Update (edited)
I finally decided to determine how long the 1.5V sensor supply must be toggled to reset the DRO scales to zero.
My reader is described here:
https://www.eevblog.com/forum/mechanical-engineering/mitutoyo-982-537-1-dro-display-with-series-572-linear-scales/msg5171796/#msg5171796I used a CD4066 NO analog switch at 5V VDD to toggle the 1.5V going to the scale* and a Rigol DG1032Z ro produce a zero pulse of various widths with a period of 10 s (i.e., high was set to 9.998 seconds for a 2-ms pulse). The on state was about 5V as well. It is important to have the 1.5V on for several seconds or the scale freezes up.
With a 1-ms pulse, zero would reset occasionally. A 2-ms pulse consistently reset to zero. Shorter pulses did not work.
EDIT: I was concerned that the circuit used in the experiments above didn't pull the 1.5V to ground. I got a pair of SPST NC and NO analog switches (TS12A4515 and TS12A4514, respectively) and wired them as a SPDT as shown in the attachment.** Transitions were much cleaner all the way to a 100-us pulse and reached ground. A pulse of 200us caused repeatable reset; 100us didn't.
John
*I initially tried to use the enable function of the MIC2951-adj voltage regulator. I could not get short pulses with that because of the relatively large output capacitor used. A pull-down load resistor was tired briefly, but was abandoned. I didn't have an NC switch, which would have made setting up the pulse generator a little simpler.
**The symbol for each switch has the correct pin label for com, but the diagram of the switch is opposite what I have usually seen. I was too lazy to change it. The switches are not directional.