This power supply was bought new for the very small price of 50 euros (including VAT) from a dealer in a electronics fair.
He had lots of them, and told me that he was buying them in 100 pieces lots from the manufacturer.
I had urgent need of using it to test some other equipment, so I put it to work immediately after opening the box. It operated without any evident problem, the only negative thing was noise from the internal cooling fan, always on.
After the initial hurry, I put it on one side of the instrument rack, and forgot about it.
Some days ago I decided to open it and install a thermal switch to control the fan, since it was mostly used at low current, and I though that the fan could be off most of the time, for a better environment it the lab.
The inside of the unit was full of surprises, some off them really nasty,
including some wiring faults that presented real safety hazards. The unit: it's a Chinese clone of another Chinese clone…. made by the
HK Longwey company. It was really difficult to find the manufacturer's name, because it is labeled LWDQGS… in any case I traced the company's site:
http://en.hklongwei.com/newEbiz1/EbizPortalFG/portal/html/index.htmlThey have a big range of simple laboratory instruments, each of them resembling a big-name model.
The unit can supply a nominal 30 Vdc at 5 A max.
A couple of rotary pots (coarse and fine) sets output voltage, another couple sets current limit value. A pushbutton switch, not well documented in the too simple user's manual, changes the current limit range from 0 to 5 A (HI) to 0 to 2.5 A (LO). I see no real use for this…
Two LEDs show operation mode (CV or CC).
Actual voltage and current values are displayed by two 3-digits large LED displays.
Output terminals are 5-way binding posts, that do not (at least I believe they do not) respect actual safety regulations.
Opening the case shows a very large and hefty transformer, that does not show signs of overheating even when heavily loaded, and a lot of cheaply-made, single sided circuit boards.
The dual meters, based on the classic 7107 ic, are assemble on two boards interconnected by a flat ribbon cable, and there are wires everywhere….
I cannot understand the economics of this choice: a smaller dual sided board could handle both meters, and the final cost should be lower, even in Hong Kong or China.
The main board shows three 8-pins ICs and more than 10 transistors.
I could not see the details because the wiring did not allow to reach near it, but two relays were obviously used to switch the transformer's secondary windings, to reduce power dissipation.
The heath sink was a simple piece of sheet aluminum, and the two power transistors were custom labeled units, coded: LW H2388C LONG WEI MEX.
This make me believe they have a great use of this parts…
The soldering to the power transistors pins was abysmal: see the photos.
One wire connecting the two collectors lugs was not well soldered, and it detached with a simple pull. If this happened in operation, the transistor (or both) could be easily damaged.
Wires (including mains) from unit's rear to unit's front
were placed inside a slot in the heat sink, that had sharp corners that could damage the insulation, and that can become really hot.
The wire from the protective hearth pin of the IEC power inlet was not soldered (or at least the joint was detached by a simple pull), but kept in place by heat-shrink tube that was kept in place by a blob of hot melt glue (hot glue was everywhere in the unit's inside).
I must say that I had verified continuity from mains-in PE terminal to instrument case ground before first use of the unit, and it tested OK: maybe the heath shrinking was helping continuity.
All mains-connected wires did not respect any safety regulation, in terms of colors and wiring/soldering method. Live, neutral and PE were all white….
I ended up with a new ground wire (yellow-green) with adequate heat-shrinking, and some mechanical protection for the other wires.
After reassembling the unit (whit the thermostat-controlled fan and the modified wiring), I carefully tested it with loads from 1 to 3 A .
The meters read correctly (at least +- 1 LSD)
Ripple was about 1 mV p-p (see photo) at any voltage and current, but increased tenfold in constant current mode.
In searching the Net, I've found a report on a Swedish site:
http://www.elektronikforumet.com/wiki/index.php/LW_PS-305DI translated it with Google, but not every thing was clear to me.
I understand that the tester measured a large overshoot when powering down.
I've recorded many on/off sequences with my DSO, and they seem correct (no overshoot) at any output voltage and with/without load.
Maybe the unit in Sweden had an unconnected wire….
What should I write as the last comment?
This unit is CHEAP, but crap-made.
There are safety risks in actual use.If you have one, open it up and check the wiring before further use.
If you don't have one, DO NOT buy it.
The problem that bothers me is: this manufacturer can supply OEM units with other logos on it, and you can end up with a more expensive, crappy unit, with big name on it…