've never seen this done elsewhere before.
 